High MyAir Score but O2 Desats Continue
Re: High MyAir Score but O2 Desats Continue
Either post some 1 to 2 minute images of the bad area in question or upload that entire night to Dropbox or Drive for better review. The Night Summary (aka SkyLab View) is not helpful for the kind of analysis you're looking for.
It enlarges for me:
During the period in question your HR increases (to wake levels). Is all that SWJ?
And you've got hundreds of PCs. What's up with that?
It enlarges for me:
During the period in question your HR increases (to wake levels). Is all that SWJ?
And you've got hundreds of PCs. What's up with that?
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High MyAir Score but O2 Desats Continue
OK so if my adds and subtracts don't fail me 28,000 beats/night, 1450 PCs is ~5%.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
- latskogkatt
- Posts: 64
- Joined: Tue Jul 19, 2022 4:30 pm
- Location: Pacific Northwest
Re: High MyAir Score but O2 Desats Continue
There are a couple acronyms you used here that aren't in the glossary or wiki, and since they're less than 4 characters I can't search them. If you don't mind, I'd like to know what SWJ and PCs refer to.During the period in question your HR increases (to wake levels). Is all that SWJ?
And you've got hundreds of PCs. What's up with that?
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Additional Comments: I switch between 4 different masks, as needed Settings: 11-20, no ramp, epr 3 |
Re: High MyAir Score but O2 Desats Continue
PC Pulse Change. A change in heart rate of 6 beats per minute but user-adjustable.
SWJ Sleep-Wake Junk. A term I invented denoting the interface between sleep and wake when breathing can be unstable. If someone is having difficulty taking sleep, waveforms during this period can be one big mess.
SWJ Sleep-Wake Junk. A term I invented denoting the interface between sleep and wake when breathing can be unstable. If someone is having difficulty taking sleep, waveforms during this period can be one big mess.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High MyAir Score but O2 Desats Continue
So glad you asked!latskogkatt wrote: ↑Tue Jul 26, 2022 3:56 amThere are a couple acronyms you used here that aren't in the glossary or wiki, and since they're less than 4 characters I can't search them. If you don't mind, I'd like to know what SWJ and PCs refer to.During the period in question your HR increases (to wake levels). Is all that SWJ?
And you've got hundreds of PCs. What's up with that?
And so glad you replied!Rubicon wrote: ↑Tue Jul 26, 2022 4:21 amPC Pulse Change. A change in heart rate of 6 beats per minute but user-adjustable.
SWJ Sleep-Wake Junk. A term I invented denoting the interface between sleep and wake when breathing can be unstable. If someone is having difficulty taking sleep, waveforms during this period can be one big mess.
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Re: High MyAir Score but O2 Desats Continue
so glad you repeated it!ozij wrote: ↑Wed Jul 27, 2022 5:12 amSo glad you asked!latskogkatt wrote: ↑Tue Jul 26, 2022 3:56 amThere are a couple acronyms you used here that aren't in the glossary or wiki, and since they're less than 4 characters I can't search them. If you don't mind, I'd like to know what SWJ and PCs refer to.During the period in question your HR increases (to wake levels). Is all that SWJ?
And you've got hundreds of PCs. What's up with that?
And so glad you replied!Rubicon wrote: ↑Tue Jul 26, 2022 4:21 amPC Pulse Change. A change in heart rate of 6 beats per minute but user-adjustable.
SWJ Sleep-Wake Junk. A term I invented denoting the interface between sleep and wake when breathing can be unstable. If someone is having difficulty taking sleep, waveforms during this period can be one big mess.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 6
- Joined: Mon Jul 18, 2022 3:20 pm
- Location: Massachusetts
Re: High MyAir Score but O2 Desats Continue
Yes, what's up with that indeed. I have been told by ECG technicians that I have respiratory sinus arrythmia. https://www.healthline.com/health/sinus-arrhythmia When I breath in my heart rate increases, when I exhale it decreases. This is what I have been told. I see heart rate spikes that seem to line up to low SPO2 events. Part of what I am trying to solve actually is why my heart rates seems to change so much versus others whom I have seen.Rubicon wrote: ↑Tue Jul 26, 2022 3:05 amEither post some 1 to 2 minute images of the bad area in question or upload that entire night to Dropbox or Drive for better review. The Night Summary (aka SkyLab View) is not helpful for the kind of analysis you're looking for.
During the period in question your HR increases (to wake levels). Is all that SWJ?
And you've got hundreds of PCs. What's up with that?
Here is a particularly bad moment. You can see it is OA, then breathing, OA then breathing, etc, etc. and oddly no leaks. It isn't calling it a clear airway event - calling it obstructive. O2 for sure dropping...
I am going to buy a better mask but since I see no leaks just feels like my air hole is plugging up and saying sorry, not sorry to my heart and brain. I am also going to try a cervical collar.
I'd appreciate any insight.
Re: High MyAir Score but O2 Desats Continue
That screen looks more like HF HRV to me.
Those PCs could be LF HRV, HF HRV, SA, the PVCs or other. Really need an EKG channel to precisely explain the 1450 PCs.
See viewtopic/t57907/To-gvz.html for some explanation.
Those PCs could be LF HRV, HF HRV, SA, the PVCs or other. Really need an EKG channel to precisely explain the 1450 PCs.
See viewtopic/t57907/To-gvz.html for some explanation.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
-
- Posts: 6
- Joined: Mon Jul 18, 2022 3:20 pm
- Location: Massachusetts
Re: High MyAir Score but O2 Desats Continue
I have a personal ECG, but I don't think I can add that channel to Oscar, can I? I have had numerous heart work ups. ECG, Stress Test, Echo, etc. My diagnosis is Benign PVCs. I can't speak to the Heart Rate changes except that I have also used ReadMyEcg (an app that you can submit your ECGs to be read by a tech) and they come back frequently with Respiratory Sinus Arrythmia.Rubicon wrote: ↑Sun Aug 07, 2022 3:03 amThat screen looks more like HF HRV to me.
Those PCs could be LF HRV, HF HRV, SA, the PVCs or other. Really need an EKG channel to precisely explain the 1450 PCs.
See viewtopic/t57907/To-gvz.html for some explanation.
Does LF = Low Frequency, HF = High Frequency, HRV = Heart Rate Variation, SA = Sinus Arrythmia?
Re: High MyAir Score but O2 Desats Continue
The August 4th snippet you posted isn't SA, it's LF HRV.ArthurDent1978 wrote: ↑Sun Aug 07, 2022 11:26 amReadMyEcg (an app that you can submit your ECGs to be read by a tech) and they come back frequently with Respiratory Sinus Arrythmia.
The PCs that night don't start until 2.5 hours into the night. Seems to me SA would be consistent regardless of the circumstances.
IDK about the qualifications of the ReadMyECG Techs. They could be LSsOS.
What are the desat and PC parameters set for?
IIWM I'd magnify the Pulse Rate channel, go to 60 second view and try to match up the PCs with the corresponding respiratory event.
Determine the duration of the PC events. SA will be up to 2.5 seconds.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High MyAir Score but O2 Desats Continue
Specifically:
HF 0.15 - 0.40 hz or 6.7 seconds to 2.5 seconds
LF 0.04 - 0.15 hz or 25 seconds to 6.7 seconds
HF 0.15 - 0.40 hz or 6.7 seconds to 2.5 seconds
LF 0.04 - 0.15 hz or 25 seconds to 6.7 seconds
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High MyAir Score but O2 Desats Continue
Or down to, depending on your POV.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
Re: High MyAir Score but O2 Desats Continue
This reminds me of the good ol' days:
You know what this says in 12 words or less?
In central apnea and CompSA the apneas are all the same length.
You know what this says in 12 words or less?
In central apnea and CompSA the apneas are all the same length.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.
-
- Posts: 6
- Joined: Mon Jul 18, 2022 3:20 pm
- Location: Massachusetts
Re: High MyAir Score but O2 Desats Continue
Help me understand better. What is CompSA? I'm confused - are you saying that a sinus arrythmia can cause apneas? Sorry I'm just not following your logic.
Re: High MyAir Score but O2 Desats Continue
Did you look at my link above? You need to deep dive into those PCs to see the exact timing. If the entire change occurs within a single breath, slow during expiration, faster during inspiration, then it's a sinus arrhythmia. If it occurs during a respiratory event, slow during apnea, faster during resumption of breathing, it's LF HRV. If there's arrhythmia, then that's the cause.
In your screenshot it looks like LF HRV there. Put the pulse value parameters at like 50-80 and it should stand out a little more.
Then look for respiratory variation throughout the night to see if there's a pattern.
In your screenshot it looks like LF HRV there. Put the pulse value parameters at like 50-80 and it should stand out a little more.
Then look for respiratory variation throughout the night to see if there's a pattern.
Freeze this moment a little bit longer.
Make each sensation a little bit stronger.
Experience slips away.
Make each sensation a little bit stronger.
Experience slips away.