AHI Total AI Centra AI

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ejbpesca
Posts: 214
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

AHI Total AI Centra AI

Post by ejbpesca » Tue Sep 20, 2022 7:34 am

In lay terms, could someone please tell me what the following means? Please, no acronyms.

Set to display six months, my Resmed Airsense 10 shows on its sleep report the following:

AHI 1.1
Total AI 0.8
Central AI 0.3

What does each individual term and its number mean? (definition)
By apnea standards are these good or bad scores?
Do these scores indicate my machine is working?
Do these scores indicate anything about my level of apnea? (low, med. high?)

A single day report may show AHI 0.3, Total AI 0.3, Central AI 0.3

_________________
MachineMask

User avatar
Pugsy
Posts: 64026
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: AHI Total AI Centra AI

Post by Pugsy » Tue Sep 20, 2022 7:41 am

ejbpesca wrote:
Tue Sep 20, 2022 7:34 am
What does each individual term and its number mean? (definition)
http://sleepyhead.sourceforge.net/wiki/ ... r_Glossary
ejbpesca wrote:
Tue Sep 20, 2022 7:34 am
By apnea standards are these good or bad scores?
very good
ejbpesca wrote:
Tue Sep 20, 2022 7:34 am
Do these scores indicate my machine is working?
yes
ejbpesca wrote:
Tue Sep 20, 2022 7:34 am
Do these scores indicate anything about my level of apnea? (low, med. high?)
These are TREATED scores and have zero relation as to what your pre cpap category of severity might be.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Miss Emerita
Posts: 3444
Joined: Sun Nov 04, 2018 8:07 pm

Re: AHI Total AI Centra AI

Post by Miss Emerita » Tue Sep 20, 2022 10:21 am

As you can see from the resource Pugsy linked to, AHI is apnea-hypopnea index, AI is apnea index, and Central AI is the index for central apneas only. "Index" is the number of events in the specified category divided by hours of sleep.

What about the obstructive apnea index? You can figure it out by subtracting the Central AI from the AI.

And what about the hypopnea index? You can figure it out by subtracting the AI from the AHI.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI Total AI Centra AI

Post by zonker » Tue Sep 20, 2022 10:49 am

Miss Emerita wrote:
Tue Sep 20, 2022 10:21 am
As you can see from the resource Pugsy linked to, AHI is apnea-hypopnea index, AI is apnea index, and Central AI is the index for central apneas only. "Index" is the number of events in the specified category divided by hours of sleep.

What about the obstructive apnea index? You can figure it out by subtracting the Central AI from the AI.

And what about the hypopnea index? You can figure it out by subtracting the AI from the AHI.
m'am?

i was told there'd be no math on the test.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ChicagoGranny
Posts: 14471
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: AHI Total AI Centra AI

Post by ChicagoGranny » Tue Sep 20, 2022 11:12 am

ejbpesca wrote:
Tue Sep 20, 2022 7:34 am
AHI 1.1
Total AI 0.8
Central AI 0.3
They are so good that you should be banned from knowing the definitions. :mrgreen:

dataq1
Posts: 820
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: AHI Total AI Centra AI

Post by dataq1 » Tue Sep 20, 2022 12:11 pm

ejbpesca wrote:
Tue Sep 20, 2022 7:34 am
Set to display six months, my Resmed Airsense 10 shows on its sleep report the following:

AHI 1.1
Total AI 0.8
Central AI 0.3

A single day report may show
AHI 0.3,
Total AI 0.3,
Central AI 0.3
What may be confusing (from the math standpoint) is the difference (in the six month report) between "AHI" and "Total AI" (1.1 and 0.8 ). One might expect that to imply a 0.3 Hypopnea index over that period and a 0.5 Obstructive Index.

One the day report, one might expect that to imply Zero Hypopneas/hour and Zero obstructions/hour; the only events that night were centrals- that the cpap cannot treat.

Correct?
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: AHI Total AI Centra AI

Post by palerider » Tue Sep 20, 2022 2:56 pm

dataq1 wrote:
Tue Sep 20, 2022 12:11 pm
ejbpesca wrote:
Tue Sep 20, 2022 7:34 am
Set to display six months, my Resmed Airsense 10 shows on its sleep report the following:

AHI 1.1
Total AI 0.8
Central AI 0.3

A single day report may show
AHI 0.3,
Total AI 0.3,
Central AI 0.3
One might expect that to imply a 0.3 Hypopnea index over that period and a 0.5 Obstructive Index.
"one" (ie, you) would be wrong.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Miss Emerita
Posts: 3444
Joined: Sun Nov 04, 2018 8:07 pm

Re: AHI Total AI Centra AI

Post by Miss Emerita » Tue Sep 20, 2022 4:13 pm

zonker wrote:
Tue Sep 20, 2022 10:49 am
Miss Emerita wrote:
Tue Sep 20, 2022 10:21 am
As you can see from the resource Pugsy linked to, AHI is apnea-hypopnea index, AI is apnea index, and Central AI is the index for central apneas only. "Index" is the number of events in the specified category divided by hours of sleep.

What about the obstructive apnea index? You can figure it out by subtracting the Central AI from the AI.

And what about the hypopnea index? You can figure it out by subtracting the AI from the AHI.
m'am?

i was told there'd be no math on the test.
Mike: Zonker, what are you doing here?

Mike: You forgot to go to your math retake exam. Now you're going to fail for sure.

Zonker: ...

Zonker: Bummer.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: AHI Total AI Centra AI

Post by zonker » Tue Sep 20, 2022 4:53 pm

Miss Emerita wrote:
Tue Sep 20, 2022 4:13 pm
zonker wrote:
Tue Sep 20, 2022 10:49 am


m'am?

i was told there'd be no math on the test.
Mike: Zonker, what are you doing here?

Mike: You forgot to go to your math retake exam. Now you're going to fail for sure.

Zonker: ...

Zonker: Bummer.
oh my god!! you win the internet for not just today, but for the entire week!

Image

Image

Image
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
ejbpesca
Posts: 214
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Re: AHI Total AI Centra AI

Post by ejbpesca » Tue Sep 20, 2022 5:11 pm

Uh, well. Math is not working, nor do I understand yet.

I think I realize I do not have from MyAir an indication at all of how many apnea episodes I had the night before, or on average over a week, month etc. by indexes.

How am I to interpret these indexes? What are they supposed to be telling me?

Is there a way to convert Total AI to the number of apneas I had during a period of sleep?

Thanks.

_________________
MachineMask

dataq1
Posts: 820
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: AHI Total AI Centra AI

Post by dataq1 » Tue Sep 20, 2022 5:34 pm

palerider wrote:
Tue Sep 20, 2022 2:56 pm
"one" (ie, you) would be wrong.
I was sincerely hoping that you PR had elected to not see anything I posted….. sigh

Anyway, just pronouncing your opinion that I am incorrect without offering an explanation (specifically how am I incorrect) doesn’t help anyone.
Please explain.
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
Julie
Posts: 19910
Joined: Tue Feb 28, 2006 12:58 pm

Re: AHI Total AI Centra AI

Post by Julie » Tue Sep 20, 2022 6:09 pm

Uh... to get back to the OP - Myair leaves a LOT to be desired and you'd be way better off by getting Oscar (see Stickies top of front pg) which is so comprehensive, so concise and people here can help you interpret, and if necessary (tho' it doesn't seem likely now) help with tweaking whatever might need it. You could always post a couple of nights here for all to see and talk about.

User avatar
Respirator99
Posts: 343
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: AHI Total AI Centra AI

Post by Respirator99 » Tue Sep 20, 2022 6:37 pm

dataq1 wrote:
Tue Sep 20, 2022 12:11 pm
ejbpesca wrote:
Tue Sep 20, 2022 7:34 am
Set to display six months, my Resmed Airsense 10 shows on its sleep report the following:

AHI 1.1
Total AI 0.8
Central AI 0.3

A single day report may show
AHI 0.3,
Total AI 0.3,
Central AI 0.3
What may be confusing (from the math standpoint) is the difference (in the six month report) between "AHI" and "Total AI" (1.1 and 0.8 ). One might expect that to imply a 0.3 Hypopnea index over that period and a 0.5 Obstructive Index.

One the day report, one might expect that to imply Zero Hypopneas/hour and Zero obstructions/hour; the only events that night were centrals- that the cpap cannot treat.

Correct?
CORRECT
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

User avatar
Respirator99
Posts: 343
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: AHI Total AI Centra AI

Post by Respirator99 » Tue Sep 20, 2022 7:59 pm

ejbpesca wrote:
Tue Sep 20, 2022 5:11 pm
Uh, well. Math is not working, nor do I understand yet.

I think I realize I do not have from MyAir an indication at all of how many apnea episodes I had the night before, or on average over a week, month etc. by indexes.

How am I to interpret these indexes? What are they supposed to be telling me?

Is there a way to convert Total AI to the number of apneas I had during a period of sleep?

Thanks.
The indexes are an indication of how many events you had per hour. We don't usually worry about the raw number of events as that will change depending on how long you sleep. It's the index which tells us how good or bad your apnea is. As a very rough rule, an AHI under 5.0 is regarded as "treated". Personally I need to get down below 2 to have a decent rest.

The index is simply calculated by dividing the number of events by hours slept. So if you slept for 8 hours (according to the machine) and recorded 20 events then your AHI would be 20 / 8 = 2.5.

If you really want to know the number of events, multiply the index by the hours you slept. eg if your total AI was 2.5 and you slept for 8 hours, then you had 2.5 x 8.0 = 20 events. In your case the number will be quite low as your apnea is well controlled (based on the numbers you've given us).

<edit to add>
Looking at your six month numbers, you had:
AHI 1.1
Total AI 0.8
Central AI 0.3
So your AHI (apnea-hypopnea index) is the sum of all apneas and hypopneas divided by total number of hours = 1.1
Your total AI is the sum of all apneas (obstructive, central and unclassified) divided by the hours = 0.8.
The central AI is the number of central apneas divided by the hours = 0.3.

From this it's easy to see that the obstructive obstructive apnea index is Total AI - Central AI = 0.8 - 0.3 = 0.5
The hypopnea index is the AHI - Total AI = 1.1 - 0.8 = 0.3.

For what it's worth, these are all very good numbers indicating your apnea is well under control.
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

User avatar
ejbpesca
Posts: 214
Joined: Tue Dec 10, 2013 12:09 pm
Location: Alabama

Re: AHI Total AI Centra AI

Post by ejbpesca » Wed Sep 21, 2022 8:25 am

Thank you Respirator99 for the detailed explanation.

Now on to why MyAir reports fail to keep accurate usage times.

_________________
MachineMask