Can anyone recommend CPAP Pillow?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Can anyone recommend CPAP Pillow?

Post by jumblegirl » Tue Jun 21, 2022 8:43 am

Can anyone recommend a pillow that is less likely to make the mask fall off or slide around and leak. I keep getting one or 2 hours missing data where I don't know what happened but assume mask came off. Anything that is comfortable but will help keep the mask on. I use a soft down pillow now, for stomach sleepers which I can't really do anymore but it works ok.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
Pugsy
Posts: 64016
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can anyone recommend CPAP Pillow?

Post by Pugsy » Tue Jun 21, 2022 8:55 am

If you are using SmartStop enabled or turned on....turn it off.
That way if the mask slips off your placement on the face....the machine will keep blowing air and you can see it on the reports.
This will help you figure out if the missing 2 hours or whatever you see on the reports is related to mask coming off or if it points to something else.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Can anyone recommend CPAP Pillow?

Post by jumblegirl » Tue Jun 21, 2022 10:48 am

so what is weird is that when I looked at the screen on the machine at like 3 am it had 3 blue bouncing dots on it, along with the pressure level which was low like 3-4 - I have it set now to Autoset for her just to try it out so I don't know if it's from that but I thought this was during the time when there's no data on the graph.

unless I hallucinated it haha but I'll take a photo tonight :)

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
Pugsy
Posts: 64016
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Can anyone recommend CPAP Pillow?

Post by Pugsy » Tue Jun 21, 2022 10:57 am

I don't remember seeing a report where there was a large block of empty space on the report graphs from you.
Did I just miss it?
Do you ever take a nap with the machine on during the day?

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Can anyone recommend CPAP Pillow?

Post by jumblegirl » Tue Jun 21, 2022 11:35 am

I didn't post it yet because I am so bad at Imgur importing I"ll do it now:

https://imgur.com/a/00UGQRg

Hope that works

Theres also no red pressure line just the green EPAP I don't know if that's to do with my mask or the setting of Autoset for her
(I will go ahead an post it in my original thread to keep it in one place, sorry)

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
ozij
Posts: 10141
Joined: Fri Mar 18, 2005 11:52 pm

Re: Can anyone recommend CPAP Pillow?

Post by ozij » Tue Jun 21, 2022 12:02 pm

jumblegirl wrote:
Tue Jun 21, 2022 11:35 am
I didn't post it yet because I am so bad at Imgur importing I"ll do it now:

https://imgur.com/a/00UGQRg

Hope that works

Theres also no red pressure line just the green EPAP I don't know if that's to do with my mask or the setting of Autoset for her
(I will go ahead an post it in my original thread to keep it in one place, sorry)
You've got EPR off - hence only one pressure line. Same pressure for inhale and exhale.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
LoneStar
Posts: 187
Joined: Wed Mar 11, 2009 1:35 pm
Location: south central Texas

Re: Can anyone recommend CPAP Pillow?

Post by LoneStar » Tue Jun 21, 2022 12:41 pm

As far as pillows go, I use a down pillow from LL Bean. I can squish it around my mask easily and shape it just about any way I need. I use the Nuance nasal mask for reference.

There are pillows with c-shaped cut-outs on the side for larger masks - I think Amazon has them and probably CPAP.com does, too. Not everything works for everyone, you may just have to try a few to find what you like.

_________________
Machine: Airsense 10 Card to Cloud
Mask: Nuance & Nuance Pro Nasal Pillow CPAP Mask with Gel Nasal Pillows
Humidifier: HC150 Heated Humidifier With Hose, 2 Chambers and Stand
Additional Comments: APAP, humidifier on passover, heated hose
With xPAP therapy, do or do not; there is no try.

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Can anyone recommend CPAP Pillow?

Post by zonker » Tue Jun 21, 2022 6:01 pm

um, guys? may i humbly suggest most of this conversation belongs in jumble's therapy thread? she started this for pillow advice.

it can get kind of confusing for those of us following along at home.

thank you!
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34390
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Can anyone recommend CPAP Pillow?

Post by chunkyfrog » Tue Jun 21, 2022 7:02 pm

Down-alternative, medium to low height for me.
I found mine on clearance at B3--likely the last of that model on earth.
Down is too spendy to try, and I slept (not well) on feathers as a child.
I suspect possible allergies . . .
My good pillow has a nice satin case--the cat loves it.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can anyone recommend CPAP Pillow?

Post by palerider » Tue Jun 21, 2022 8:25 pm

jumblegirl wrote:
Tue Jun 21, 2022 8:43 am
Can anyone recommend a pillow that is less likely to make the mask fall off or slide around and leak. I keep getting one or 2 hours missing data where I don't know what happened but assume mask came off. Anything that is comfortable but will help keep the mask on. I use a soft down pillow now, for stomach sleepers which I can't really do anymore but it works ok.
I've spent the last 13 years just sleeping towards the edge of a regular pillow, no muss, no fuss, no extra expense.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can anyone recommend CPAP Pillow?

Post by palerider » Tue Jun 21, 2022 8:29 pm

jumblegirl wrote:
Tue Jun 21, 2022 8:43 am
Can anyone recommend a pillow that is less likely to make the mask fall off or slide around and leak. I keep getting one or 2 hours missing data
Also, do you sleep past noon?

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Can anyone recommend CPAP Pillow?

Post by zonker » Tue Jun 21, 2022 9:34 pm

palerider wrote:
Tue Jun 21, 2022 8:25 pm
jumblegirl wrote:
Tue Jun 21, 2022 8:43 am
Can anyone recommend a pillow that is less likely to make the mask fall off or slide around and leak. I keep getting one or 2 hours missing data where I don't know what happened but assume mask came off. Anything that is comfortable but will help keep the mask on. I use a soft down pillow now, for stomach sleepers which I can't really do anymore but it works ok.
I've spent the last 13 years just sleeping towards the edge of a regular pillow, no muss, no fuss, no extra expense.
me too. and come to think of it, it may have been you that suggested it to me.

this works very well with my p10 nasal pillows. i have no idea how it would work with other types of masks, though.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Can anyone recommend CPAP Pillow?

Post by jumblegirl » Wed Jun 22, 2022 10:39 am

palerider wrote:
Tue Jun 21, 2022 8:29 pm
jumblegirl wrote:
Tue Jun 21, 2022 8:43 am
Can anyone recommend a pillow that is less likely to make the mask fall off or slide around and leak. I keep getting one or 2 hours missing data
Also, do you sleep past noon?
No I never sleep past noon - I guess that would affect Oscar data then? Lucky if I sleep past 5am.

Sorry zonked I will try to keep it in the therapy thread :)

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
ozij
Posts: 10141
Joined: Fri Mar 18, 2005 11:52 pm

Re: Can anyone recommend CPAP Pillow?

Post by ozij » Wed Jun 22, 2022 11:33 am

palerider wrote:
Tue Jun 21, 2022 8:25 pm
jumblegirl wrote:
Tue Jun 21, 2022 8:43 am
Can anyone recommend a pillow that is less likely to make the mask fall off or slide around and leak. I keep getting one or 2 hours missing data where I don't know what happened but assume mask came off. Anything that is comfortable but will help keep the mask on. I use a soft down pillow now, for stomach sleepers which I can't really do anymore but it works ok.
I've spent the last 13 years just sleeping towards the edge of a regular pillow, no muss, no fuss, no extra expense.
The years I did that successfully depended on the type of mask I had. And when unsuccessful, it resulted in cramped neck and shoulders.
I ordered a CPAP pillow from Amazon - based my choice on some of the opinions there. e.g. - avoided pillows reported as having a smell.... and tried to imagine what would fit my mask and favorite sleep position.

What works for one of us may not for another, and vice versa....

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Can anyone recommend CPAP Pillow?

Post by palerider » Wed Jun 22, 2022 5:33 pm

jumblegirl wrote:
Wed Jun 22, 2022 10:39 am
palerider wrote:
Tue Jun 21, 2022 8:29 pm
jumblegirl wrote:
Tue Jun 21, 2022 8:43 am
Can anyone recommend a pillow that is less likely to make the mask fall off or slide around and leak. I keep getting one or 2 hours missing data
Also, do you sleep past noon?
No I never sleep past noon - I guess that would affect Oscar data then? Lucky if I sleep past 5am.

Sorry zonked I will try to keep it in the therapy thread :)
Resmed machines roll their data over at 'noon' (machine time).

Look at Oscar, if you've got data missing in the middle of the night, then ... see if you can figure out when, if there's a pattern.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.