New mask - now getting dry mouth
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
New mask - now getting dry mouth
For about six months, I've been using a "Brand X" full face mask that I found on Amazon, which is similar to the Mirage Quattro -- see my avatar for a photo. The therapy has been good (AHI averages about 0.4), and while I would not call it comfortable, I've been wearing it more than nine hours a night, and it's tolerable. But recently, as an experiment, I bought an F&P Evora Full Face Mask and have been using that for about a week. I've not noticed any difference in the therapy -- the average AHI is about what it has been with the old mask -- but it's definitely more comfortable, and I'm awakened fewer times a night by mask leaks. The one downside is that I'm now waking up with a very dry mouth, which I was not experiencing with the other mask. Any insights or suggestions would be welcome.
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
Re: New mask - now getting dry mouth
Anyone have any ideas?
Re: New mask - now getting dry mouth
XYLIMELTS
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
Re: New mask - now getting dry mouth
Just to be sure it's the mask and not something else, can you try going back to the old mask for a couple of nights?
Unless.... unless you're sleeping deeper and your mouth drops open?
When I use a full face mask, my non-taped mouth only gets troublesomely dry if the mask is leaking....
Have you looked at the objective leak rate? Could it be that for some mask related reason you're leaking without being woken up by it?
I mean, maybe the leak is larger but doesn't wake you up because it makes a different amount of noise, or doesn't hit you where it can wake you?
These two sentences sound paradoxical.earlvillestu wrote: ↑Sun Feb 05, 2023 1:08 pmI'm awakened fewer times a night by mask leaks. The one downside is that I'm now waking up with a very dry mouth, which I was not experiencing with the other mask.
Unless.... unless you're sleeping deeper and your mouth drops open?
When I use a full face mask, my non-taped mouth only gets troublesomely dry if the mask is leaking....
Have you looked at the objective leak rate? Could it be that for some mask related reason you're leaking without being woken up by it?
I mean, maybe the leak is larger but doesn't wake you up because it makes a different amount of noise, or doesn't hit you where it can wake you?
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- Okie bipap
- Posts: 3552
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: New mask - now getting dry mouth
The mask in your avatar appears to go under your chin. If it did, it would help keep your mouth closed at night. My wife and I are both use a firm foam cervical collar with the Evora mask. The collar keeps our chin from dropping and our mouth from coming open. You can use a cervical collar or a chin strap to keep your mouth closed.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
Re: New mask - now getting dry mouth
That may be it. I hadn't pulled the SD card and loaded it into OSCAR since switching masks. Now that I did so, it shows a very high leak rate every night since the switch. They are silent leaks, compared to those with the other mask, when there were fewer leaks but they howled and awakened me.ozij wrote: ↑Tue Feb 07, 2023 8:56 am
When I use a full face mask, my non-taped mouth only gets troublesomely dry if the mask is leaking....
Have you looked at the objective leak rate? Could it be that for some mask related reason you're leaking without being woken up by it?
I mean, maybe the leak is larger but doesn't wake you up because it makes a different amount of noise, or doesn't hit you where it can wake you?
Thanks for the suggestion. I'll tighten down the straps and see if that helps.
- ChicagoGranny
- Posts: 14463
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: New mask - now getting dry mouth
Starting with the basics, you are breathing some through your mouth. As little as five minutes of mouth-breathing can cause a very dry mouth. Some suggestions:
1. Wear a foam cervical collar to help prevent jaw-drop. The collars also hold the neck in the optimal position for airway patency by preventing the chin from moving toward the chest. Otherwise, when the chin moves toward the chest, the airway tends to be crimped. (My pressure requirements are much less when wearing a collar.)
2. Practice holding the tip of the tongue on the roof of the mouth and keeping the mouth closed.
3. Tape your lips before you don the mask. I use a similar mask and tape with sleep strips - https://www.amazon.com/Advanced-Breathi ... ref=sr_1_5
4. The Evora should be sized and adjusted so that the bottom is near the lower lip. This can help prevent the lips from parting.
- nightnight
- Posts: 14
- Joined: Sun Jun 10, 2007 12:14 pm
Re: New mask - now getting dry mouth
That mask is HUGE.
Try a smaller size.
Try a smaller size.
- ChicagoGranny
- Posts: 14463
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: New mask - now getting dry mouth
Yes, in the avatar, he is wearing it underneath the chin. The proper size places the bottom of the mask between the chin and lower lip. In other words, in the labiomental groove.
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
Re: New mask - now getting dry mouth
I've ordered a cervical mask. I am wearing the Evora with the bottom just below the lower lip.ChicagoGranny wrote: ↑Tue Feb 07, 2023 2:24 pmStarting with the basics, you are breathing some through your mouth. As little as five minutes of mouth-breathing can cause a very dry mouth. Some suggestions:
1. Wear a foam cervical collar to help prevent jaw-drop. The collars also hold the neck in the optimal position for airway patency by preventing the chin from moving toward the chest. Otherwise, when the chin moves toward the chest, the airway tends to be crimped. (My pressure requirements are much less when wearing a collar.)
2. Practice holding the tip of the tongue on the roof of the mouth and keeping the mouth closed.
3. Tape your lips before you don the mask. I use a similar mask and tape with sleep strips - https://www.amazon.com/Advanced-Breathi ... ref=sr_1_5
4. The Evora should be sized and adjusted so that the bottom is near the lower lip. This can help prevent the lips from parting.
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
Re: New mask - now getting dry mouth
The mask pictured in the avatar is not the one I'm having mouth-breathing/dry-mouth troubles with. The small photo in the avatar may look like I'm wearing it under the chin, but the bottom of it is really between the chin and the lower lip.ChicagoGranny wrote: ↑Wed Feb 08, 2023 1:41 pmYes, in the avatar, he is wearing it underneath the chin. The proper size places the bottom of the mask between the chin and lower lip. In other words, in the labiomental groove.
If the Evora doesn't work out, I'll go back to the original, which was tolerable and gave me good results.
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
Re: New mask - now getting dry mouth
Last night I used a chinstrap that I had bought but never used. The leaks were reduced, my mouth was not quite as dry, and Mr. ResMed had a green smiley face for me this morning. I'll keep using the strap until the cervical mask arrives next week. I've also ordered some mouth tape. I'm going to do some experimenting to see what works best leak-wise and balance that against what is most comfortable.
Thanks, everyone, for the guidance.
Thanks, everyone, for the guidance.
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
Re: New mask - now getting dry mouth
First night with the cervical mask last night. Wow! What a difference! No dry mouth at all. Got 8 hours of good sleep. Spent much of the night on my back, which is the most comfortable position for me, but it usually drives my AHI very high. But the AHI for the night was under 0.2. I think this is the solution.
Re: New mask - now getting dry mouth
you keep posting cervical "mask".earlvillestu wrote: ↑Fri Feb 17, 2023 2:12 pmFirst night with the cervical mask last night. Wow! What a difference! No dry mouth at all. Got 8 hours of good sleep. Spent much of the night on my back, which is the most comfortable position for me, but it usually drives my AHI very high. But the AHI for the night was under 0.2. I think this is the solution.
you mean cervical "collar", right?
anyway, glad it helps!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- earlvillestu
- Posts: 64
- Joined: Mon Jul 18, 2022 9:40 am
Re: New mask - now getting dry mouth
Yes, cervical collar. My bad.