Can anyone help me with my graphs

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
imrdyman
Posts: 21
Joined: Tue Feb 07, 2023 3:48 pm

Can anyone help me with my graphs

Post by imrdyman » Tue Feb 07, 2023 3:59 pm

Hello how do you do. Ive finally been able to sleep with my CPAP on but im still waking up tired. Was wondering if anyone can help me decipher what these numbers on the graphs mean. I dont have a resmed or phillips so its not OSCAR graphs so i wasnt able to get it all on one screenshot because the format is a bit strange. Also, theres a graph for "events" but it seems completely empty so i didnt include it. Can anyone take a quick peek and tell me whats going on. Thanks.



https://www.flickr.com/photos/197657176 ... W37z0957nK

User avatar
Miss Emerita
Posts: 3444
Joined: Sun Nov 04, 2018 8:07 pm

Re: Can anyone help me with my graphs

Post by Miss Emerita » Tue Feb 07, 2023 4:45 pm

Welcome! Could you tell us the make and model of your machine? Ideally you will also put that information, plus your mask information, in the signature block for your profile. That way it will be available to readers every time you post.

Was this your first night sleeping with the machine?

I'm having some trouble figuring out the "trend" view for your data. Maybe it means all your events happened within a short period of time??

At any rate, for the night you did pretty well by the numbers, with an AHI of 2.7, almost all of that from hypopneas. Many people here would say an AHI under 2 is good and under 1 is excellent.

Does your machine have a feature that allows the pressure to drop each time you exhale? Using that might be of some help with the hypopneas.

But what really matters once your numbers are pretty good is how you're feeling. If you're still in the early days, it's not at all surprising you're not feeling better yet. For most of us, it takes time. Your body may not be relaxing into good sleep quite yet; it can take time for it to get used to the new sensations of the mask and the pressure. You may also need some time to recover from "sleep debt."

I notice that the time you spent using the machine was 6.7 hours. I bet you weren't asleep for all that time, so you may have had only 6 or 6.5 hours of sleep. For a fair number of people, that isn't enough. Can you try arranging your schedule so you can sleep longer? If you could aim for 7.5 or 8 hours of actual sleep, that would be great.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

imrdyman
Posts: 21
Joined: Tue Feb 07, 2023 3:48 pm

Re: Can anyone help me with my graphs

Post by imrdyman » Tue Feb 07, 2023 5:05 pm

Miss Emerita wrote:
Tue Feb 07, 2023 4:45 pm
Welcome! Could you tell us the make and model of your machine? Ideally you will also put that information, plus your mask information, in the signature block for your profile. That way it will be available to readers every time you post.

Was this your first night sleeping with the machine?

I'm having some trouble figuring out the "trend" view for your data. Maybe it means all your events happened within a short period of time??

At any rate, for the night you did pretty well by the numbers, with an AHI of 2.7, almost all of that from hypopneas. Many people here would say an AHI under 2 is good and under 1 is excellent.

Does your machine have a feature that allows the pressure to drop each time you exhale? Using that might be of some help with the hypopneas.

But what really matters once your numbers are pretty good is how you're feeling. If you're still in the early days, it's not at all surprising you're not feeling better yet. For most of us, it takes time. Your body may not be relaxing into good sleep quite yet; it can take time for it to get used to the new sensations of the mask and the pressure. You may also need some time to recover from "sleep debt."

I notice that the time you spent using the machine was 6.7 hours. I bet you weren't asleep for all that time, so you may have had only 6 or 6.5 hours of sleep. For a fair number of people, that isn't enough. Can you try arranging your schedule so you can sleep longer? If you could aim for 7.5 or 8 hours of actual sleep, that would be great.

Yes its a Resplus autocpap from Beyond medical and using a resmed n20 nasal cushion mask with memory foam, ive also updated it on my profile. Yes im having trouble figuring out the "trend" view as well, thats what was mostly confusing me. Overall it shows 2.7 but in the trend view it says 11.2 per hour so i was thinking maybe thats why im so tired. This is about my 6th night sleeping with the machine. Yes my machine has exhalation relief, right now i have it at 2. I find that the machine has trouble detecting my inhale when i have it at 3 (3 being the most exhalation relief it can provide). Do you know if the graphs for my leak rates and the one for "Flow" look normal? If they do then maybe i need to just give it time and hopefully ill feel better in the future. Thanks for your response

User avatar
Miss Emerita
Posts: 3444
Joined: Sun Nov 04, 2018 8:07 pm

Re: Can anyone help me with my graphs

Post by Miss Emerita » Wed Feb 08, 2023 11:10 am

The first page pretty clearly says your AHI for the night was 2.7, so that's the number I'd trust.

About leaks, I don't think we have a way to tell how you're doing. We don't know what level of leaks your machine can compensate for, and we don't know whether the leak values shown on the report include "intentional leaks," that is, necessary venting that allows the release of CO2.

About "flow": I'm not sure what part of the chart you're referring to. Tidal volume? That looks normal for someone who has a fairly large build.

What does the "static view" look like?
Oscar software is available at https://www.sleepfiles.com/OSCAR/

imrdyman
Posts: 21
Joined: Tue Feb 07, 2023 3:48 pm

Re: Can anyone help me with my graphs

Post by imrdyman » Wed Feb 08, 2023 2:46 pm

Miss Emerita wrote:
Wed Feb 08, 2023 11:10 am
The first page pretty clearly says your AHI for the night was 2.7, so that's the number I'd trust.

About leaks, I don't think we have a way to tell how you're doing. We don't know what level of leaks your machine can compensate for, and we don't know whether the leak values shown on the report include "intentional leaks," that is, necessary venting that allows the release of CO2.

About "flow": I'm not sure what part of the chart you're referring to. Tidal volume? That looks normal for someone who has a fairly large build.

What does the "static view" look like?

Oh ok. Yea it just seems to give me an average leak and a max leak. Is it difficult to see whats going on with leaks because of the obscure machine and program they having me using? I cant seem to get alot of info on them online hehe. According to google an acceptable leak rate is 20-24L/min so maybe im ok? If you click on the first link again, on the last picture, theres a graph thats labeled "flow" in between the "pressure" and "vleak" one. Ive tried to compare my graphs with the OSCAR graphs floating around here but im not sure what to make of them, they look a bit different. Also theres information that u can see if you hover over the graphs on my end, obviously it cant be done with just a screenshot. My doctor said he doesnt know much about cpap or sleep apnea, he was just able to prescribe one for me so im kind of on my own. Anyway, if its still all confusing its ok, ill try to do a bit of trial and error on my end with the pressure maybe that'll help. Heres a link to the "static view" if you still want to take a peek.

https://flic.kr/p/2ofW8zS

User avatar
ozij
Posts: 10142
Joined: Fri Mar 18, 2005 11:52 pm

Re: Can anyone help me with my graphs

Post by ozij » Wed Feb 08, 2023 11:41 pm

Yes im having trouble figuring out the "trend" view as well, thats what was mostly confusing me.
Looks like the person who did the programming didn't know what "index" stands for.

I don't see a reason to trust one number (AHI of 2.6) over the other (11.2 "times an hour"). Which is why the first thing I'd do it is ask to be shown the events "Event" in the curve view. It's not checked on the present image.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

imrdyman
Posts: 21
Joined: Tue Feb 07, 2023 3:48 pm

Re: Can anyone help me with my graphs

Post by imrdyman » Thu Feb 09, 2023 2:02 am

ozij wrote:
Wed Feb 08, 2023 11:41 pm
Yes im having trouble figuring out the "trend" view as well, thats what was mostly confusing me.
Looks like the person who did the programming didn't know what "index" stands for.

I don't see a reason to trust one number (AHI of 2.6) over the other (11.2 "times an hour"). Which is why the first thing I'd do it is ask to be shown the events "Event" in the curve view. It's not checked on the present image.
Yea im looking around at the manual for the program and theres alot of broken english, i think its a program and machine from China. When i do pull up the "event" graph its completely empty it doesnt show me any events. Im trying to look around for info on this machine and program but theres not alot. Maybe id be better off trying to get a more mainstream machine to get some more help from you folks, OSCAR graphs seem to be more useful.

User avatar
ozij
Posts: 10142
Joined: Fri Mar 18, 2005 11:52 pm

Re: Can anyone help me with my graphs

Post by ozij » Thu Feb 09, 2023 2:17 am

imrdyman wrote:
Thu Feb 09, 2023 2:02 am
Maybe id be better off trying to get a more mainstream machine to get some more help from you folks, OSCAR graphs seem to be more useful.
I'd do that if I was in your place.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Can anyone help me with my graphs

Post by zonker » Thu Feb 09, 2023 11:45 am

imrdyman wrote:
Thu Feb 09, 2023 2:02 am
Maybe id be better off trying to get a more mainstream machine to get some more help from you folks, OSCAR graphs seem to be more useful.
if you can afford to do so, i think you'd be much better off.

good luck!
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Miss Emerita
Posts: 3444
Joined: Sun Nov 04, 2018 8:07 pm

Re: Can anyone help me with my graphs

Post by Miss Emerita » Thu Feb 09, 2023 12:32 pm

Yes, maybe I'm too trusting on the "AHI." I think we'll never figure out those graphs without knowing what the units on x-axis are.

It would be great if you could get a different machine. The ResMed Airsense 10 Autoset is an excellent machine with an excellent track record. You could also get a ResMed Airsense 10 Autoset For Her, which has one additional mode of operation thought to be of benefit to many women and some men. The ResMed Airsense 11 Autoset came out fairly recently. It includes the For Her modality. I think the jury is still out on whether it is superior to the 10 machines.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

imrdyman
Posts: 21
Joined: Tue Feb 07, 2023 3:48 pm

Re: Can anyone help me with my graphs

Post by imrdyman » Thu Feb 09, 2023 2:41 pm

Yea the Resmed 10 is the one i was thinking about getting. Ah, well thanks for the replys

User avatar
Pugsy
Posts: 64020
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can anyone help me with my graphs

Post by Pugsy » Thu Feb 09, 2023 3:15 pm

Side note just in case you are thinking of buying from cpap.com.
Sign up for their "newsletter" aka soft spam and you will get notices about sales and promo codes.
The SuperBowl is coming up soon and if history repeats itself they may have another sale.
Last really good sale was last fall and the Card to Cloud AirSense 10 AutoSet went for around 500 bucks.
I don't know that they will do it again but they might.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

imrdyman
Posts: 21
Joined: Tue Feb 07, 2023 3:48 pm

Re: Can anyone help me with my graphs

Post by imrdyman » Fri Feb 10, 2023 5:47 pm

Pugsy wrote:
Thu Feb 09, 2023 3:15 pm
Side note just in case you are thinking of buying from cpap.com.
Sign up for their "newsletter" aka soft spam and you will get notices about sales and promo codes.
The SuperBowl is coming up soon and if history repeats itself they may have another sale.
Last really good sale was last fall and the Card to Cloud AirSense 10 AutoSet went for around 500 bucks.
I don't know that they will do it again but they might.
Yes i just signed up for their "Vip" or something and i was able to purchase it for 500. Nice! Should be here by Monday. Thank you

User avatar
Miss Emerita
Posts: 3444
Joined: Sun Nov 04, 2018 8:07 pm

Re: Can anyone help me with my graphs

Post by Miss Emerita » Fri Feb 10, 2023 9:38 pm

Oh, well done!
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
chunkyfrog
Posts: 34394
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Can anyone help me with my graphs

Post by chunkyfrog » Fri Feb 10, 2023 11:38 pm

(Doing the froggy happy dance)

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her