WheezyRider's CPAP therapy
-
- Posts: 42
- Joined: Wed Jun 23, 2021 3:56 pm
WheezyRider's CPAP therapy
It's been taking me awhile to actually fall asleep. Last night my AHI peaked at 31 before I was even sleeping, according to Oscar. Then fell off to around 7 for the rest of the night, presumably after I fell asleep. I've noticed this on other nights where I wake up and then have trouble falling back to sleep and the AHI is high while I'm awake. Is this normal?
edit:
learning how to post pictures:
edit2: single page, thumbs UP?
edit:
learning how to post pictures:
edit2: single page, thumbs UP?
- Attachments
-
- single.png (66.27 KiB) Viewed 41815 times
Last edited by wheezyrider on Sun Dec 26, 2021 9:31 am, edited 3 times in total.
Re: How can AHI be high while awake?
If you're not breathing regularly while awake, then it's "normal". (and awake breathing is often quite irregular.)wheezyrider wrote: ↑Fri Dec 24, 2021 7:39 amIt's been taking me awhile to actually fall asleep. Last night my AHI peaked at 31 before I was even sleeping, according to Oscar. Then fell off to around 7 for the rest of the night, presumably after I fell asleep. I've noticed this on other nights where I wake up and then have trouble falling back to sleep and the AHI is high while I'm awake. Is this normal?
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 52
- Joined: Wed Dec 22, 2021 3:20 am
Re: How can AHI be high while awake?
i don't know, but I do know that my oxygen saturation falls when I prepare for sleep, while I am yet still awake. I don't so far have enough data from my CPAP machine to know if other things happen as well (I've only had it a week).
In any case, I know that it's possible for my breathing pattern to change whilst I'm awake but sleepy -- enough to cause a desaturation, but not enough that I become aware of it.
I also would be quite interested to know why this kind of thing happens.
Best wishes
Lars.
In any case, I know that it's possible for my breathing pattern to change whilst I'm awake but sleepy -- enough to cause a desaturation, but not enough that I become aware of it.
I also would be quite interested to know why this kind of thing happens.
Best wishes
Lars.
ResMed S9 AutoSet / Hoffrichter standard full face
Re: How can AHI be high while awake?
Concentrate on asleep breathing - the machine won't do valid awake therapy or results.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
Re: How can AHI be high while awake?
unlike santa claus, the machine doesn't know when you are sleeping or when you are awake!wheezyrider wrote: ↑Fri Dec 24, 2021 7:39 amIt's been taking me awhile to actually fall asleep. Last night my AHI peaked at 31 before I was even sleeping, according to Oscar. Then fell off to around 7 for the rest of the night, presumably after I fell asleep. I've noticed this on other nights where I wake up and then have trouble falling back to sleep and the AHI is high while I'm awake. Is this normal?
as palerider says, awake breathing isn't smooth breathing.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- Miss Emerita
- Posts: 3443
- Joined: Sun Nov 04, 2018 8:07 pm
Re: How can AHI be high while awake?
What kinds of events were you having while you were awake? Centrals? Obstructive apneas? Hypopneas?
I'm also curious about the rest of your night. If your AHI while you were asleep was around 7, it'd sure be nice to get it down lower. Could you post an Oscar chart so people here can take a look?
I'm also curious about the rest of your night. If your AHI while you were asleep was around 7, it'd sure be nice to get it down lower. Could you post an Oscar chart so people here can take a look?
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
- Posts: 42
- Joined: Wed Jun 23, 2021 3:56 pm
Re: How can AHI be high while awake?
what parts should I post? Not sure I can post pics yet though.Miss Emerita wrote: ↑Fri Dec 24, 2021 12:34 pmWhat kinds of events were you having while you were awake? Centrals? Obstructive apneas? Hypopneas?
I'm also curious about the rest of your night. If your AHI while you were asleep was around 7, it'd sure be nice to get it down lower. Could you post an Oscar chart so people here can take a look?
-
- Posts: 42
- Joined: Wed Jun 23, 2021 3:56 pm
Re: How can AHI be high while awake?
Here is my chart. Are centrals the same as Clear Airways? I definitely want to get lower, but lately AHI's have been really high. This was a good night. I have restless leg and it has flair up, so many of my recent readings are not good from that, I think because I'm awake alot of the night. I took a muscle relaxant, very small dose, and my PCP knows I have SA and using CPAP, but I think it helped.Miss Emerita wrote: ↑Fri Dec 24, 2021 12:34 pmWhat kinds of events were you having while you were awake? Centrals? Obstructive apneas? Hypopneas?
I'm also curious about the rest of your night. If your AHI while you were asleep was around 7, it'd sure be nice to get it down lower. Could you post an Oscar chart so people here can take a look?
thanks!
EDIT: charts removed to save space, see 1st comment for report.
Last edited by wheezyrider on Sun Dec 26, 2021 8:30 am, edited 1 time in total.
- Miss Emerita
- Posts: 3443
- Joined: Sun Nov 04, 2018 8:07 pm
Re: How can AHI be high while awake?
Many thanks for posting the chart. A key question for you is whether you were awake for around two and a half hours -- when the bulk of those centrals (aka clear airway, aka CA) occurred. If you were that'll be one story, but if you weren't, it's another. The right advice for you will depend on what you can tell us about the length of your awake period.
No need at all to repost this chart, but next time you post one, please show only these graphs:
Events
Flow rate
Pressure
Leaks
Snores
Flow limitations.
You can squeeze them into a single screenshot by grabbing the horizontal gray bars that separate the graphs and pushing them up a little. You will also need to reorder your graphs. Left-click and hold while you push graphs up or down.
About the muscle relaxant: please let your PCP know that you are experiencing central apneas, just to make sure the drug is right for you. The PCP may assume that the only potential concern would be obstructive events.
No need at all to repost this chart, but next time you post one, please show only these graphs:
Events
Flow rate
Pressure
Leaks
Snores
Flow limitations.
You can squeeze them into a single screenshot by grabbing the horizontal gray bars that separate the graphs and pushing them up a little. You will also need to reorder your graphs. Left-click and hold while you push graphs up or down.
About the muscle relaxant: please let your PCP know that you are experiencing central apneas, just to make sure the drug is right for you. The PCP may assume that the only potential concern would be obstructive events.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Oscar software is available at https://www.sleepfiles.com/OSCAR/
-
- Posts: 42
- Joined: Wed Jun 23, 2021 3:56 pm
Re: How can AHI be high while awake?
Miss Emerita wrote: ↑Fri Dec 24, 2021 5:08 pmwhether you were awake for around two and a half hours
About the muscle relaxant
Looks like 1-1/2 hours and I am carful about the meds and only 2mg's. It was for my back before bed when needed as prescribed. The legs have a jittery sensation. It was in enough discomfort that the sensation in my legs needed to chill so I can get some needed sleep. Insomnia is my nemesis. and I wasn't too afraid that 2mg Tizanidine would put me in a coma. My doctor should be aware that my apneas are central because he sent me to a ENT before the sleep study. I'll try to fall asleep au naturel, but if I'm still staring at the clock at midnite, I'm poppin' one.
Thanks for your analysis. That really helps.
I'll update the top pictures,
Re: How can AHI be high while awake?
Muscle relaxant meds are more likely to worse the obstructive side of sleep apnea than cause centrals.
Muscle relaxants tend to make the floppy airway tissues more floppy. At least that is what my sleep doc told me and cautioned me to not take muscles relaxants to help me sleep.
Muscle relaxants tend to make the floppy airway tissues more floppy. At least that is what my sleep doc told me and cautioned me to not take muscles relaxants to help me sleep.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
- chunkyfrog
- Posts: 34391
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: How can AHI be high while awake?
Consciously breathing regularly is very relaxing.
It works when you don't want to waste that last alprazolam.
It works when you don't want to waste that last alprazolam.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
-
- Posts: 42
- Joined: Wed Jun 23, 2021 3:56 pm
Re: How can AHI be high while awake?
That makes sense to my little brain. I'm also concerned with falling asleep faster. My mind takes awhile to wind down some times. It's like I'm hooked up to an alien probe to upload the important news from planet earth. there's a lot of crap to brief them on.Pugsy wrote: ↑Fri Dec 24, 2021 6:48 pmMuscle relaxant meds are more likely to worse the obstructive side of sleep apnea than cause centrals.
Muscle relaxants tend to make the floppy airway tissues more floppy. At least that is what my sleep doc told me and cautioned me to not take muscles relaxants to help me sleep.
Does the rest of the night look OK? I'm really just trying to learn to read the report. On many nights the ahi is going up when I am waking up. Though the night posted looks like it glides into morning. Like my sleep pressure is wearing off. Is that good? Mental image of me surfing down a killer wave and just ride it all the way to the beach for a soft landing..........
Sleep Well!
-
- Posts: 42
- Joined: Wed Jun 23, 2021 3:56 pm
Re: How can AHI be high while awake?
had to look it up.chunkyfrog wrote: ↑Fri Dec 24, 2021 7:20 pmConsciously breathing regularly is very relaxing.
It works when you don't want to waste that last alprazolam.
"Is alprazolam safe to use?
[Alprazolam is a safe and effective medication when used as directed. Benzodiazepines may produce emotional and/or physical dependence (addiction) even when used as recommended."
I'll stick with indica.
- chunkyfrog
- Posts: 34391
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: How can AHI be high while awake?
I wish it was legal here.
We found an acapulco giant strain in the yard of a house we just moved into.
Removed the plants before the evictees could sneak back in for it.
We found an acapulco giant strain in the yard of a house we just moved into.
Removed the plants before the evictees could sneak back in for it.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |