Noobie Needs Help
Re: Noobie Needs Help
You might benefit from the additional features of the VAuto. It can go to 25 cm if needed.
And it has the ability to give you higher relief during exhale than the AutoSet if you end up needing higher pressures you might like more relief than EPR can give you.
When people start seeing pressures in the teens being needed then the extra relief often comes in handy.
The AutoSet has EPR 1,2 or 3.
The VAuto has something similar in function called PS pressure support and it will do 1 to 5, or 6 or more.
And it has the ability to give you higher relief during exhale than the AutoSet if you end up needing higher pressures you might like more relief than EPR can give you.
When people start seeing pressures in the teens being needed then the extra relief often comes in handy.
The AutoSet has EPR 1,2 or 3.
The VAuto has something similar in function called PS pressure support and it will do 1 to 5, or 6 or more.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
-
- Posts: 157
- Joined: Tue May 17, 2022 2:56 am
- Location: Kansas City, Mo.
Re: Noobie Needs Help
Thanks for your unofficial, unmedical opinion.Pugsy wrote: ↑Sun May 29, 2022 8:46 pmYou might benefit from the additional features of the VAuto. It can go to 25 cm if needed.
And it has the ability to give you higher relief during exhale than the AutoSet if you end up needing higher pressures you might like more relief than EPR can give you.
When people start seeing pressures in the teens being needed then the extra relief often comes in handy.
The AutoSet has EPR 1,2 or 3.
The VAuto has something similar in function called PS pressure support and it will do 1 to 5, or 6 or more.
I will talk to the pulmonologist tomorrow and ask about getting the AC10 VPAP.
I am going to take a printout of the OSCAR screenshots with me, for them to look at.
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Re: Noobie Needs Help
if it were me, i'd take the sd card as well. not many doctors are aware of the oscar software.Iamstumped wrote: ↑Mon May 30, 2022 2:05 pm
I will talk to the pulmonologist tomorrow and ask about getting the AC10 VPAP.
I am going to take a printout of the OSCAR screenshots with me, for them to look at.
good luck with the vpap.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 157
- Joined: Tue May 17, 2022 2:56 am
- Location: Kansas City, Mo.
Re: Noobie Needs Help
I was using an N30i (SW) on a (Sm) frame.
Started wearing a chin strap, but I removed that in the middle of the night, as it became sore to continue to wear it.
I also took 2, 5mg tablets of Melatonin before going to bed.
Fell asleep fairly quickly, did not wake up during the night, that I know of.
Feel somewhat good today.
P.S. the pulmonologist took 8 days of data, and determined that my 30 day average of my 95% pressure was only 10.5 cmh2O
REALLY?
How can you get a 30 day average with only 8 days worth of data?
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Re: Noobie Needs Help
That last chart was a real keeper. Whatever you did, try and do it more often. I am not an advanced math whiz, but reasonably proficient with plain arithmetic. That said, I can't see how you can take 8 days of data and determine a 30 day average which is LESS than your 8 day average. The good news is that you can change your settings if doc sets to 10.5.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
ResMed AirSense 10 Autoset with built in humidifier
Resmed P10 pillow mask
Resmed P10 pillow mask
Re: Noobie Needs Help
ex·tra·po·la·tionIamstumped wrote: ↑Fri Jun 03, 2022 4:02 pm
How can you get a 30 day average with only 8 days worth of data?
/ikˌstrapəˈlāSH(ə)n/
Learn to pronounce
noun
the action of estimating or concluding something by assuming that existing trends will continue or a current method will remain applicable.
"sizes were estimated by extrapolation"
MATHEMATICS
the extension of a graph, curve, or range of values by inferring unknown values from trends in the known data.
"extrapolation of the logarithmic curve yielded an estimate of 66 species"
sorry for the poor formatting. it's the first thing that pops up when one googles "extrapolation".
https://www.google.com/search?q=extrapo ... e&ie=UTF-8
math is fun!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 157
- Joined: Tue May 17, 2022 2:56 am
- Location: Kansas City, Mo.
Re: Noobie Needs Help
Seems like every time that I use a FFM, my AHI numbers go up, and there is an UGLY cluster of UA's, in the beginning, while I am trying to fall asleep, but when I use a nasal mask, my AHI numbers do go down, and there is very few UA's, in the beginning while I am trying to fall asleep.
When I use a FFM, should I up my min pressure? Currently at 14cmh2O. I was looking at the manuals that came with each FFM, and it looks like the maker recommends increasing the pressure due to a pressure drop.
Or maybe I should decrease the EPR from 3 to 2, or even lower it from 3 to 1, or maybe even turn it off completely?
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
- Miss Emerita
- Posts: 3444
- Joined: Sun Nov 04, 2018 8:07 pm
Re: Noobie Needs Help
The large leak line somehow got nixed from your events graph. It’d be good to restore it. If you look on the left panel, you’ll see you were running large leaks almost the whole time. So check you long hose your short hose, and all your hose connections. If it wasn’t your mask leaking, it’s got to be something else.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
- Posts: 157
- Joined: Tue May 17, 2022 2:56 am
- Location: Kansas City, Mo.
Re: Noobie Needs Help
Actually, I unchecked it, under the events tab. I will restore it right away.Miss Emerita wrote: ↑Tue Jun 07, 2022 11:03 pmThe large leak line somehow got nixed from your events graph. It’d be good to restore it. If you look on the left panel, you’ll see you were running large leaks almost the whole time. So check you long hose your short hose, and all your hose connections. If it wasn’t your mask leaking, it’s got to be something else.
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
-
- Posts: 157
- Joined: Tue May 17, 2022 2:56 am
- Location: Kansas City, Mo.
Re: Noobie Needs Help
According to OSCAR, my 95% pressure is just barely above the min pressure setting (12 cmh2O) on my machine.
Question: what is that, that is just below the SD Card slot, on the left side, on the AS10
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Re: Noobie Needs Help
Why?Iamstumped wrote: ↑Mon Jun 13, 2022 10:52 amShould I adjust the min pressure settings on my machine?
According to OSCAR, your pressure goes abov1 12 only 5% of the time.
So what do you mean by "adjust"? Is it bothering you and you'd prefer it to be lower part of the time?
How do you feel?
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Machine: Resmed AirSense10 for Her with Climateline heated hose ; alternating masks. |
And now here is my secret, a very simple secret; it is only with the heart that one can see rightly, what is essential is invisible to the eye.
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
Antoine de Saint-Exupery
Good advice is compromised by missing data
Forum member Dog Slobber Nov. 2023
-
- Posts: 157
- Joined: Tue May 17, 2022 2:56 am
- Location: Kansas City, Mo.
Re: Noobie Needs Help
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Re: Noobie Needs Help
this is fine for now. next time, would like to see the "events flags" across the top. and we don't need the ahi chart.Iamstumped wrote: ↑Fri Jun 17, 2022 1:57 pmLet me know if this worked.
https://imgur.com/xhLUNEB
It looks like it did
how'd you sleep? looks like a decent enough chart all told.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 157
- Joined: Tue May 17, 2022 2:56 am
- Location: Kansas City, Mo.
Re: Noobie Needs Help
No left side bar, just the actual charts?zonker wrote: ↑Fri Jun 17, 2022 6:55 pmthis is fine for now. next time, would like to see the "events flags" across the top. and we don't need the ahi chart.Iamstumped wrote: ↑Fri Jun 17, 2022 1:57 pmLet me know if this worked.
https://imgur.com/xhLUNEB
It looks like it did
how'd you sleep? looks like a decent enough chart all told.
Fair, I guess. That was just a test, to see if I could link to imgur, like that instead of doing an attachment to here since it is getting full again.
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Re: Noobie Needs Help
Left side bar is always needed.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.