Oscar chart

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
ChicagoGranny
Posts: 14470
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Oscar chart

Post by ChicagoGranny » Sat Jun 18, 2022 10:10 am

jumblegirl wrote:
Sat Jun 18, 2022 7:28 am
what is the y axis of the flow rate ?
It is liters per minute. Positive values for inhalation, negative for exhalation.

jumblegirl wrote:
Sat Jun 18, 2022 7:28 am
why is it not 0-100
If you run or lift weights in your sleep, you could hit 100.
jumblegirl wrote:
Sat Jun 18, 2022 7:28 am
There is a 43 sec OA that's the worst one.
Can't comment without zooming in on this section of the graph.

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Oscar chart

Post by jumblegirl » Sat Jun 18, 2022 11:47 am

Heres the zoom in of the OA

https://imgur.com/ivsRk2F

I was thinking Flow Rate would be in percentage, thanks.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Oscar chart

Post by jumblegirl » Sat Jun 18, 2022 2:13 pm

so if apnea is 80-100% reduction in flow, it's whatever brings it to 0 flow? How does the % reduction work. Sorry if stupid question.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
ozij
Posts: 10142
Joined: Fri Mar 18, 2005 11:52 pm

Re: Oscar chart

Post by ozij » Sat Jun 18, 2022 9:14 pm

jumblegirl wrote:
Sat Jun 18, 2022 2:13 pm
so if apnea is 80-100% reduction in flow, it's whatever brings it to 0 flow? How does the % reduction work. Sorry if stupid question.
The machines measure the flow of your breath, computing a running average, and a reduction in the flow will be considered an apnea, or a hypopnea. Any reduction in flow below 80% of that running average is an apnea.
The pressurized air is supposed to help keep you air flow as close to 100% as possible.

The therapy's aim is keep your airway open so that your sleep breathing flows stably, at a constant volume without too many reductions.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks.
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery

Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Oscar chart

Post by jumblegirl » Mon Jun 20, 2022 11:51 am

Here is today's Oscar graphs,a few more centrals mostly toward the end, I have chart zoomed I don't know if these are from waking up or what the breathing before and after means. Seems like a lot more leaks than usual.

https://imgur.com/a/c1JhpGG

and zoomed Centrals

https://imgur.com/a/aHQStd6

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
Pugsy
Posts: 64023
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Oscar chart

Post by Pugsy » Mon Jun 20, 2022 12:09 pm

You didn't zoom in enough. You have to be able to see each breath to evaluate the shape and pattern.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Oscar chart

Post by jumblegirl » Mon Jun 20, 2022 2:41 pm

I think i must have posted the wrong one I mean to zoom more

https://imgur.com/a/aHQStd6

https://imgur.com/a/JrIdiuB

Is that zoomed enough ? There are a bunch in a row but I don't know which part of the breathing is most helpful to see.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Oscar chart

Post by jumblegirl » Mon Jun 20, 2022 2:52 pm

This is zoomed closer 22 sec CA, and there was another CA right before it and another maybe 3 min later

https://imgur.com/a/YgNfsEf

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
Pugsy
Posts: 64023
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Oscar chart

Post by Pugsy » Mon Jun 20, 2022 2:55 pm

jumblegirl wrote:
Mon Jun 20, 2022 2:52 pm
This is zoomed closer 22 sec CA, and there was another CA right before it and another maybe 3 min later

https://imgur.com/a/YgNfsEf
this one you didn't include enough of the flow rate immediately prior (1 minute at least) to adequately evaluate it.
The breathing after the flag is obviously awake/arousal breathing but we can't see enough of the breathing prior to the flag to make a determination.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Pugsy
Posts: 64023
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Oscar chart

Post by Pugsy » Mon Jun 20, 2022 3:02 pm

jumblegirl wrote:
Mon Jun 20, 2022 2:41 pm
I don't know which part of the breathing is most helpful to see.
The breathing in front of the flagged event...at least a minute or so.
Often the flagged event has to be to the far right of the graph.

I know we like to cram as many flagged events as we can into one image but when we do that then it's hard to know what all started the cluster of events. I have seen one real OA (on my own reports) trigger an arousal and then several awake/arousal related event flags. It's rare that I have that happen now but occasionally I still see it happen. This is why it is important to deal with OAs....because they cause arousals. The arousal related flags really aren't important in terms of AHI because they don't count but in terms of being related to crappy sleep because of the one real asleep OA...now that crappy sleep is important. When we see a lot of arousal breathing then we know there's a lot of crappy sleep....with or without flagging of any events.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Oscar chart

Post by jumblegirl » Mon Jun 20, 2022 3:04 pm

what about the 2 links 2 message above, they aren't zoomed as much but have more time before the apnea

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
Pugsy
Posts: 64023
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Oscar chart

Post by Pugsy » Mon Jun 20, 2022 3:15 pm

jumblegirl wrote:
Mon Jun 20, 2022 3:04 pm
what about the 2 links 2 message above, they aren't zoomed as much but have more time before the apnea
The bottom link....iffy as to it being asleep related...can't always tell easily. There's a big gulp of air (we don't take big gulps of air when we are asleep) and then something that looks sort of like asleep breathing that is very brief. Is it maybe a sleep onset central???? Maybe. Then after it the flag the flow rate isn't easy to evaluate. IMHO

The top link....I don't think sound asleep for the first 2 but maybe asleep for the last one. Again maybe sleep onset?
I don't know for sure. I have this sort of thing myself on occasion and I can't always tell for sure.

Again...I think you are likely having a mix of arousal related central apnea flagging and some asleep central flagging.

Since there's really not much we can do about centrals except maybe play with EPR (which only helps IF EPR is a factor in the centrals) I still say work on what we know we can fix with the machine tweaking and that's OAs and hyponeas.
And also give yourself some time to just get adjusted to and sleeping with the mask and machine....and THEN see what is left that might be a problem.
If an OA causes the arousal which then causes some arousal/awake central flagging or again some sleep onset centrals...then what we need to do is deal with OA/hyponea causing the arousal.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Oscar chart

Post by jumblegirl » Mon Jun 20, 2022 3:18 pm

so I just posted 3 close ups with more time breathing before and I don't know where they went here is the last one

https://imgur.com/a/Xw3xjvp

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Oscar chart

Post by zonker » Mon Jun 20, 2022 3:25 pm

jumblegirl wrote:
Mon Jun 20, 2022 3:18 pm
so I just posted 3 close ups with more time breathing before and I don't know where they went here is the last one

https://imgur.com/a/Xw3xjvp
it might help you if, after you've composed your message, you use the preview button to make sure it "took". if it did, then hit submit. if not, try again!
:wink:
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

jumblegirl
Posts: 101
Joined: Sun Jun 12, 2022 7:34 am

Re: Oscar chart

Post by jumblegirl » Mon Jun 20, 2022 3:30 pm

Thanks, I feel like I replied in the middle of the thread like on a different page, but assumed it would land at the end of the thread anyway so I don't know where it went. I am really terrible with a laptop and so much more used to phone. But for Oscar I will use the laptop :)

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 For Her Full Face CPAP Mask with Headgear
Additional Comments: APAP min 7/max 12, EPR off