ResMed AirSense 10 No cell connection

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
JeffL
Posts: 502
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

ResMed AirSense 10 No cell connection

Post by JeffL » Thu Nov 24, 2022 11:49 am

First let me say that the new AirSense 11 that I got yesterday is working fine. It has a cell connection and it uploaded data that appears on MyAir.

Now, what maybe considered looking up up a dead horse's butt:

I haven't looked at MyAir since I don't remember when. I forgot about it until I got my new machine. When I did check it yesterday, I saw that there hasn't been any data recorded since 4/28/22. I got more curious about it this morning, and plugged it in. When the screen came up, it showed a line through a circle where the signal strength bars should be. My new 11 shows bars, so I know I have good enough signal strength. Besides, it worked fine in the same location until 4/28. I wonder if there is a setting to make it work again that would possibly send the missing data.

At worse, all is not lost. The data was written to the SD card and I was able to download it to ResScan.

_________________
MachineMask

User avatar
Jlfinkels
Posts: 468
Joined: Sun Jun 20, 2021 3:51 pm

Re: ResMed AirSense 10 No cell connection

Post by Jlfinkels » Thu Nov 24, 2022 12:12 pm

JeffL wrote:
Thu Nov 24, 2022 11:49 am
Now, what maybe considered looking up up a dead horse's butt:
One is led to ask, why would you do that?
JeffL wrote:
Thu Nov 24, 2022 11:49 am
I haven't looked at MyAir since I don't remember when. I forgot about it until I got my new machine. When I did check it yesterday, I saw that there hasn't been any data recorded since 4/28/22. I got more curious about it this morning, and plugged it in. When the screen came up, it showed a line through a circle where the signal strength bars should be. My new 11 shows bars, so I know I have good enough signal strength. Besides, it worked fine in the same location until 4/28. I wonder if there is a setting to make it work again that would possibly send the missing data.

At worse, all is not lost. The data was written to the SD card and I was able to download it to ResScan.
Did you associate the new machine with MyAir? You need to go into MyAir, select “profile”, then “My Equipment”, then “update machine”. You will then enter the machine serial number and it will be associated with your account.
Sometimes it is the very people who no one imagines anything of who do the things that no one imagines

User avatar
Dog Slobber
Posts: 3961
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: ResMed AirSense 10 No cell connection

Post by Dog Slobber » Thu Nov 24, 2022 12:18 pm

Your Air10 likely had an older 3g cell radio, last april the 3g cell towers were decommissioned. Your Air11 uses 4g.
Last edited by Dog Slobber on Thu Nov 24, 2022 12:19 pm, edited 1 time in total.

User avatar
Pugsy
Posts: 64026
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed AirSense 10 No cell connection

Post by Pugsy » Thu Nov 24, 2022 12:19 pm

My best guess for a reason the old machine stopped transmitting data the last of April is that it has a 3 G cellular modem thingy in it and that's about when 3 G cellular was simply discontinued and put out to pasture.

All the older machines that had 3 G cellular simply quit transmitting data.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
JeffL
Posts: 502
Joined: Fri Sep 16, 2011 10:02 pm
Location: Los Angeles
Contact:

Re: ResMed AirSense 10 No cell connection

Post by JeffL » Thu Nov 24, 2022 1:39 pm

The 3G issue makes a lot of sense since I had to replace our cell phone. I'm going with that :mrgreen: Thanks!
Jlfinkels wrote:
Thu Nov 24, 2022 12:12 pm
JeffL wrote:
Thu Nov 24, 2022 11:49 am
Now, what maybe considered looking up up a dead horse's butt:
One is led to ask, why would you do that?
Just in case there is a slim possibility that you aren't familiar with horse expressions :lol:



"Meaning of LOOK UP A DEAD HORSE'S ASS in English

do a worthless task, do a pointless exercise, pissing into the wind".


If I were to continue to pursue it, you could say I was beating a dead horse :lol:

waste energy on a lost cause or a situation that cannot be changed.

_________________
MachineMask

User avatar
JoyD.
Posts: 525
Joined: Sat Feb 14, 2009 7:21 pm
Location: Central Indiana, USA

Re: ResMed AirSense 10 No cell connection

Post by JoyD. » Fri Nov 25, 2022 3:15 am

.
I understand that some AirSense 10's use 3G, and some 4G. -- Is there a way to tell if one's AirSense 10 uses 3G or 4G?

_________________
Machine: Airsense 10 Card to Cloud
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: HumidAire H4i™ Heated Humidifier
Additional Comments: CPAP Pr 13.8; EPR:3 Full; Heated Hose; Nexcare blue mouth-tape; PurSleep Buckwheat Hull Pillow; Caldera Releaf Collar.
Other Accessories & Software: Wellue O2 Ring; OSCAR; SleepHQ Pro.

User avatar
Respirator99
Posts: 343
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: ResMed AirSense 10 No cell connection

Post by Respirator99 » Fri Nov 25, 2022 3:24 am

JoyD. wrote:
Fri Nov 25, 2022 3:15 am
.
I understand that some AirSense 10's use 3G, and some 4G. -- Is there a way to tell if one's AirSense 10 uses 3G or 4G?
In the Options screen scroll all the way down to About. Click that, then scroll all the way down to Internal Modem. You'll find it there.
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

User avatar
Dog Slobber
Posts: 3961
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: ResMed AirSense 10 No cell connection

Post by Dog Slobber » Fri Nov 25, 2022 7:50 am

JoyD. wrote:
Fri Nov 25, 2022 3:15 am
.
I understand that some AirSense 10's use 3G, and some 4G. -- Is there a way to tell if one's AirSense 10 uses 3G or 4G?
On the back of the machine will be an IC number and an FCC ID. If it ends with:
  • CDMA - it's a real old device with pre-LTE cell
  • 3G - it's 3G
  • 4G - it's 4G

User avatar
JoyD.
Posts: 525
Joined: Sat Feb 14, 2009 7:21 pm
Location: Central Indiana, USA

Re: ResMed AirSense 10 No cell connection

Post by JoyD. » Fri Nov 25, 2022 2:09 pm

Respirator99 wrote:
Fri Nov 25, 2022 3:24 am
In the Options screen scroll all the way down to About. Click that, then scroll all the way down to Internal Modem. You'll find it there.
Thanks, Respirator! I went into the Clinical Menu and followed your instructions to happily learn that I have 4G in my AirSense 10 !

_________________
Machine: Airsense 10 Card to Cloud
Mask: Zest Nasal CPAP Mask with Headgear
Humidifier: HumidAire H4i™ Heated Humidifier
Additional Comments: CPAP Pr 13.8; EPR:3 Full; Heated Hose; Nexcare blue mouth-tape; PurSleep Buckwheat Hull Pillow; Caldera Releaf Collar.
Other Accessories & Software: Wellue O2 Ring; OSCAR; SleepHQ Pro.