AurCurve 19 ASV - my exhalations are not venting
AurCurve 19 ASV - my exhalations are not venting
For a year I've been on cpap/bipap/ now asv with treatment emergent Apnea. Recently I noticed my f20 mask elbow exhaust valve isn't working. When I was on cpao/BiPAP I had exhaust flow when I exhaled. The little flaps are inside the elbow joint but never close to vent my exhalations so I exhale into the climateline tube and rebreath. Is that how it's supposed to work? Is it safe to rebreath all night long?
Re: AurCurve 19 ASV - my exhalations are not venting
I assume AirCurve 19 is a slip and you meant Aircurve 10 ASV.dnuffer wrote: ↑Tue Nov 29, 2022 7:26 amFor a year I've been on cpap/bipap/ now asv with treatment emergent Apnea. Recently I noticed my f20 mask elbow exhaust valve isn't working. When I was on cpao/BiPAP I had exhaust flow when I exhaled. The little flaps are inside the elbow joint but never close to vent my exhalations so I exhale into the climateline tube and rebreath. Is that how it's supposed to work? Is it safe to rebreath all night long?
The little flaps are supposed to be closed, in order for you to have the pressure you need, supplied through your tube.
The exhaled air - when therapy is working properly - does not, repeat not go out through those openings, but rather through the little holes in the elbow.
The flaps open when there's no pressure, in order to ensure your exhalations escape, and fresh air comes in.
I don't know why the flaps in your F20 don't close when your machine is on, but you have to get that taken care of.
With the flaps/vents open, the machine does not give you the therapy you need.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: AurCurve 19 ASV - my exhalations are not venting
Thanks very much.
Re: AurCurve 19 ASV - my exhalations are not venting
If you were re-breathing your exhaled breath all night long, you'd be dead.dnuffer wrote: ↑Tue Nov 29, 2022 7:26 amFor a year I've been on cpap/bipap/ now asv with treatment emergent Apnea. Recently I noticed my f20 mask elbow exhaust valve isn't working. When I was on cpao/BiPAP I had exhaust flow when I exhaled. The little flaps are inside the elbow joint but never close to vent my exhalations so I exhale into the climateline tube and rebreath. Is that how it's supposed to work? Is it safe to rebreath all night long?
If you're dead, and posting, let us know, because we'd like to notify the news media.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- Respirator99
- Posts: 341
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: AurCurve 19 ASV - my exhalations are not venting
Ozij, I just had a close look at my F20 and I think you've got it a bit back-to-front. The flaps remain open at all times except when there is no pressure from the machine and you exhale. The flaps remain open at all times under machine pressure, both inhale and exhale. They only close if the machine is not generating pressure. When the machine is running there is enough pressure to keep the flaps open. As you said, the exhaled breath goes through the elbow vent (tiny holes).
The idea is that the airflow from the machine is sufficient to maintain therapy pressure while continually purging some of the machine's airflow and your exhalation out through the elbow vent. If the machine goes off, there is no pressure coming up the tube but the pressure of your exhalation is sufficient to close the flaps and cause the exhaled air to exit through the vent.
The idea is that the airflow from the machine is sufficient to maintain therapy pressure while continually purging some of the machine's airflow and your exhalation out through the elbow vent. If the machine goes off, there is no pressure coming up the tube but the pressure of your exhalation is sufficient to close the flaps and cause the exhaled air to exit through the vent.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
Re: AurCurve 19 ASV - my exhalations are not venting
Maybe its a question of terminology. (as in "Is the glass half empty or half full?")Respirator99 wrote: ↑Fri Dec 02, 2022 12:19 amOzij, I just had a close look at my F20 and I think you've got it a bit back-to-front. The flaps remain open at all times except when there is no pressure from the machine and you exhale. The flaps remain open at all times under machine pressure, both inhale and exhale. They only close if the machine is not generating pressure. When the machine is running there is enough pressure to keep the flaps open. As you said, the exhaled breath goes through the elbow vent (tiny holes).
I took another look at my F30 elbow - here's what I mean:
The F30 (F20) use the same elbow.
The elbow has a large openings on either side.
There are flaps at the elbow's angle, and air can flow all around them. They leave that opening open unless there's pressure in the system.
When there's pressure, the flaps get pressed onto those large openings - closing them. If they don't, you get noise, and not enough positive air pressure. When the large openings are closed, the exhaled air only goes out through the tiny holes.
When there's no pressure, the flaps fall down. You breath in fresh in through those large opening on the side and it flows in around those flaps. Breathing out may close the flaps, but they drop back down easily from the large side opening when you inhale. There's room for air to come in all around them. I don't consider their natural position (not under pressure position)"closed" - because in that position, they don't stop your inhale of fresh air.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
- Respirator99
- Posts: 341
- Joined: Mon Jul 05, 2021 12:39 am
- Location: Australia
Re: AurCurve 19 ASV - my exhalations are not venting
Ozij, I had another look at the elbow after reading your explanation and it didn't work the way I thought. You are right. My apologies.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
* Oscar help
* An alternative to Oscar - try SleepHQ
I have no medical training or qualifications. Take my advice for what it's worth.
- chunkyfrog
- Posts: 34390
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: AurCurve 19 ASV - my exhalations are not venting
The tiny holes are for exhaust.
The big flappy things are the anti-asphyxia valve, which, as mentioned,
only moves out of the way to keep you from suffocating
when the electricity goes off.
The big flappy things are the anti-asphyxia valve, which, as mentioned,
only moves out of the way to keep you from suffocating
when the electricity goes off.