Resmed Airsense 10 humidifier seal

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
BradleyBuck1984
Posts: 10
Joined: Sat Nov 06, 2021 3:55 pm

Resmed Airsense 10 humidifier seal

Post by BradleyBuck1984 » Sat Nov 06, 2021 4:00 pm

Hello, I'm having a recurring problem where I'll use my CPAP effectively for a couple of months, but eventually it always gets to the point where, although the ahi is still under 1, I don't feel as if I'm getting enough pressure. Usually, when everything is working properly, I inhale and the machine forces the air in without much work on my part. When I exhale I feel resistance but if I relax it all works fine. I think this is all normal, but eventually I lose that sense of resistance on exhale and I actually have to work to inhale rather than just let the machine do it. It makes it difficult to fall asleep, and although the ahi doesn't really change I wake up more and feel worse in the morning. I've noticed that replacing the humidifier tank solves this problem, but the NHS insists that they should last at least a year. I'm not able to get past 3 months before I have this issue. I wash the tank and seal daily, as instructed and dry them in front of a fan. Same goes for my mask. (FYI, my minimum pressure is 11.6) I'm gathering that the silicone is degrading and not making as good a seal, but I'm not sure how to avoid this. Can I purchase a replacement gasket? Is there a way to improve the seal after the silicone has begun to degrade? Thanks for your help.

User avatar
ChicagoGranny
Posts: 14459
Joined: Sun Jan 29, 2012 1:43 pm
Location: USA

Re: Resmed Airsense 10 humidifier seal

Post by ChicagoGranny » Sat Nov 06, 2021 4:51 pm

BradleyBuck1984 wrote:
Sat Nov 06, 2021 4:00 pm
I wash the tank and seal daily, as instructed and dry them in front of a fan. Same goes for my mask. (FYI, my minimum pressure is 11.6)
Besides the "instructions" are there any other reasons you are washing and drying the humidifier tank daily? Very few of us do this. When I use a humidifier, I just dump the water in the morning and replace it at bedtime. I wonder if your daily removing, washing, drying and replacing of the tank is causing abnormal wear.

Also, what is your leak level value on the nights you have the problem?

User avatar
LSAT
Posts: 13232
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Resmed Airsense 10 humidifier seal

Post by LSAT » Sat Nov 06, 2021 5:07 pm

I have used 3 different Resmed machines in the 13 years I have been on CPAP. I have never replaced a water tub or the tub seal in that time. I use distilled water and wash my water tub about once a year. That's not you problem unless you are damaging the seal by pulling it out and washing it so often. It takes a long time to degrade the seals.

BradleyBuck1984
Posts: 10
Joined: Sat Nov 06, 2021 3:55 pm

Re: Resmed Airsense 10 humidifier seal

Post by BradleyBuck1984 » Mon Nov 08, 2021 7:21 am

ChicagoGranny wrote:
Sat Nov 06, 2021 4:51 pm

Besides the "instructions" are there any other reasons you are washing and drying the humidifier tank daily? Very few of us do this. When I use a humidifier, I just dump the water in the morning and replace it at bedtime. I wonder if your daily removing, washing, drying and replacing of the tank is causing abnormal wear.

Also, what is your leak level value on the nights you have the problem?

It's only because the sleep tech instructed me to wash it daily. I actually used to wash it weekly, but would still remove the gasket for drying. I'm not sure about the leak level. The machine gives a green happy face in the morning, so there shouldn't be a large leak.

BradleyBuck1984
Posts: 10
Joined: Sat Nov 06, 2021 3:55 pm

Re: Resmed Airsense 10 humidifier seal

Post by BradleyBuck1984 » Mon Nov 08, 2021 7:24 am

LSAT wrote:
Sat Nov 06, 2021 5:07 pm
I have used 3 different Resmed machines in the 13 years I have been on CPAP. I have never replaced a water tub or the tub seal in that time. I use distilled water and wash my water tub about once a year. That's not you problem unless you are damaging the seal by pulling it out and washing it so often. It takes a long time to degrade the seals.
Would using regular tap water cause the silicone to degrade? I live in the UK, and distilled water isn't common here.

User avatar
Respirator99
Posts: 341
Joined: Mon Jul 05, 2021 12:39 am
Location: Australia

Re: Resmed Airsense 10 humidifier seal

Post by Respirator99 » Mon Nov 08, 2021 8:35 am

BradleyBuck1984 wrote:
Mon Nov 08, 2021 7:24 am
Would using regular tap water cause the silicone to degrade? I live in the UK, and distilled water isn't common here.
Unless your tap water is unusually high in minerals (eg calcium) there should be no problem. Even so, daily washing of the tank is unrealistic and probably doing more harm than good. These tanks should last a very long time - the one on my S9 was still like new after 7 years, with occasional washing only.

If your machine is giving you a green smiley face, that's an indication that your leaks are not very excessive - it's not an indication that there are no leaks. I'd suggest you download and install Oscar, which is free software that will give you (and us, if you choose to share) a much more detailed look at what's going on. https://www.sleepfiles.com/OSCAR/
* Download Oscar
* Oscar help
* An alternative to Oscar - try SleepHQ

I have no medical training or qualifications. Take my advice for what it's worth.

BradleyBuck1984
Posts: 10
Joined: Sat Nov 06, 2021 3:55 pm

Re: Resmed Airsense 10 humidifier seal

Post by BradleyBuck1984 » Mon Nov 08, 2021 9:38 am

I've installed OSCAR and looked at the data. Aside from last night, I see few leaks. Last night I used a new hose and cushion (brevida) and for the last two nights I've been using my wife's tank (which she hasn't been using) and I noticed a clear improvement with her tank and moreso with the new hose and cushion, but the leak data shows no real difference.

User avatar
Dog Slobber
Posts: 3960
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Resmed Airsense 10 humidifier seal

Post by Dog Slobber » Mon Nov 08, 2021 10:27 am

I hate to be the guy that suggests, this but there may not be a problem at all, or a problem with your tank.

We are not really very good at judging pressures, especially after we've become desensitised to it.. Case in point when we start the CPAP as night we feel the pressure, but shortly there after, we don't feel it at all. I have nights where I feel the pressure more, and other nights where there is just no resistance (I actually don't like this feel at all). But I don't attribute the feeling to a difference in behaviour of the machine. Chances are if there was a difference, it would show up in the OSCAR graphs. Pressure trace, Mask Pressure Trace, Leak lines.

The other thing, you may have created a false association of this feeling with an aging tank. We are pattern recognising creatures, and often when we don't see patterns, we create them.

But some things to try:
  • In OSCAR make a note when you change tanks, compare. The traces with a well-behaved tank and non-behaving tank.
  • Swap tank parts with your wife's tank. Swap just the seal, just the top, just the bottom. Which part swap exhibits the problem.
  • Swap devices with your wife, does the problem move with the device?
  • On the bottom of misbehaving tank, with a sharpie, mark a little x, where it's not easy to see. Have your wife mix them up so you don't know which is the misbehaving one. Use it. Did you experience the problem? check if it's the x-marked tank? Do this for a few nights, You should be able to predict 100% of the time which tank has the mark.

User avatar
chunkyfrog
Posts: 34385
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: Resmed Airsense 10 humidifier seal

Post by chunkyfrog » Mon Nov 08, 2021 11:39 am

Your sleep tech is full of feces. (Faeces--your spelling)
Daily cleaning WEARS STUFF OUT
Stop taking everything apart--putting something back cockeyed is way too easy.
That itself will damage parts, so they may never reassemble properly again.
Check the seal for creases where they should not be.
Demand a replacement seal if that is the case.
Blame the stupid sleep tech if anyone asks what happened.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed Airsense 10 humidifier seal

Post by palerider » Mon Nov 08, 2021 12:57 pm

BradleyBuck1984 wrote:
Mon Nov 08, 2021 7:21 am
. The machine gives a green happy face in the morning, so there shouldn't be a large leak.
You get a green face if you've spent less than 30% of the night with a large leak. To know what's actually happening, you need to use Oscar.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed Airsense 10 humidifier seal

Post by palerider » Mon Nov 08, 2021 12:58 pm

BradleyBuck1984 wrote:
Mon Nov 08, 2021 7:24 am
Would using regular tap water cause the silicone to degrade? I live in the UK, and distilled water isn't common here.
No, it won't harm the silicone at all, that's why your manual doesn't specify distilled water :)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

BradleyBuck1984
Posts: 10
Joined: Sat Nov 06, 2021 3:55 pm

Re: Resmed Airsense 10 humidifier seal

Post by BradleyBuck1984 » Mon Nov 08, 2021 2:50 pm

chunkyfrog wrote:
Mon Nov 08, 2021 11:39 am
Your sleep tech is full of feces. (Faeces--your spelling)
Daily cleaning WEARS STUFF OUT
Stop taking everything apart--putting something back cockeyed is way too easy.
That itself will damage parts, so they may never reassemble properly again.
Check the seal for creases where they should not be.
Demand a replacement seal if that is the case.
Blame the stupid sleep tech if anyone asks what happened.

Maybe that's it. I haven't always washed the tank daily, but I typically remove the gasket for drying. I suppose the repeated removal could worsen the seal. Since I've started washing daily, it definitely seems worse. Do you think the nasal pillows and parts which have direct contact with the face need washed daily? I assumed the skin oils would degrade the silicone without washing.

BradleyBuck1984
Posts: 10
Joined: Sat Nov 06, 2021 3:55 pm

Re: Resmed Airsense 10 humidifier seal

Post by BradleyBuck1984 » Mon Nov 08, 2021 2:55 pm

Dog Slobber wrote:
Mon Nov 08, 2021 10:27 am
I hate to be the guy that suggests, this but there may not be a problem at all, or a problem with your tank.

We are not really very good at judging pressures, especially after we've become desensitised to it.. Case in point when we start the CPAP as night we feel the pressure, but shortly there after, we don't feel it at all. I have nights where I feel the pressure more, and other nights where there is just no resistance (I actually don't like this feel at all). But I don't attribute the feeling to a difference in behaviour of the machine. Chances are if there was a difference, it would show up in the OSCAR graphs. Pressure trace, Mask Pressure Trace, Leak lines.

The other thing, you may have created a false association of this feeling with an aging tank. We are pattern recognising creatures, and often when we don't see patterns, we create them.

But some things to try:
  • In OSCAR make a note when you change tanks, compare. The traces with a well-behaved tank and non-behaving tank.
  • Swap tank parts with your wife's tank. Swap just the seal, just the top, just the bottom. Which part swap exhibits the problem.
  • Swap devices with your wife, does the problem move with the device?
  • On the bottom of misbehaving tank, with a sharpie, mark a little x, where it's not easy to see. Have your wife mix them up so you don't know which is the misbehaving one. Use it. Did you experience the problem? check if it's the x-marked tank? Do this for a few nights, You should be able to predict 100% of the time which tank has the mark.
I've gone back and forth between a new tank and the old one before. There's definitely a noticable difference. Also, I'm just more tired and wake up more frequently until the tank is replaced, and I don't think that would be the case if it were just a matter of getting used to the pressure. I've been using CPAP for about 4 years now and I'm as used to it as I'll be.

I have already replaced the device. No change.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Resmed Airsense 10 humidifier seal

Post by palerider » Mon Nov 08, 2021 3:50 pm

BradleyBuck1984 wrote:
Mon Nov 08, 2021 2:50 pm
Maybe that's it. I haven't always washed the tank daily, but I typically remove the gasket for drying. I suppose the repeated removal could worsen the seal.
Sounds reasonable.
BradleyBuck1984 wrote:
Mon Nov 08, 2021 2:50 pm
I assumed the skin oils would degrade the silicone without washing.
Yes, they can. but all you need to do is wipe the oils off with a baby wipe, or an alcohol pad, or a dab of hand sanitizer on a kleenex (I used that last for years, removes the facial oils nicely.)

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Resmed Airsense 10 humidifier seal

Post by zonker » Mon Nov 08, 2021 4:26 pm

BradleyBuck1984 wrote:
Mon Nov 08, 2021 2:50 pm
Do you think the nasal pillows and parts which have direct contact with the face need washed daily? I assumed the skin oils would degrade the silicone without washing.
i'm with palerider on this. i wipe my mask with a baby wipe every morning.

i don't wash my mask except for maybe 2-3 times per year. and that's not on any schedule. i'm usually reminded to do it by someone like you posting about washing masks, so thanks for that. :lol:

i'm convinced that this is how i've gotten my p10 nasal pillow to last literally for years.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg