ejbpesca wrote: ↑Mon Oct 03, 2022 11:30 am
In reply to RobySue:
Let's see if I have the concept right.
Min pressure will continue if there are no events. This means either the min pressure is preventing them, or the patient (me) is not having events.
When events occur, pressure increases on an APAP, and throttles towards the max setting to prevent more events, and give relief to the current event. Higher pressure will remain for some period of time, especially with more events as it tries to give therapy to those events.
APAPs will increase the pressure when it detects
snoring,
flow limitations,
RERAs,
hypopneas and
obstructive apneas. The most
common reason a Resmed APAP is going to increase pressure is flow limitations: In other words, if the machine is not happy with the way you are breathing and it thinks the breathing pattern indicates the airway is at risk of collapsing or partially collapsing, it will increase the pressure in an effort to
prevent the serious events (hypopneas and obstructive apneas) from occurring the first place.
Trying to determine why the minimum pressure is maintained during a certain time period requires knowing about your
untreated apnea: If your untreated apnea was significantly worse during REM or on your back, then the minimum pressure might be maintained because you just aren't having events during that time period. But it could also be that the minimum pressure is enough to prevent events from occurring during the
parts of your sleep when the apnea is not at its worst. Without knowing what your diagnostic sleep study said about
when your events were occurring
before you started PAPing, there's no way to tell.
Aerophagia could be caused by both min and max pressures, but I suspect it is the jump up to the max pressure setting that is causing aerophagia.
It's a reasonable hypothesis to think that the aerophagia might be caused by the sudden increase in pressure when the machine decides to increase the pressure in response to a deteriorating breathing pattern. But in this case, we would expect that increasing the
minimum pressure from 4 to 6 should have not triggered a return of the aerophagia since the jump in pressure would have been less severe.
In general, the higher the minimum pressure is, the smaller the increase the machine needs to make in order to stabilize your breathing when it detects things are starting to go down hill. To try to make this clearer:
If your minimum pressure is at 4cm and your machine has not increased pressure in response to anything AND you go into REM and the breathing deteriorates rapidly, the machine may increase the pressure from 4 to your max of 11 in a matter of 5 or 10 minutes. That's in increase of 7cm over a relatively small amount of time.
On the other hand, if your minimum pressure is at 8 cm your machine has not increased pressure in response to anything AND you go into REM and the breathing deteriorates rapidly, the machine may increase the pressure from 8 to your max of 11 in a matter of 5 or 10 minutes. That's in increase of only 3 cm over the same amount of time.
It's also a reasonable hypothesis that your particular sphincter value is super-sensitive to pressure in the first place, and in that case you might do better with a lower minimum pressure since the lower minimum pressure seems to keep
most of your apnea under control. The question then becomes what is the
highest setting for the minimum pressure that your stomach can tolerate. And given that minimum pressure setting, how much additional pressure can the stomach take for
shorter periods of time to try to prevent nasty clusters of events from occurring?
The way to tease out which of the two hypotheses better fits you and your stomach is to cautiously change only
one thing at a time from the baseline 4-10cm range that did not seem to trigger the aerophagia problems and gather enough data at each pressure setting to be sure that you're not just looking at one particularly good or bad night in terms of both the apnea and the aerophagia.
Last night:
Once awake around 3am, I stayed up until 7am for 3 hours. I tried to get more sleep without therapy, but naps are some twilight zone that is not asleep and not totally awake. I think I may be drifting in and out, without knowing if I ever went to sleep. Miserable.
What happens if you try to nap with the machine? More aerophagia?
I discontinued 2 meds I had been taking daily for high blood pressure for ten years and two other meds for depression and anxiety I had taken for 30 years. I also discontinued potassium. My daily dose of a pain med was discontinued six years ago. So, that leaves only 1 med I take daily and that is Losartan for blood pressure. I do take other meds than the one, but they are on an as needed basis and some needed often.
Do any of those medications have known affects on your acid reflux problems?
So, each day with meds is different and yes, some do affect sleep. A nerve pain med can induce sleep. I have many permanent injuries that give me chronic pain that can leap to level 9 in an instant. The pain is part of why I do not sleep well.
It may be worth keeping a log to see if the chronic pain and bad sleep issues corollate with the worst of the aerophagia.
In other words, it's worth asking:
Is your aerophagia less severe (or absent) on the nights where the chronic pain doesn't leap up to a level 9 in an instant?
Is your aerophagia worse on the nights where the chronic pain flairs up significantly? Is it worse on the nights where the chronic pain's background level is higher?
Is your aerophagia less severe (or absent) on the nights where you subjectively feel you have gotten as decent of a night's sleep as you are capable of getting?
Is your aerophagia worse on nights when you subjectively feel you have gotten a rotten night's sleep for whatever reason?