Did I damage my APAP?
Did I damage my APAP?
Hi All,
I have an Airsense 10 APAP. Last night, I knocked it off of the nighttable and onto the floor. (I've never done this before, by the way!)
I picked up the APAP, put it right side up, and tried to go back to sleep. The APAP kept running, but there was a very loud gurgling noise every time I breathed. It turned out that either when the APAP fell, or when I picked it up and righted it, it had turned over enough that water from the humidifier had gotten into the pump and gotten all the way to my hose.
After I removed the hose and drained it, the machine seemed to function fine for the rest of the night. But I'm worried about what the water might have done to the air pumping mechanism or other innards of the machine.
Can anyone tell me if I did or did not damage the machine?
Thanks,
Andy
I have an Airsense 10 APAP. Last night, I knocked it off of the nighttable and onto the floor. (I've never done this before, by the way!)
I picked up the APAP, put it right side up, and tried to go back to sleep. The APAP kept running, but there was a very loud gurgling noise every time I breathed. It turned out that either when the APAP fell, or when I picked it up and righted it, it had turned over enough that water from the humidifier had gotten into the pump and gotten all the way to my hose.
After I removed the hose and drained it, the machine seemed to function fine for the rest of the night. But I'm worried about what the water might have done to the air pumping mechanism or other innards of the machine.
Can anyone tell me if I did or did not damage the machine?
Thanks,
Andy
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Using OSCAR |
Resmed Airsense Autoset 10
Resmed Airfit F20 Mask
Resmed Airfit F20 Mask
Re: Did I damage my APAP?
Generally speaking if the machine seems to run normally now then the water didn't make inside the machine to fry the electronics.
Probably what happened with the overturn was water took the path of least resistance and when from water chamber into the long hose and that's where all the noise came from.
While not totally impossible for water to go into the motor side of things....not nearly as common as it was years ago.
The water chambers or purposely designed to not be so easy for upturned water chamber contents to go into the motor side of things. Instead designed to go to the long hose.
When electronics meet water...you usually know it...either it won't power up at all or there is a god awful noise.
If it is moving air like it should and not making a god awful noise in the motor....it's probably fine.
Probably what happened with the overturn was water took the path of least resistance and when from water chamber into the long hose and that's where all the noise came from.
While not totally impossible for water to go into the motor side of things....not nearly as common as it was years ago.
The water chambers or purposely designed to not be so easy for upturned water chamber contents to go into the motor side of things. Instead designed to go to the long hose.
When electronics meet water...you usually know it...either it won't power up at all or there is a god awful noise.
If it is moving air like it should and not making a god awful noise in the motor....it's probably fine.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Did I damage my APAP?
Well put!! Thanks for your help, Pugsy!When electronics meet water...you usually know it
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Using OSCAR |
Resmed Airsense Autoset 10
Resmed Airfit F20 Mask
Resmed Airfit F20 Mask
Re: Did I damage my APAP?
this reminds of my stint as an admin at sony service center. at one time, sony was running ads showing happy young people cavorting on the beach capturing their precious memories on video tape. this drove our manager nuts. "don't they realize that water and electronic don't mix?!?!".
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Did I damage my APAP?
Actually they do mix, but the result is less than satisfying. With the XPAP running unless the air intake was in the water stream, water would not get in the pump, the air flow would Keep the water out of the machine, and blow it into the hose from HH. Causing the noises you heard. Waterboarding Anyone?zonker wrote: ↑Thu Jan 13, 2022 11:44 am
this reminds of my stint as an admin at sony service center. at one time, sony was running ads showing happy young people cavorting on the beach capturing their precious memories on video tape. this drove our manager nuts. "don't they realize that water and electronic don't mix?!?!".
Use data to optimize your xPAP treatment!
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
"The art of medicine consists in amusing the patient while nature cures the disease." Voltaire
- chunkyfrog
- Posts: 34377
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Did I damage my APAP?
I remember the Lucas magic smoke.
Click and Clack--one podcast I miss terribly.
Click and Clack--one podcast I miss terribly.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Did I damage my APAP?
So, there's the blower unit, THEN the humidifier, THEN the hose, so it's entirely possible, especially if it was running at the time, that no water got into the machine itself, that it just got blown into the hose.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Did I damage my APAP?
Ah, I didn't think about it that way! Thank you, that's very comforting. I thought it was humidifier, then blower, then hose.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Using OSCAR |
Resmed Airsense Autoset 10
Resmed Airfit F20 Mask
Resmed Airfit F20 Mask
Re: Did I damage my APAP?
I did the exact same thing about a week ago. And sat it back up on the table upside down, lol. It is working fine and seems to have suffered no damage. I have since put some "museum gel" on the bottom so it won't go anywhere.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
ResMed AirSense Autoset FOR HER/F30i
Re: Did I damage my APAP?
Easy enough mistake to make with the way the air circuit appears to go back into the machine, but if you look in the hole with the chamber out, the left opening goes to the blower assembly, and the right opening is an elbow that goes out the back of the machine.andy88488 wrote: ↑Thu Jan 13, 2022 2:11 pmAh, I didn't think about it that way! Thank you, that's very comforting. I thought it was humidifier, then blower, then hose.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Did I damage my APAP?
Yep! You never want to lose the magic smoke.chunkyfrog wrote: ↑Thu Jan 13, 2022 1:24 pmI remember the Lucas magic smoke.
Click and Clack--one podcast I miss terribly.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
Do not regret growing older. It is a privilege denied to many...
- babydinosnoreless
- Posts: 2325
- Joined: Fri Nov 02, 2018 2:53 pm
Re: Did I damage my APAP?
I did this with my aircurve. I put it in front of a fan and let it dry all day, then plugged it in the next night and it worked. Thank goodness !
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
Re: Did I damage my APAP?
Yeah, I took a look at that last night, and sure enough, you are 100% correct. Thanks, you've set my mind at ease.palerider wrote: ↑Thu Jan 13, 2022 4:49 pmEasy enough mistake to make with the way the air circuit appears to go back into the machine, but if you look in the hole with the chamber out, the left opening goes to the blower assembly, and the right opening is an elbow that goes out the back of the machine.andy88488 wrote: ↑Thu Jan 13, 2022 2:11 pmAh, I didn't think about it that way! Thank you, that's very comforting. I thought it was humidifier, then blower, then hose.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: ResMed AirFit™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: Using OSCAR |
Resmed Airsense Autoset 10
Resmed Airfit F20 Mask
Resmed Airfit F20 Mask
- SleepyCPAP
- Posts: 333
- Joined: Wed Dec 08, 2010 6:01 am
Re: Did I damage my APAP?
I took apart my ResMed Airsense 10 AutoSet apart when I first got it (it was well used, and needed to be de-smelled). There is no way for the water to go from the chamber to the electronics board(s), it would just come out the filter if you were adept at pouring water in there.
It is possible to get water from the humidifier into the motor chamber on a fall with water in the humidifier, but you have to be particularly skilled at doing it. I apparently do have that skill, when my machine came off the stand one night. I shook the machine to get drops out of the motor chamber (never opened it up) and let it dry out for three days while I used my backup machine. I don’t think water ever made it into the motor/blower itself. It runs as well as ever.
- SleepyCPAP
It is possible to get water from the humidifier into the motor chamber on a fall with water in the humidifier, but you have to be particularly skilled at doing it. I apparently do have that skill, when my machine came off the stand one night. I shook the machine to get drops out of the motor chamber (never opened it up) and let it dry out for three days while I used my backup machine. I don’t think water ever made it into the motor/blower itself. It runs as well as ever.
- SleepyCPAP
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Use OSCAR. Combine AlaxoStent with VAuto for perfect 0.0 AHI at PS 3.6 over 4cm EPAP |
-- SleepyCPAP
Sleep study in 2010 (11cm CPAP). Pillows (Swift FX>TAP PAP >Bleep). PRS1 “Pro” 450/460 until recall, now Aircurve 10 VAuto. Tape mouth. Palatal Prolapse solved by AlaxoStent & VAuto EPAP 4cm, PS 3.6cm = 0.0 AHI
Sleep study in 2010 (11cm CPAP). Pillows (Swift FX>TAP PAP >Bleep). PRS1 “Pro” 450/460 until recall, now Aircurve 10 VAuto. Tape mouth. Palatal Prolapse solved by AlaxoStent & VAuto EPAP 4cm, PS 3.6cm = 0.0 AHI