Hi, I’ve been using my AirSense 11 Autoset with the AirFit F30i for a few weeks and things were great… until the whapping.
For the first few weeks, when the machine started (at a 4), the mask would make an annoying little “whap” just as I began to inhale. It would do this for a few breaths and then it would stop and everything would be fine. Sleep! Amazing!
Last night, though, the “whap” didn’t stop. Every inhale, I got the whap. I checked the hoses, the mask fit… everything seems fine. But I can’t use the thing Please help me stop the whappening!
Problem with quick burst of air on inhale
Problem with quick burst of air on inhale
_________________
Machine: AirSense 11 Autoset |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
Re: Problem with quick burst of air on inhale
On my F30, which is very similar to the F30i, the "whap" happens when the mask's valves snap shut.
The valves are two flaps in the elbow - they're open when you start therapy, and "whap" shut when the airflow starts.
Could it be that the elbow isn't connected firmly enough to the mask? Did you try connecting and reconnecting from the mask side, not the hose side?
Could be that there are water drops in the elbow after you washed it? If so. try running the mask fit test a number of times - its' a great way of drying masks and their parts.
Another possibility is that you need to take the humidifier fully out and reseat it.
Download OSCAR to see if there's another source of leak or if the machine is dropping pressure for another reason.
You'll find the necessary instructions here:
viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html
The valves are two flaps in the elbow - they're open when you start therapy, and "whap" shut when the airflow starts.
Could it be that the elbow isn't connected firmly enough to the mask? Did you try connecting and reconnecting from the mask side, not the hose side?
Could be that there are water drops in the elbow after you washed it? If so. try running the mask fit test a number of times - its' a great way of drying masks and their parts.
Another possibility is that you need to take the humidifier fully out and reseat it.
Download OSCAR to see if there's another source of leak or if the machine is dropping pressure for another reason.
You'll find the necessary instructions here:
viewtopic/t172378/Sticky--Newbies-PLEAS ... STING.html
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: Problem with quick burst of air on inhale
Thanks so much! I’ll try everything here!
_________________
Machine: AirSense 11 Autoset |
Mask: ResMed AirFit F30i Full Face CPAP Mask with Headgear |
- chunkyfrog
- Posts: 34394
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Problem with quick burst of air on inhale
Many years ago, when I was using a full face mask, we had a power outage.
The slapping sound of the AA (anti-asphyxia valve) just doing its thing, woke me.
The slapping sound of the AA (anti-asphyxia valve) just doing its thing, woke me.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
- VinnieBarbarino
- Posts: 69
- Joined: Mon Oct 24, 2016 2:17 pm
- Location: Kansas City, USA
Re: Problem with quick burst of air on inhale
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Backup/Travel: DeVilbiss IntelliPAP Auto with SmartFlex and Heated Humidifier |
Proud recipient of multiple Gold Badges from ResMed MyAir
Re: Problem with quick burst of air on inhale
Great video Vinnie. Good share. thanksVinnieBarbarino wrote: ↑Mon Nov 28, 2022 10:52 pmOh great, now I've got that Supremes tune in my head: https://www.youtube.com/watch?v=_VuxfQMsUK4
_________________
Machine: AirSense 11 Autoset |
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear |
Additional Comments: Newbie who loves her machine! |
Beware the schoolyard bullies, mean girls, and fragile male egos. Move along if you can’t be kind.