Airsense 11 tank almost empty every night

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
JeffL
Posts: 502
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Airsense 11 tank almost empty every night

Post by JeffL » Tue Nov 29, 2022 12:40 pm

Has anyone noticed the Airsense 11 using more water than the Airsense 10?

I don't know if the local humidity suddenly dropped or what. I only had to refill the tub on the Airsense 10 every 3 days. I've had the Airsense 11 for 6 days and it's used nearly all 12 ounces every night. I've tried setting it for auto, I've tried putting in the same settings as I had on the 10, and I've tired lowering the climateline temperature and humidity. It still nearly runs out. I had the humidity set at 4 last night. I'll try 3 tonight.

_________________
MachineMask

User avatar
Pugsy
Posts: 64016
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 11 tank almost empty every night

Post by Pugsy » Tue Nov 29, 2022 1:23 pm

What is the ambient humidity in your house?
How do your leak numbers look?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
JeffL
Posts: 502
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: Airsense 11 tank almost empty every night

Post by JeffL » Tue Nov 29, 2022 2:41 pm

Pugsy wrote:
Tue Nov 29, 2022 1:23 pm
What is the ambient humidity in your house?
How do your leak numbers look?
I don't have a hygrometer so I don't know the inside humidity. I think it's low because the thermostat is set for 71, I'm still chilled and my hands feel dry. The outside humidity is 55% according to Yahoo WX.

Leak numbers since I've had the machine are 6,23,34,28,23,and 24L/min.

_________________
MachineMask
Last edited by JeffL on Tue Nov 29, 2022 8:03 pm, edited 1 time in total.

User avatar
jlsmithseven
Posts: 250
Joined: Mon Dec 20, 2021 6:18 pm

Re: Airsense 11 tank almost empty every night

Post by jlsmithseven » Tue Nov 29, 2022 3:10 pm

My first one that came with the machine leaked like crazy, I just got a new one and it's so much better.

User avatar
JeffL
Posts: 502
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: Airsense 11 tank almost empty every night

Post by JeffL » Tue Nov 29, 2022 4:26 pm

jlsmithseven wrote:
Tue Nov 29, 2022 3:10 pm
My first one that came with the machine leaked like crazy, I just got a new one and it's so much better.
What leaked, the mask or the tub?

_________________
MachineMask

User avatar
jlsmithseven
Posts: 250
Joined: Mon Dec 20, 2021 6:18 pm

Re: Airsense 11 tank almost empty every night

Post by jlsmithseven » Tue Nov 29, 2022 5:18 pm

The tub. I think from the water heated under the base, the seal of that got through the base and it would just leak.

User avatar
JeffL
Posts: 502
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: Airsense 11 tank almost empty every night

Post by JeffL » Tue Nov 29, 2022 7:44 pm

jlsmithseven wrote:
Tue Nov 29, 2022 5:18 pm
The tub. I think from the water heated under the base, the seal of that got through the base and it would just leak.
I guess there can be a manufacturing defect once in awhile. I've never had a tub leak, but I've had water come through the air hose if I go to far past the MAX line. :lol:

_________________
MachineMask

jsmit86
Posts: 222
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: Airsense 11 tank almost empty every night

Post by jsmit86 » Tue Nov 29, 2022 7:58 pm

jlsmithseven wrote:
Tue Nov 29, 2022 5:18 pm
The tub. I think from the water heated under the base, the seal of that got through the base and it would just leak.
Is this an Airsense 10, or Airsense 11?
Leaks in the 10's are not uncommon.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

User avatar
Pugsy
Posts: 64016
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Airsense 11 tank almost empty every night

Post by Pugsy » Tue Nov 29, 2022 8:11 pm

Do you still have the AirSense 10 machine to use to see if the water consumption drastically changes?

3 things affect water consumption

1..ambient humidity....this is a biggy....the drier the ambient humidity the more water will be consumed if someone doesn't change the number 2
2..humidity setting
3..leaks when excessive and prolonged.....and also pressure needs..the higher the pressure the more water will be consumed.

While it wouldn't be impossible for there to be some sort of mechanical issue I would first suspect ambient humidity changes.
Ambient air became drier which then means the machine has to use more water to keep the humidity at the selected setting.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
JeffL
Posts: 502
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: Airsense 11 tank almost empty every night

Post by JeffL » Thu Dec 01, 2022 12:14 pm

Pugsy wrote:
Tue Nov 29, 2022 8:11 pm
Do you still have the AirSense 10 machine to use to see if the water consumption drastically changes?

3 things affect water consumption

1..ambient humidity....this is a biggy....the drier the ambient humidity the more water will be consumed if someone doesn't change the number 2
2..humidity setting
3..leaks when excessive and prolonged.....and also pressure needs..the higher the pressure the more water will be consumed.

While it wouldn't be impossible for there to be some sort of mechanical issue I would first suspect ambient humidity changes.
Ambient air became drier which then means the machine has to use more water to keep the humidity at the selected setting.
Last night's settings were 70 degrees with a humidity setting of 2. This morning I had a little over a half tank of water left. No discomfort, so I think I'll leave it there. I also think the weather has become a little more humid. Next time I'm out and about, I'll see if I can find a digital hygrometer so I'll know for sure. I'll order one if I can't find one.

The AS10 is packed up and stored away. Besides, a perfect A/B comparison would require me using both at the same time :?

_________________
MachineMask

User avatar
LSAT
Posts: 13232
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Airsense 11 tank almost empty every night

Post by LSAT » Thu Dec 01, 2022 1:55 pm

jsmit86 wrote:
Tue Nov 29, 2022 7:58 pm
jlsmithseven wrote:
Tue Nov 29, 2022 5:18 pm
The tub. I think from the water heated under the base, the seal of that got through the base and it would just leak.
Is this an Airsense 10, or Airsense 11?
Leaks in the 10's are not uncommon.
I would think a leak would be obvious..there would be a puddle.

jsmit86
Posts: 222
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: Airsense 11 tank almost empty every night

Post by jsmit86 » Thu Dec 01, 2022 7:50 pm

When the tub had just started to develop the leak, only the bottom was a bit wet... I wasn't sure if it was leaking, or perhaps condensation. Later, there was no doubt it was leaking. One time it dumped water all over my nightstand. I ended up replacing the machine with an Airsense 11 autoset. Not even a hint of a leak on that unit so far.

A few internet searches will show multiple articles and videos specifically showing exactly why the seal on the bottom edge of the Airsense 10 tub sometimes develops leaks. My machine seemed to get progressively worse with age. My son''s DME said that he believes some machines had the plates develop an uneven heating issue over time, that accelerated the problem. That seemed likely with my Airsense 10 unit. Every tub I purchased in the last 18 months or so, developed a leak with a few weeks. I know that doesn't happen to everybody, but with my machine continually doing that, I opted for replacement. If that had not been an issue, I would still be using the Airsense 10. Both machines are excellent in by view. I just had this weird problem.

So far, the Airsense 11 does not seem to be prone to the leaking tub issue, probably due to the design change which is a bit of a mixed bag for other reasons.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

Laos Film Director
Posts: 71
Joined: Tue Mar 16, 2010 3:01 pm
Location: Los Angeles

Re: Airsense 11 tank almost empty every night

Post by Laos Film Director » Fri Dec 02, 2022 11:32 am

My tank needs to be refilled every 2 days when I use a Climateline hose. I've switched to the standard hose and the water in the tank lasts much longer.

_________________
MachineMask

User avatar
chunkyfrog
Posts: 34390
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Airsense 11 tank almost empty every night

Post by chunkyfrog » Fri Dec 02, 2022 4:19 pm

Central heat reduces indoor humidity.
Hygrometers are inexpensive, and useful for predicting water usage.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
ejbpesca
Posts: 214
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Re: Airsense 11 tank almost empty every night

Post by ejbpesca » Sun Dec 04, 2022 9:43 am

My AirSense 10 will use up a tank in high humidity house within 5 hours so I still wake with an incredibly dry mouth.

_________________
MachineMask