New user -- high, constant pressure /loud noise
New user -- high, constant pressure /loud noise
Hi folks,
I'm new to CPAP, and my first machine is a ResMed AirMini. I purchased this while waiting for my primary machine to arrive (still several months from now). I've had this for six days so far. My first night was fine, but then...
At least once each night, I'm awakened (as is my spouse!) by a very loud noise accompanied by constant, very high pressure. Unlike normal operation, where the pressure modulates as I breathe in and out, this pressure is constant, with a large air flow out my mask vent. The mask doesn't leak. I've been very careful to insure a good fit, and this behavior is quite different from what happens if I temporarily displace the mask while adjusting my sleep position, etc.
Is this normal? Is my machine malfunctioning? Right now, I don't see how I can continue using the device with this happening every night. In each case I've had to turn it off and remove the mask. Sometimes I'll wake up again and put it back one, but not always. So this seems bad.
I'd appreciate any info from folks who know what's going on. Thanks!
I'm new to CPAP, and my first machine is a ResMed AirMini. I purchased this while waiting for my primary machine to arrive (still several months from now). I've had this for six days so far. My first night was fine, but then...
At least once each night, I'm awakened (as is my spouse!) by a very loud noise accompanied by constant, very high pressure. Unlike normal operation, where the pressure modulates as I breathe in and out, this pressure is constant, with a large air flow out my mask vent. The mask doesn't leak. I've been very careful to insure a good fit, and this behavior is quite different from what happens if I temporarily displace the mask while adjusting my sleep position, etc.
Is this normal? Is my machine malfunctioning? Right now, I don't see how I can continue using the device with this happening every night. In each case I've had to turn it off and remove the mask. Sometimes I'll wake up again and put it back one, but not always. So this seems bad.
I'd appreciate any info from folks who know what's going on. Thanks!
- Deborah K.
- Posts: 433
- Joined: Mon Feb 24, 2020 11:15 pm
- Location: Colorado
Re: New user -- high, constant pressure /loud noise
Bumping this up for you. Maybe someone who can help you will see your post and respond.
Machine: Resmed AirSense 10 Autoset For Her
Mask: Bleep Dreamport mask system
Mask: Bleep Dreamport mask system
Re: New user -- high, constant pressure /loud noise
Call whoever you got the machine from and ask them....
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
- ChicagoGranny
- Posts: 14469
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: New user -- high, constant pressure /loud noise
> When this happens, do you turn the machine off and them back on?
Sometimes yes. Sometimes I just turn if off and give up for the night.
> What are your pressure settings?
Originally 8 to 14. After describing the problem with my physician's assistant, we are trying 4 to 10. So far (one night), the event happened again but with less pressure. And the app reported a maximum pressure of 10. I'm still unclear on whether this is normal, but perhaps it is? Super annoying though.
And yes, of course I called the vendor. They replied saying that loud noises are usually caused by mask leaks. I'm sure that's true, but that's not what I'm experiencing.
Sometimes yes. Sometimes I just turn if off and give up for the night.
> What are your pressure settings?
Originally 8 to 14. After describing the problem with my physician's assistant, we are trying 4 to 10. So far (one night), the event happened again but with less pressure. And the app reported a maximum pressure of 10. I'm still unclear on whether this is normal, but perhaps it is? Super annoying though.
And yes, of course I called the vendor. They replied saying that loud noises are usually caused by mask leaks. I'm sure that's true, but that's not what I'm experiencing.
Re: New user -- high, constant pressure /loud noise
Take your machine back - a 'very loud noise' is not normal and setting your min. pressure (the one that does the job the machine is supposed to do) at 4, makes no sense.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
- ChicagoGranny
- Posts: 14469
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: New user -- high, constant pressure /loud noise
If your airway is on the verge of collapse at pressures below 10, it's perfectly normal for the machine to raise the pressure to 10. That's what we want it to do - raise the pressure to keep the airway open.
Is it possible you have the vents pointed toward a sheet or pillow which is magnifying the normal sound of the flow from the vents?
A pressure of 10 is in the middle of a typical pressure range. It would not be considered a high pressure.
Re: New user -- high, constant pressure /loud noise
I have only used my AirMini on a handful of trips and it has always been quite loud, not only from the device but the HumidX area of the mask!
Anyways, I took a video of the noises coming from my AirMini after being awoken by it. @OP are these the noises you are hearing too? I hear alternating loud whining jet engine noise and breathing noise in the video. Link to video: https://www.youtube.com/watch?v=cYCvZgKzzm0
Anyways, I took a video of the noises coming from my AirMini after being awoken by it. @OP are these the noises you are hearing too? I hear alternating loud whining jet engine noise and breathing noise in the video. Link to video: https://www.youtube.com/watch?v=cYCvZgKzzm0
Re: New user -- high, constant pressure /loud noise
I agree with the notion of taking the machine back to the vendor. Phone call won't work because it's easier for vendor to brush you off when it's just a phone connection. But if you are physically there and say "listen to this, it can't be normal" that is probably a better approach. They are not likely to physically throw you out the door. If they ask you to have a seat and wait, have a seat and START the machine. This will bug the receptionist and be very off-putting for prospective customers. I'm betting you won't have to wait long.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask
Resmed P10 pillow mask