dataq1 wrote: ↑Tue Nov 15, 2022 4:33 pm
robysue1 wrote: ↑Tue Nov 15, 2022 11:55 am
above
Are you saying that SleepHQ is misrepresenting flow-rate curves? (FOT present in OSCAR's depiction but not in SleepHQ's depictions for the same time period)?
I'm saying that the two apnea-like "missed" events in lynninnj's data that she posted at
https://sleephq.com/public/3a22af32-0e9 ... a5466f4010 don't look like CAs to me. They look like OAs: There are no FOT oscillations in the flow rate and the machine responds to these two "missed" events by increasing the pressure right after they end.
I have no idea how these two events look in Oscar. But I would suspect they look much the same. If there had been oscillations in the flow rate graph, I don't think the machine would have increased the pressure since there was no activity to speak of in either the flow limitation graph or the snore graph that could explain the pressure increase.
The Mask Pressure graph as drawn at SleepHQ is clearly way, way off. As drawn in both ResScan and Oscar, the Mask Pressure graph shows the transition of pressure caused by EPR on each and every breath and the "height" of the bumps is roughly equal to the EPR setting---i.e. if EPR = 2, the height of the bumps from bottom to top will be approximately 2 cmH2O units in height. SleepHQ's Mask Pressure graph does not show the individual bumps once you zoom in close enough to see the individual breaths.
Here's a snippet of a CA flagged in my own Flow Rate data from last night, along with the Pressure graph and the Mask Pressure graph as displayed in Oscar:
Because you can't reorder the graphs in SleepHQ, I have to show these three graphs in two screen shots. The first screen shot shows the Pressure Graph and the Flow Rate Graph:

And here's the Mask Pressure Graph:
It should be quite clear that the Mask Pressure graph from SleepHQ is not displaying (all) the data that was written on the SD card correctly. So clearly you
can't use the Mask Pressure graph from SleepHQ to help check whether an event (or a "missed event") ought to be considered a "central" event or an "obstructive" event because you can't see the FOT oscillations in the Mask Pressure graph to compare them to oscillations in the Flow Rate graph.
It's also worth noting that I am using a bi-level machine---i.e. an AirCurve 10 VAuto running in Auto mode and PS = 4. That's why the bumps in the Mask Pressure graph in Oscar all have height about 4 cm H2O. The Oscar Pressure graph correctly shows the current "set" IPAP and EPAP pressures throughout this whole snippet, and the fact that the machine's program identifies this bit of troubled breathing as "central" explains why there is no pressure increase in response to this bit of bad breathing.
Finally, if you'd like to look at my data in SleepHQ, here's the link:
https://sleephq.com/account/teams/vWDQle