dataq1 wrote: ↑Fri Nov 11, 2022 12:42 pm
I understand what you are saying, but there is ONE
recommended standard (H3A) and ONE
alternate standard (H4A). Yes?
But Medicare, the biggest insurer of folks with OSA does
NOT accept standard H4a. And hence standard H4a is still a
AASM approved standard for scoring hypopneas. And yes, there are still labs out there that use H4a to score hypopneas on PSGs. And many AASM approved home sleep studies also use H4a to score hypopneas.
You seem to think that somehow the fact that H4a is an "alternate" means it is not a valid standard. But it is still a valid AASM approved standard.
Whether the AASM should push back against Medicare and inform them that in the collective eyes of the AASM, hypopneas scored under H3a are just as significant and just as bad as those scored under H4a and that Medicare ought to accept the collective opinions of the medical experts is another question altogether. But the fact that the AASM has
not pushed back on Medicare's insistence that the only hypopneas that count are those with 4% O2 desats means that there are folks in the AASM who either believe H4a is the "right" definition of a hypopnea or there are folks in the AASM who don't really care which definition is used.
The major difference between the two (other than the degree of desaturation), is that the H4A standard does not rely on arousals at all.
And that's my main problem with H4a: If you happen to arouse
before the desats happen, under H4a you can wind up being declared as "normal" because none of your hypopneas with arousal (which count under H3a) count under H4a. That's how you get someone like me: Under H3a, I have moderate OSA because my diagnostic PSGs in an accredited labs came up with H3a scored AHIs of 23 and 19. But if those PSGs had been scored under H4a, I would have been diagnosed as not having apnea at all because my H4a scored AHIs would have been 3.5 and 3.0.
In other words, the fact that the AASM still sanctions scoring PSGs under H4a means that there is not yet an overwhelming consensus that hypopneas with arousal (and no O2 desat) are "real" hypopneas that are as serious as those hypopneas scored under H4a.
Since the home machines currently have no reliable method for detecting arousals, the simplest standard for at-home machines to adopt is the H4 standard.
I disagree. Having a home CPAP machine adopt the H4 standard means that if a person's OSA is primarily caused by hyponeas with arousal and the machine's pressure is set too low to prevent those H3a hypopneas with arousal from happening, the person may well wind up with a machine reported AHI = 0.0 night after night, but in reality the number of their H3a hypopneas with arousal remain significantly high because the person actually needs more pressure to treat them than the current pressure settings allow.
And what would happen in that case? The patient can be fully compliant with xPAP, their "treated AHI" would be close to 0 (because they don't have any hypopneas with desats to begin with and the home machine is not scoring them under your scenario), but their real AHI (under H3a) remains essentially unchanged because the machine is not set high enough to prevent the H3a hypopneas with arousal from happening. So the person continues to arouse many, many times for each hour of sleep; their sleep is still badly fragmented; and they still suffer all the daytime symptoms that convinced them to get a sleep test in the first place.
In other words, making a home CPAP machine use standard H4a will result in the machine
under reporting the true number of events in every patient who has a significant problem with H3a hypopneas that are scored because of the
arousals. That can lead to
under treatment of the person's OSA because the machine scored H4a AHI will not score all those H3a hypopneas with arousal that
are scored if the machine is using only the wave flow to score hypopneas.
In other words, when a machine uses only the wave flow to score hypopneas, it will count both H4a and H3a hypopneas, along with some false events that occur when the patient is awake.
Using your idea of having a machine use oximeter data to score
only H4 hypopneas means the machine will miss every single H3a hypopnea "with arousal" that does not have a corresponding deast. Your idea might eliminate a few false positives---i.e. "scoring non-events as events", but it's going to also create a situation with a heck of a lot of false negatives---i.e. not scoring real hypopneas that would be scored under H3a because the person was asleep, but had an arousal before the desat occurred.
Statistically speaking,
over reporting the treated AHI by counting 'non-real' events is better for the patients than
under reporting created by routinely not scoring one whole category of event---i.e. not scoring H3a hypopneas with arousal that do not involve an O2 desat.
But current at home machines can't really adopt either standard because they lack the capability of detecting any desaturations at all.
Both standards are based on flow rate. The machines
can measure flow rate. By flagging any period of a >=30% reduction in flow that lasts at least 10 seconds as a "hypopnea", every
real hypopnea (under either H3a or H4a) will be counted. Yes, there will be some "false positives" thrown in. But that is not just because the oximeter data is missing. Just as important, perhaps more important, the EEG data is missing. The machines cannot tell when we're awake and when we're asleep. So anything that looks like sleep disordered breathing (including OAs and CAs as well as Hs) gets scored---even if we're wide awake with our eyes open.
But that's not a problem: The
real treated AHI, based on both oximeter and EEG data, as scored on a PSG in an accredited lab will be
less than or equal to the AHI calculated by the machine when it scores events based solely on flow rate. So if the machine's AHI is consistently low enough, we can correctly conclude the real, treated AHI is also low enough.
If you make the machine discount every >=30% flow reduction that lasts for at least 10 seconds just because there is not a corresponding (3% or 4%) O2 desat in a built-in oximeter, then
every real 3Ha hypopnea that has an arousal, but no corresponding desat will not be scored by the machine. In other words, for people with a lot of hypopneas with arousal, you wind up under counting the events. And now the machine's AHI is worthless because the machine's AHI could well be a
lower bound on the true status of your treated OSA rather than an upper bound. In other words, for people like me, the machine's AHI could be significantly lower than 5 while the real treated AHI might remain virtually unchanged if the pressure is not set sufficiently high enough to keep the hypopneas with arousal (but no desats) from happening while the person is asleep with the mask on your face.
BTW, If oximetry were built into at home machines, it could be a simple matter to select the threshold desaturation (3 or 4%) in machine setup.
That still will lead to a situation where H3a hypopneas
with arousal but no corresponding O2 desat are not counted. And that means you are
under counting the number of real hypopneas.
Relative to Medicare: Medicare (and likely the other insurers) rely on the sleep study results (that maybe either H3 or H4), not the on-going day-to-day home results. Since I'm addressing home results, Medicare position on H4 versus H3 has no bearing.
And I'm telling you that your ideas would mean that for a person like myself, there would be no way to tell if the machine is treating my apnea. If the only events that the machine flags are "hypopneas with a 3% or 4% (your choice) desat" then every one of my hypopneas that still occur while I'm using the machine would not be counted. In other words, in theory I could have a CPAP reporting an AHI = 0.0 while still having 15-20 hypopneas with arousal per hour. In other words I could have 15-20 places in each hour of sleep where there is A >= 30 reduction in flow for 10 seconds or more with a corresponding arousal, but no O2 desat while using the machine, and your machine would not count a single one of those in its AHI because not a single one of them had an O2 desat of 3%.