zero AHI on 4cm

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Seb-Sanfilippo
Posts: 37
Joined: Mon Jan 16, 2023 1:49 am

zero AHI on 4cm

Post by Seb-Sanfilippo » Mon Jan 30, 2023 12:33 pm

If I have zero AHI and no other event does that mean I don't have sleep apnea or UARS?
I only get a few events when I bump up the pressure to 6cm

https://sleephq.com/public/d61df18d-179 ... 53eca6f081

User avatar
lazarus
Posts: 1294
Joined: Sat Mar 12, 2022 3:21 pm
Location: NY

Re: zero AHI on 4cm

Post by lazarus » Mon Jan 30, 2023 2:03 pm

[Pugsy's following responses are better than what I had said.]
Last edited by lazarus on Mon Jan 30, 2023 5:06 pm, edited 1 time in total.
The people who confuse "entomology" and "etymology" really bug me beyond words.
---
A love song to a CPAP? Oh please!:
https://youtu.be/_e32lugxno0?si=W4W9EnrZZTD5Ow6p

User avatar
Pugsy
Posts: 64020
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: zero AHI on 4cm

Post by Pugsy » Mon Jan 30, 2023 2:35 pm

viewtopic.php?f=1&t=185935&p=1432964#p1432964
Just got the results of my psg from the doctor. She says there’s nothing to explain my extreme fatigue. She said there were a few hypotonias and some sleep fragmentation but according to her nothing that would cause such debilitating symptoms.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

Seb-Sanfilippo
Posts: 37
Joined: Mon Jan 16, 2023 1:49 am

Re: zero AHI on 4cm

Post by Seb-Sanfilippo » Mon Jan 30, 2023 2:46 pm

So what point are you making pugsy?
PSG negative so I am chasing clouds?

User avatar
Pugsy
Posts: 64020
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: zero AHI on 4cm

Post by Pugsy » Mon Jan 30, 2023 2:55 pm

Seb-Sanfilippo wrote:
Mon Jan 30, 2023 2:46 pm
So what point are you making pugsy?
PSG negative so I am chasing clouds?
No. Not trying to make any point at all.

All I was doing was giving the others who might read your post and need a bit of history about you so they can make a better more informed response to you. They probably wouldn't take the time to go looking at your past posts. Most people won't go digging in old posts. You don't have OSA....sleep study documented. Now as to what it might be that you do have....I have no friggin idea.

I was trying to help you get better and more accurate responses.

You might want to go remove that chip from your shoulder. It's not doing you any favors.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

Seb-Sanfilippo
Posts: 37
Joined: Mon Jan 16, 2023 1:49 am

Re: zero AHI on 4cm

Post by Seb-Sanfilippo » Mon Jan 30, 2023 3:02 pm

Sorry no chip on my shoulder, English is not my first language so I might come across all wrong

User avatar
Pugsy
Posts: 64020
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: zero AHI on 4cm

Post by Pugsy » Mon Jan 30, 2023 3:30 pm

Seb-Sanfilippo wrote:
Mon Jan 30, 2023 3:02 pm
Sorry no chip on my shoulder, English is not my first language so I might come across all wrong
That's okay. English is my first language and sometimes what I say comes across in not the way I meant it as well.

You don't have OSA but you are trying cpap. You are going to get nice low or 0.0 AHI numbers simply because you don't have OSA unless you are seeing some events flagged and you weren't sound asleep but that's a different thing.
I can understand wanting to try cpap but you are going to have to rely more on how you feel or don't feel than you could rely on the AHI or other numbers. Those numbers are going to be low anyway and not because of using the cpap.

I am sorry but I have no idea about your fatigue issues. It's not OSA but I suppose it could be some other sleep disordered breathing problem or maybe just a crappy sleep problem if you have extremely fragmented sleep.

I wish you the best though.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Pugsy
Posts: 64020
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: zero AHI on 4cm

Post by Pugsy » Mon Jan 30, 2023 3:32 pm

As far as UARS goes....nothing the machine can or will show you will help you make that diagnosis.
It's a difficult diagnosis to make unless someone gets one of those in lab sleep studies using a Pes device.
Then you need another sleep study done with the Pes and cpap to get the right pressure if it is UARS.
The machine won't do a good job in auto mode for UARS treatment.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: zero AHI on 4cm

Post by zonker » Mon Jan 30, 2023 4:52 pm

Seb-Sanfilippo wrote:
Mon Jan 30, 2023 12:33 pm
If I have zero AHI and no other event does that mean I don't have sleep apnea or UARS?
I only get a few events when I bump up the pressure to 6cm

https://sleephq.com/public/d61df18d-179 ... 53eca6f081
zero ahi means you PERFECTLY treated your sleep apnea FOR THAT NIGHT. and that is all that it means.

we don't sleep the same way every night. you will have different numbers on your ahi on different nights. it doesn't mean that when you get zero ahi that you don't still need treatment. you do.

does that make sense to you?
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

dataq1
Posts: 820
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: zero AHI on 4cm

Post by dataq1 » Mon Jan 30, 2023 7:17 pm

Did you get your sleep study results?
Will you share that report?
Did you sleep study report a ZERO AHI?
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

Seb-Sanfilippo
Posts: 37
Joined: Mon Jan 16, 2023 1:49 am

Re: zero AHI on 4cm

Post by Seb-Sanfilippo » Tue Jan 31, 2023 10:45 am

Here is a link to my PSG, it's in french but maybe being a standard thing you might make sense of it.

https://pdfhost.io/v/dB.~uDLXy_Sanfilip ... ta_1201231

I just translated the following:

Total awakenings : 11
Changes of stages: 37
Sleep efficiency: 49.9%

Neurological events
Numbers of Micro awakenings 104 (0 after movements, 0 after apnea or Hypopnea)
Numbers of myoclonus: 61 (33 in non REM, 28 in REM)
Index of awakenings + micro awakenings: 23.8/H
(not sure what is) Index PLM on total sleep time: 12.6

Respiratory events base on the sleep stages and position
no supine
in REM: zero events
in non-Rem: 27 events

Obsturctive Hypopnea: 27 , Index 5.6

IAH: 5.6/h
IAHO : 5,6/h
IAHC : 0,0/h
IAO: 0,0/h
non-REM : 0.0/h
REM : 21.7/h

no brutism
no snoring
no leg movement
no flow limitation

Desaturation: 10.3/h

Conclusion: Mild apnea (IAHO: 5.6/H, IDO3: 10,3/) linked to REM sleep (IAHOREM: 21.7/h)

Slight insomnia (to get to sleep)
Sever insomnia (to stay asleep)

No indication for CPAP
Go to ENT (which I’ve done already) I’ve swollen turbinates

Treat insomnia with sleep hygene therapy and medication

User avatar
robysue1
Posts: 897
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: zero AHI on 4cm

Post by robysue1 » Tue Jan 31, 2023 3:50 pm

Seb-Sanfilippo wrote:
Tue Jan 31, 2023 10:45 am
Respiratory events base on the sleep stages and position
no supine
in REM: zero events
in non-Rem: 27 events

Obsturctive Hypopnea: 27 , Index 5.6

IAH: 5.6/h
IAHO : 5,6/h
IAHC : 0,0/h
IAO: 0,0/h
non-REM : 0.0/h
REM : 21.7/h

no brutism
no snoring
no leg movement
no flow limitation
So there's very mild sleep disordered breathing.

And
Seb-Sanfilippo wrote:
Tue Jan 31, 2023 10:45 am
Total awakenings : 11
Changes of stages: 37
Sleep efficiency: 49.9%

Neurological events
Numbers of Micro awakenings 104 (0 after movements, 0 after apnea or Hypopnea)
Numbers of myoclonus: 61 (33 in non REM, 28 in REM)
Index of awakenings + micro awakenings: 23.8/H

...

Slight insomnia (to get to sleep)
Sever insomnia (to stay asleep)
And there's a real, genuine problem with sleep maintenance insomnia, along with a minor problem with sleep onset insomnia.

Sleep efficiency of 49.9% is more than enough to explain why you would feel exhausted all the time.


Finally:
No indication for CPAP
Go to ENT (which I’ve done already) I’ve swollen turbinates

Treat insomnia with sleep hygene therapy and medication
Take the recommendation to treat the insomnia seriously. If you don't want to take prescription sleeping medication, ask for a referral to someone who does cognitive behavior therapy for insomnia or CBT-I for short. Note, however, CBT-I requires some real dedication on the patient's part, along with realistic expectations. And it's a long-term solution---in other words, don't expect to magically feel better after a night or two. It could take 2-4 months for CBT-I to make a noticeable difference in how you are feeling.

Good luck on dealing with the insomnia.
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls