Tear in Hose Tubing: Recommendations for Repair
Re: Tear in Hose Tubing: Recommendations for Repair
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
Re: Tear in Hose Tubing: Recommendations for Repair
You've out'd me Zonker, as an unabashed fan of the ol' Red Green show.
"If it doesn't stick and it should, apply Duct Tape. If it sticks and it shouldn't, apply WD40."
But seriously, any health concern over "fumes from duct tape" needs to be tempered with what is it's "defined purpose in life" -- which is to SEAL AIR VENTILATION DUCTS. Maybe it causes cancer, I don't know, but if it does then we are all doomed. DOOMED I say.
"If it doesn't stick and it should, apply Duct Tape. If it sticks and it shouldn't, apply WD40."
But seriously, any health concern over "fumes from duct tape" needs to be tempered with what is it's "defined purpose in life" -- which is to SEAL AIR VENTILATION DUCTS. Maybe it causes cancer, I don't know, but if it does then we are all doomed. DOOMED I say.
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P30i Nasal Pillow CPAP Mask with Headgear Starter Pack |
Re: Tear in Hose Tubing: Recommendations for Repair
well, let's drag this WAAAYYY off topic, then.
long after the last episode, i got the chance to see steve smith on tour. as you may imagine, he did it as red. it was quite an enjoyable show, even without the rest of the cast.
before going in, we had a drink at the theater's bar. was drinking wine and looking out of the second story window. i'm the right age and have that same type of beard. many people going into the show down below kept nudging each other and pointing at me.
i think they really thought *I* was steve!
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Tear in Hose Tubing: Recommendations for Repair
Instead of worrying about unknown evaporative chemicals,
my "emergency" rule is if it smells bad keep it away from my face, nose, etc.
Duck brand is practically odorless--I would only use it for the shortest time possible.
Some generics REEK--don't even buy them.
my "emergency" rule is if it smells bad keep it away from my face, nose, etc.
Duck brand is practically odorless--I would only use it for the shortest time possible.
Some generics REEK--don't even buy them.
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Tear in Hose Tubing: Recommendations for Repair
I'm sure that your machine, like my AirCurve, came with a standard non-heated hose, right? Why not use this till you get a replacement heated hose?bkeberl wrote: ↑Thu Dec 09, 2021 10:32 pmHi All,
This morning while trying to remove the mask elbow from the silicone sleeve on the hose tubing, the tubing pulled away from the silicone end (see picture below). Any recommendations for a quick repair?
IMG_6451.jpg
I have the ResMed AirSense S11, so most replacement parts will be hard to come by over the next few months until the device becomes more widely available.
Will traditional electrical or duck tape work? Alternatively, I’ve considered cutting away the ripped plastic and gluing the tubing back into the elbow ‘housing’.
Any advice is appreciated as I’m very new to CPAP.
Thanks,
Brandon
My son recently got the same machine that you have. His DME sent a heated hose along with his machine. Unfortunately, they sent the wrong hose, it was for a AirSense 10. He informed them of the mistake, and they sent the correct hose right away and told him to toss the incorrect hose. Unfortunate for him but fortunate for me as this is the hose that fits my machine.
I see a lot of talk here of using tape to fix your broken hose. A few folks are worried about the chemicals that you may be inhaling from the tape. Think about how a CPAP works for a moment. How would any chemicals from the tape get into your hose when there is positive pressure in the hose? If there was a vacuum in the hose yes, positive pressure, no.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Swift™ FX Nasal Pillow CPAP Mask with Headgear |
Humidifier: HumidAire H4i™ Heated Humidifier |
ResMed AirCurve 10 Vauto Swift FX
Do not regret growing older. It is a privilege denied to many...
Do not regret growing older. It is a privilege denied to many...
Re: Tear in Hose Tubing: Recommendations for Repair
I think I would avoid glue or adhesive since breathing the fumes would not be nice. If your worried about the smell of adhesive on tape I would look at the self fusing tapes that are available. They do not use adhesive to make the bond. They typically are made of silicone like our mask cushions and bond to themselves when you stretch and wrap them around something. That's what I would try if I needed to repair the hose and I wouldn't worry about leaving it that way indefinitely. My local hardware stores carry a brand called Rescue tape in the plumbing section. They claim it's waterproof and air tight. Hopefully the planets will align and you can find a new hose soon.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N10 Nasal CPAP Mask with Headgear |
Additional Comments: Resmed F10 Mask for colds. When camping on battery power I use P10 mask and PR 560p machine. |
Re: Tear in Hose Tubing: Recommendations for Repair
I have had to repair many hoses, usually due to an aggressive cat.
Whatever the reason, I have used this type of tape many times.
https://www.amazon.com/Seal-Self-Fusing ... 468&sr=8-5
You can find similar tape at your hardware store.
Whatever the reason, I have used this type of tape many times.
https://www.amazon.com/Seal-Self-Fusing ... 468&sr=8-5
You can find similar tape at your hardware store.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset |
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Re: Tear in Hose Tubing: Recommendations for Repair
Will echo the earlier reply that you need to get a new hose. Any repair should be considered as a temporary fix until the replacement arrives. Even with the perfect tape repair you now have a cracked hose interior that will collect dust, moisture and possibly grow bacteria. Also with a tape repair it will be challenging to periodically wash the hose and is typically recommended.
Re: Tear in Hose Tubing: Recommendations for Repair
If you still need a repair go to Walmart and get some self adhering tape https://www.walmart.com/search?q=self%2 ... ing%20tape
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1 |
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Tear in Hose Tubing: Recommendations for Repair
Bkeberi asked the question over a month ago.
It would be nice to hear how he made out.
Only 2 posts--quit cpap?
Or just this forum?
It would be nice to hear how he made out.
Only 2 posts--quit cpap?
Or just this forum?
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Re: Tear in Hose Tubing: Recommendations for Repair
That type of self adhering tape as shown in the link above is porous. The air will leak out.squid13 wrote: ↑Tue Jan 18, 2022 9:52 amIf you still need a repair go to Walmart and get some self adhering tape https://www.walmart.com/search?q=self%2 ... ing%20tape
This is great stuff for bandage use. Not so good for CPAP.
_________________
Machine: AirSense 11 Autoset |
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear |
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset |
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto
Re: Tear in Hose Tubing: Recommendations for Repair
Actually I have some of that tape and in a pinch it could be used and just make multiple layer wraps and it will stop the leak well enough until a replacement hose can be obtained.jsmit86 wrote: ↑Thu Jan 20, 2022 7:44 pmThat type of self adhering tape as shown in the link above is porous. The air will leak out.squid13 wrote: ↑Tue Jan 18, 2022 9:52 amIf you still need a repair go to Walmart and get some self adhering tape https://www.walmart.com/search?q=self%2 ... ing%20tape
This is great stuff for bandage use. Not so good for CPAP.
One layer probably not enough though....and one can look at their leak graphs to see if it stopped the leak well enough.
Most holes in the hose ...or tears...will show a big enough leak consistently that it will be obvious when the leak gets stopped on the leak graphs.
Been there and done that just recently when it was determined that I had a small hole in my hose.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: Tear in Hose Tubing: Recommendations for Repair
Like Pugsy said you would use multiple layer wraps, wrap it good enough and it will stop the leak.
_________________
Machine: ResMed AirCurve 10 ASV Machine with Heated Humidifier |
Mask: Evora Full Face Mask - Fitpack |
Additional Comments: AirCurve 10 ASV, Oscar V1.0.1-r-1 |
US Navy Retired 1973,AirCurve 10 ASV, Mode: ASV Auto, Min EPAP: 7.2, Max EPAP: 15.0, Min PS:4.0, Max PS: 15.0, Mask ResMed Airtouch F20, Backup: (2) AirCurve 10 ASV
- chunkyfrog
- Posts: 34378
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Tear in Hose Tubing: Recommendations for Repair
The first post was December 9, 2021.
---well over a month ago.
I think we can all assume the issue was resolved--somehow . . .
---well over a month ago.
I think we can all assume the issue was resolved--somehow . . .
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |