My Airfit P10 nose pillows slip off
-
- Posts: 13
- Joined: Wed Jul 07, 2021 8:19 pm
- Location: Thailand
My Airfit P10 nose pillows slip off
Hello everyone
I have a Resmed S9 and Airfit P10 and use an ergonomic pillow that has a 'valley' to help stabilize my head. At some point at night I toss and turn and the nose pillows get loose. Since they're no longer airtight they start blowing and whistling, waking my partner up.
I've had the cpap machine and Airfit P10 for just under a year, so they're all pretty new.
I was wondering if a good alternative would be the Airfit N20? Or should I go with something more involved?
thank you!
I have a Resmed S9 and Airfit P10 and use an ergonomic pillow that has a 'valley' to help stabilize my head. At some point at night I toss and turn and the nose pillows get loose. Since they're no longer airtight they start blowing and whistling, waking my partner up.
I've had the cpap machine and Airfit P10 for just under a year, so they're all pretty new.
I was wondering if a good alternative would be the Airfit N20? Or should I go with something more involved?
thank you!
_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Resmed Airsense 9 |
Resmed S9 36005A
36900A H5i
Airfit P10 nose pillows
36900A H5i
Airfit P10 nose pillows
- Dog Slobber
- Posts: 3961
- Joined: Thu Feb 15, 2018 2:05 pm
- Location: Ontario, Canada
Re: My Airfit P10 nose pillows slip off
Is the strap on your p10 the original strap?
If it is it likely doesn't have the elasticity to keep it tight. Replace it.
Newer P10 straps have a built in clip that gives the wearer more control over the tightness.
If it is it likely doesn't have the elasticity to keep it tight. Replace it.
Newer P10 straps have a built in clip that gives the wearer more control over the tightness.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Additional Comments: Min EPAP: 8.2, Max IPAP: 25, PS:4 |
Re: My Airfit P10 nose pillows slip off
If you like the P10’s it may simply be you need a new strap. My P10 straps stretch out after about 3 months. I have spares and rotate in a new one, then take the one stretched out and wash it with the delicates, run it through the dryer, and can typically use it for another 3 months. After a few cycles it eventually needs to be thrown out, but for me at least washing/drying them gets more life out of them.
As Dog Slobber said the new straps have a clip on the side that allows you to tighten them (vs the original clips that required a degree in engineering to use). I have one of those and it worked well when I tried it, but I have not used it for an extended time as yet.
You can check Amazon and there are some knock-off P10 straps being sold, but I have not tried them but others may want to chime in.
As Dog Slobber said the new straps have a clip on the side that allows you to tighten them (vs the original clips that required a degree in engineering to use). I have one of those and it worked well when I tried it, but I have not used it for an extended time as yet.
You can check Amazon and there are some knock-off P10 straps being sold, but I have not tried them but others may want to chime in.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: 5-25 PS 4 |
Sometimes it is the very people who no one imagines anything of who do the things that no one imagines
-
- Posts: 508
- Joined: Sat Feb 13, 2021 2:40 am
Re: My Airfit P10 nose pillows slip off
If you decide to buy the newer P10 headgear with the clip, buy the AirFit N30 headgear instead.
The two are the same.
This way you avoid the possibility of receiving old stock of P10 headgear without the clip.
Also make sure you don't accidently order the headgear for the AirFit N30i. It's not compatible with the P10.
The two are the same.
This way you avoid the possibility of receiving old stock of P10 headgear without the clip.
Also make sure you don't accidently order the headgear for the AirFit N30i. It's not compatible with the P10.
Last edited by GrumpyHere on Sun Feb 20, 2022 2:33 am, edited 1 time in total.
ResMed Lumis Tx - VAuto mode
- ChicagoGranny
- Posts: 14471
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: My Airfit P10 nose pillows slip off
You are getting some good advice about the headgear. Also, find your mask on youtube and carefully watch three or four videos about sizing and adjusting.
Re: My Airfit P10 nose pillows slip off
Be very careful. Mostly Made in China trash. Even the ones they claim to be Resmed straps are not.
Others claim to be Resmed compatible or replacement equal in quality. They are not.
Good luck.
JPB
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleep on a Buckwheat Hull Pillow. |
Re: My Airfit P10 nose pillows slip off
to add to the replies you're getting, did you know that putting the p10 head strap in with your laundry and dryer can shrink the strap back down a bit?
you might want to put it in a laundry bag first.
good luck!
you might want to put it in a laundry bag first.
good luck!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: My Airfit P10 nose pillows slip off
thanks for this. i knew there was a replacement, but couldn't remember the model number. i switched via dogslobber's recommendation. but i did so much cutting and hacking on it that i really need to replace it.GrumpyHere wrote: ↑Fri Jan 21, 2022 8:36 amIf you decide to buy the newer P10 headgear with the clip, buy the AirFit N30 headgear instead.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: My Airfit P10 nose pillows slip off
You asked about the N20. Although I prefer the P10 ( and the N30 ) I also have an N20. It has much more substantial headgear than the P10/N30 ( too much for my minimalist tastes ), so it might stay in place better for you than the P10.
For me the N20 is pretty comfortable, doesn't leak and the headgear holds the cushion very securely. Only negatives for me is that there is no diffuser on the mask elbow, so the exhaust can blow on your partner. It's quiet unless the vented air hits a pillow or something, and then it can be noisy. If you side-sleep then the magnetic mask buttons can sometimes press against your face and be disturbing. The cushion rests against the bridge of your nose -- not an issue for me but for some it is. I use this mask as an alternative if I need a break from a nasal pillow or nasal cradle.
For me the N20 is pretty comfortable, doesn't leak and the headgear holds the cushion very securely. Only negatives for me is that there is no diffuser on the mask elbow, so the exhaust can blow on your partner. It's quiet unless the vented air hits a pillow or something, and then it can be noisy. If you side-sleep then the magnetic mask buttons can sometimes press against your face and be disturbing. The cushion rests against the bridge of your nose -- not an issue for me but for some it is. I use this mask as an alternative if I need a break from a nasal pillow or nasal cradle.
Mask: alternate between P10 and N30; Vitera / F20 if allergies
Machine: AirSense 10 Autoset / AirMini
I came, I saw, I dozed off...
Machine: AirSense 10 Autoset / AirMini
I came, I saw, I dozed off...
- chunkyfrog
- Posts: 34395
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: My Airfit P10 nose pillows slip off
I asked my DME for the new, adjustable p10 strap.
They value my business, so they got it for me.
They value my business, so they got it for me.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: My Airfit P10 nose pillows slip off
They are now included in all the new P10 mask packages. Just got one the other day and they are promoting the new adjustable strap right on the package now. That's what cpap.com has now because that's where I got the new P10 package.chunkyfrog wrote: ↑Sat Jan 22, 2022 4:39 pmI asked my DME for the new, adjustable p10 strap.
They value my business, so they got it for me.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: My Airfit P10 nose pillows slip off
I bought three several days ago from cpap.com. When they got here yesterday all three were the new style.
So there you go. https://www.cpap.com/ - for all your P10 head gear needs.
JPB
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Sleep on a Buckwheat Hull Pillow. |
- BuckarooBanzai
- Posts: 139
- Joined: Mon Dec 27, 2021 10:48 pm
- Location: Sacramento, CA
Re: My Airfit P10 nose pillows slip off
Yes, I agree, your P10 issue likely relates to loose headgear. The new P10s do have new headgear with an adjustable strap.
I ordered an extra from thecpapshop.com just yesterday for $11.50.
By the way, I sleep on my side and stomach but have never had the pillows leak or disengage. BUT I have a large head, so the headgear is not a problem for me. Just something to think about.
Another thought: When deciding which size pillow to use, I was told to use the LARGEST pillow that would work for me. This ensures a snug and leak-free fit. Could it be you are using the wrong size pillow?
I ordered an extra from thecpapshop.com just yesterday for $11.50.
By the way, I sleep on my side and stomach but have never had the pillows leak or disengage. BUT I have a large head, so the headgear is not a problem for me. Just something to think about.
Another thought: When deciding which size pillow to use, I was told to use the LARGEST pillow that would work for me. This ensures a snug and leak-free fit. Could it be you are using the wrong size pillow?
- chunkyfrog
- Posts: 34395
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: My Airfit P10 nose pillows slip off
I can hardly wait until my hair has grown out enough to cover the grooves in my scalp.
I understand that they were always there, but seeing them is disturbing.
I understand that they were always there, but seeing them is disturbing.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |