How to determine CDMA carrier for ResMed AirSense 10

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

How to determine CDMA carrier for ResMed AirSense 10

Post by JLROhio » Tue Jan 25, 2022 6:51 pm

Does any know of a way to determine who the cell phone carrier is for a ResMed AirSense 10?

When I go into the menu, I can only see the SW version and SN of the internal modem.
Under the Provider label, it is blank.

I know that I've seen AT&T listed as a carrier on another ResMed device.
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
chunkyfrog
Posts: 34385
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by chunkyfrog » Tue Jan 25, 2022 8:23 pm

What is this obsession people have about the cellular carrier?
At least some have gotten over the impression that wi-fi has anything to do about it.

If you are afraid of 5g for some weird idea, just turn on airplane mode.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
loggerhead12
Posts: 560
Joined: Fri Sep 01, 2017 4:00 pm

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by loggerhead12 » Tue Jan 25, 2022 8:31 pm

Look at the label on the back. One of the numbers will end with 4G or CDMA.

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by JLROhio » Tue Jan 25, 2022 9:38 pm

chunkyfrog wrote:
Tue Jan 25, 2022 8:23 pm
What is this obsession people have about the cellular carrier?
At least some have gotten over the impression that wi-fi has anything to do about it.

If you are afraid of 5g for some weird idea, just turn on airplane mode.
What obsession are your referring to? I asked a simple question. :oops:

Why are you mentioning Wi-Fi? I did not mention the word or ask about Wi-Fi at all. :roll:

And where the %(@% did I mention anything about 5G? :oops:

Let me type this slower so maybe someone will understand.

- How would one determine the CDMA carrier on their ResMed AirSense 10?

It's a very simple question. No need to go down rabbit holes with obsessions, wi-fi, or 5G (OMG :shock: )
BTW, I sleep very comfortable every night next to my CDMA radiating CPAP machine, along with my cell phone, and plenty of wi-fi devices in the house. 8)
---

----
loggerhead12 wrote:
Tue Jan 25, 2022 8:31 pm
Look at the label on the back. One of the numbers will end with 4G or CDMA.
Ok, let me try to type this even slooooower.

- How would one determine the CDMA carrier on their ResMed AirSense 10?

I already know that my machine is CDMA - thus the reason for my question.
I did not ask for the difference btwn CDMA & 4G, nor how to find out if my machine is CDMA or 4G.
Although, thanks for your answer, as someone else might find it useful when researching their machine, as I found out when I researched my question and I read a post up here where someone posted a pic of the back of a 4G machine and where to look.

:wink:
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by zonker » Tue Jan 25, 2022 9:50 pm

chunkyfrog wrote:
Tue Jan 25, 2022 8:23 pm


If you are afraid of 5g for some weird idea, just turn on airplane mode.
given the current state of affairs, i find it funny that 5g and airplane used in the same sentence.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by JLROhio » Tue Jan 25, 2022 10:26 pm

zonker wrote:
Tue Jan 25, 2022 9:50 pm
...i find it funny that 5g...
Image
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

GrumpyHere
Posts: 508
Joined: Sat Feb 13, 2021 2:40 am

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by GrumpyHere » Tue Jan 25, 2022 10:58 pm

It's not going to matter by the end of year.

CDMA sunsets Stateside in 2022.
WHEN WILL AT&T, VERIZON AND T-MOBILE PHASE OUT 3G?

AT&T announced that its 3G network will be shut off by February 2022.

T-Mobile will finish shutting down Sprint’s 3G network, also known as CDMA, by March 2022, and retire Sprint’s LTE network by June 30, 2022. T-Mobile also plans to shut down its 3G networks by July 1, 2022, and will retire its older GSM 2G network, but no date has been set.

Verizon said it would finish decommissioning its 3G network by Dec. 31, 2022.
ResMed Lumis Tx - VAuto mode

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by JLROhio » Tue Jan 25, 2022 11:15 pm

GrumpyHere wrote:
Tue Jan 25, 2022 10:58 pm
It's not going to matter by the end of year.

CDMA sunsets Stateside in 2022.
WHEN WILL AT&T, VERIZON AND T-MOBILE PHASE OUT 3G?

AT&T announced that its 3G network will be shut off by February 2022.

T-Mobile will finish shutting down Sprint’s 3G network, also known as CDMA, by March 2022, and retire Sprint’s LTE network by June 30, 2022. T-Mobile also plans to shut down its 3G networks by July 1, 2022, and will retire its older GSM 2G network, but no date has been set.

Verizon said it would finish decommissioning its 3G network by Dec. 31, 2022.
Thanks - That's a bit closer to the type of info that I'm seeking.

I am hoping that someone has an answer as to how one could determine what cell phone carrier their ResMed AirSense 10 CPAP machine is using, when it does not show the carrier name in the menu.

Thanks for your answer and not going down the 5G conspiracy rabbit hole!

Image
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
Jeff8356
Posts: 39
Joined: Mon Jan 04, 2021 11:12 am
Location: North Carolina

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by Jeff8356 » Wed Jan 26, 2022 1:23 am

If the provider field is blank then it may not be connected at all. Scroll down a bit more and see if there is anything in the signal strength box. If that shows --dbm then there is no connection at all. Is it possible that CDMA is already off in your area?
My AS10 (4G) would get either ATT or Tmobile before I started using airplane mode.
MacbookAir (Intel) macOS Monterey (12.6.8)
OSCAR 1.5.1
Resmed S10 Autoset w/heated hose
F30 Mask
Pressure 9-14
OSCAR Team

southerndoc
Posts: 242
Joined: Sat Apr 23, 2011 9:32 pm
Location: Marietta, GA USA

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by southerndoc » Wed Jan 26, 2022 9:39 am

I'm guessing the OP is asking to figure out when the machine will stop sending data to ResMed for myAir. AT&T is ditching 3G next month. If it's AT&T that ResMed is using, then myAir will stop working on some AS10 machines next month.

_________________
Machine: AirSense 11 Autoset
Mask: ResMed AirFit N30 Nasal CPAP Mask with Headgear
Additional Comments: OSCAR 1.3.1, SleepHQ

User avatar
chunkyfrog
Posts: 34385
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by chunkyfrog » Wed Jan 26, 2022 10:26 am

I never turned on MyAir.
Why, when I can see everything I need on the SCREEN?
My data gets to my pneumonologist, automatically.
and I can record my data manually on a spreadsheet.
Everything works just fine.
My metadata is mine, alone--nobody's business but mine.
Just the way I like it.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by JLROhio » Wed Jan 26, 2022 1:55 pm

Jeff8356 wrote:
Wed Jan 26, 2022 1:23 am
If the provider field is blank then it may not be connected at all. Scroll down a bit more and see if there is anything in the signal strength box. If that shows --dbm then there is no connection at all. Is it possible that CDMA is already off in your area?
My AS10 (4G) would get either ATT or Tmobile before I started using airplane mode.
It is connected to someone, somewhere - three bars on the antenna most of the time.
I just can't see who the cellular provider is because it's blank.

The over-watcher (VA) of my CPAP machine has access to my data and has accessed it before, as they have included my usage in my health records.
Also, the MyAir app does record data as well and works just fine (as worthless as they data is considered by some here).
Attachments
CDMA Screen.jpg
CDMA Screen.jpg (391.12 KiB) Viewed 4451 times
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by JLROhio » Wed Jan 26, 2022 2:01 pm

chunkyfrog wrote:
Wed Jan 26, 2022 10:26 am
My metadata is mine, alone--nobody's business but mine.
Are you sure about that? I haven't read the fine print, but if someone was to read the fine print of ResMed's paperwork, I'm would venture to guess that they probably indicate in the fine print, that your usage of the machine allows them to access the data, from the machine, for their own internal uses, studies, product improvement, etc... :shock:

You're correct that metadata should be OURs (and ours to determine how it is used)...but until the lawmakers make laws giving us the rights over our own metadata, it's free to whoever's lawyers write up the fine print on product usage data sheets or EULA's.

Facebook and Google make a lot of money off of our metadata! Heck, even CPAPTalk.com probably does too!
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--

User avatar
chunkyfrog
Posts: 34385
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by chunkyfrog » Wed Jan 26, 2022 2:22 pm

My cpap metadata has no connection to my mental health or any threat to society;
Therefore it falls under HIPAA, and "should" only be accessible to me
and any RELEVANT medical or insurance provider--as per current law.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
JLROhio
Posts: 378
Joined: Tue May 09, 2017 12:04 pm

Re: How to determine CDMA carrier for ResMed AirSense 10

Post by JLROhio » Wed Jan 26, 2022 2:30 pm

chunkyfrog wrote:
Wed Jan 26, 2022 2:22 pm
My cpap metadata has no connection to my mental health or any threat to society;
Therefore it falls under HIPAA, and "should" only be accessible to me
and any RELEVANT medical or insurance provider--as per current law.
metadata that does not include your personal information might not be HIPAA protected - I'm not a certified lawyer but I was considered one in the military (barracks/sea lawyer)! :P
* This is the internet - what I write will probably NOT come across as it would in person; where one can read a person's face.
My posts are sometimes tongue & cheek w/some snarky sarcasm mixed in.
I leave it up to you to figure it out!
:wink:

--