The OSCAR graphs from last Friday: pressure restarted at 7.8. Oddly the panel on the left which says "Ramp Pressure 8.0" is wrong.


| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
No, I haven't touched it since I set it up last week.
I set it to 7.8 as suggested by Dog Slobber earlier in the thread. I did it on my old machine before I sent it back, then did it on the new machine when setting it up.What it shows in the graphs just doesn't go with the settings unless you change something in the settings at some point.
Autoramp engaged...but set to 8.0.
Where is the 7.8 coming from?

| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
Dog Slobber wrote: ↑Sat Jul 23, 2022 11:52 amConsider:
- Setting your ramp to Auto and Ramp Pressure to 7.8 (or ramp 8.0 and min pressure to 8.2). This is *not* for therapy reason but to determine if your machine is having power cycles. You won't notice the the 0.2 difference in pressure, but a pressure of 7.8 will show up on your graphs and the only explanation is a power cycle. AutoRamp does a good job of detecting sleep and switching to therapy mode, and if it doesn't detect sleep, it will also switch to therapy mode should it detect events.
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
| Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
Mask events...starting and stopping of the machine. Breaks in the therapy pressure and recording line.tisket wrote: ↑Fri Sep 02, 2022 4:08 pm"Your repaired machine has been received from the manufacturer. The manufacturer replaced the blower motor and found in the usage data that high leaks and repeated mask events were detected by the machine. The auto off feature may be causing the machine to shut down during therapy."
I don't recall seeing these events in the graphs I posted in this thread, nor do I understand why they would replace the blower motor if the power was going out or if the "auto off feature" was causing it. I think this is basically a "shrug" response.
What is the auto off feature and is there a way to disable it?
How do you see "mask events" in OSCAR?
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
| Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
| Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |
| Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
| Mask: DreamWear Nasal CPAP Mask with Headgear |