Another "newbie" ? - I am hearing gurgling sounds at times and was told to make sure the water in the reservoir isn't filled to the "max" line which it isn't.
Someone else said make sure there is no water in the hose which there isn't.
I am at loss on this one, any help is appreciated.
Gurgling noise
Gurgling noise
_________________
Machine: Luna G3 CPAP |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Re: Gurgling noise
The water tub should be at max OR BELOW.
Try to keep the machine lower than the mattress.
Water in the tube will cause a gurgling or popping noise. This is caused by condensation.
https://www.positivesleepsolutions.com/ ... nIPYE8oFwU
Try to keep the machine lower than the mattress.
Water in the tube will cause a gurgling or popping noise. This is caused by condensation.
https://www.positivesleepsolutions.com/ ... nIPYE8oFwU
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
- chunkyfrog
- Posts: 34390
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Gurgling noise
It can only be water.
Look again; remove and drain the hose.
It only takes a tiny bit of water--you may not even see it until you drain it out.
Look again; remove and drain the hose.
It only takes a tiny bit of water--you may not even see it until you drain it out.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Gurgling noise
Thank you for all your information, hopefully it will get better lol!
_________________
Machine: Luna G3 CPAP |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
- Okie bipap
- Posts: 3552
- Joined: Thu Oct 15, 2015 4:14 pm
- Location: Central Oklahoma
Re: Gurgling noise
Try to rout your hose so there are not dips in the hose. Any moisture will collect in these low spots.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: IPAP 20-25, ps 4, OSCAR software |
Growing old is mandatory, but growing up is optional.