Mask Liners

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ejbpesca
Posts: 214
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Mask Liners

Post by ejbpesca » Mon Oct 03, 2022 10:11 am

I ordered a couple of white cotton mask liners from my supplier. The feel of the mask was nice, but the leak rate went way up so I stopped using the liner. Leak rate went back down.

The maker of the liners claims those leaks are insignificant small leaks that will not affect therapy.

Anyone have experience with the liners and leaks?

Also, the price of a little patch of cotton called a liner is high. I wonder if there are less expensive alternatives to the brand I got that were $15 for two.

_________________
MachineMask

User avatar
LSAT
Posts: 13232
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Mask Liners

Post by LSAT » Mon Oct 03, 2022 10:15 am

Most machines will compensate for leaks up to about 25-30L. If you are below that and you are sleeping well with a low AHI, ignore everything else.
Look at padacheek.com for specific liners for your mask. They are more expensive but can be washed and least a long time.

User avatar
Pugsy
Posts: 64021
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Mask Liners

Post by Pugsy » Mon Oct 03, 2022 10:22 am

ejbpesca wrote:
Mon Oct 03, 2022 10:11 am
the price of a little patch of cotton called a liner is high. I wonder if there are less expensive alternatives to the brand I got that were $15 for two.
You can make your own mask liners out of an old TShirt.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
ozij
Posts: 10142
Joined: Fri Mar 18, 2005 11:52 pm

Re: Mask Liners

Post by ozij » Mon Oct 03, 2022 10:32 am

Pugsy wrote:
Mon Oct 03, 2022 10:22 am
ejbpesca wrote:
Mon Oct 03, 2022 10:11 am
the price of a little patch of cotton called a liner is high. I wonder if there are less expensive alternatives to the brand I got that were $15 for two.
You can make your own mask liners out of an old TShirt.
Especially now that you have an example to use as a template.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
robysue1
Posts: 897
Joined: Sun Sep 18, 2022 3:39 pm
Location: Buffalo, NY

Re: Mask Liners

Post by robysue1 » Mon Oct 03, 2022 3:03 pm

ejbpesca wrote:
Mon Oct 03, 2022 10:11 am
I ordered a couple of white cotton mask liners from my supplier. The feel of the mask was nice, but the leak rate went way up so I stopped using the liner. Leak rate went back down.
As with so many other things in CPAP therapy, there's a tradeoff. And how you are feeling and sleeping is a critical part of the puzzle.

In your case:

How much did the leaks go up? As LSAT points out, our machines can compensate for an increase in unintentional leaking up to about 25-30 L/min without much trouble. If your leak rate didn't increase beyond that, you don't need to worry about the increase in leaks in terms of efficacy of therapy.

Did the increase in leaks wake you up? If you slept through the leaks and they were less than about 25-30 L/min, you don't need to worry about them. If they woke you up and you had trouble getting back to sleep because of the leaks, then that's a problem regardless of their size.

Did the mask liner make the mask more comfortable? If so, it's worth using them and using a week's worth of daily data to figure out whether the increase in leaking is worth worrying about in the long run.

Anything that makes you more comfortable is going to make it easier to get to sleep (and potentially stay asleep) with the mask on your face. So if you slept better with the mask liner, keep using it until and unless you have enough leak data gathered over at least a week that indicates you might actually have a problem with large leaks that are both large enough and long enough to potentially affect the efficacy of your therapy.
Joined as robysue on 9/18/10. Forgot my password & the email I used was on a machine that has long since died & gone to computer heaven.

Correct number of posts is 7250 as robysue + what I have as robysue1

Profile pic: Frozen Niagara Falls

User avatar
Okie bipap
Posts: 3553
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: Mask Liners

Post by Okie bipap » Mon Oct 03, 2022 5:28 pm

It is quite easy to make your own mask liner. I made mine out of an old, well-used tee shirt. I found thee pattern for the mask loner here:
http://montfordhouse.com/cpap/diy-cotton-mask-liner.pdf

This size liner will fit a medium or small sized cushion. It could easily be enlarged for a large cushion. The hole in the center can be enlarges as much as you want, as long as you leave enough material to go under the mask cushion.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

tbeb
Posts: 50
Joined: Sun Nov 29, 2020 9:43 am
Location: Indiana

Re: Mask Liners

Post by tbeb » Wed Oct 05, 2022 9:44 pm

I use a Pad A Cheek mask liner with a Phillips Dreamwear full face mask. I wash liner by hand with soap and water about once a week, and it lasts about 6 months. Total cost for the liner for my mask is $27. As I understand it, it wasn't designed to lessen leaks, but it has for me and I believe others.
Phillips Dreamstation with Humidifier, Phillips Dreamwear Full Face Mask

User avatar
chunkyfrog
Posts: 34394
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Mask Liners

Post by chunkyfrog » Wed Oct 05, 2022 9:54 pm

tbeb wrote:
Wed Oct 05, 2022 9:44 pm
I use a Pad A Cheek mask liner with a Phillips Dreamwear full face mask. I wash liner by hand with soap and water about once a week, and it lasts about 6 months. Total cost for the liner for my mask is $27. As I understand it, it wasn't designed to lessen leaks, but it has for me and I believe others.
Mask liners wick away sweat, and silence the noise caused by leaks.
As long as the final leak rate is within limits, it is acceptable;
especially if the patient is better able to sleep.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her