How long should hose remain heated?

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
CyberSpiff
Posts: 19
Joined: Tue Sep 21, 2021 1:47 pm

How long should hose remain heated?

Post by CyberSpiff » Wed Oct 13, 2021 10:48 am

Just got a heated hose from my DME and have to say after 7+ months of the non-heated variety I’m enjoying it. One thing I noticed is that it does not seem to maintain a constant temp all night as I can feel the air on the rare occasion I wake up not being the same temp as when I was falling asleep. I’m assuming that is how it works, but wanted to check with those who have been using one for a while.
Amateurs guess, then apologize afterwards.

Resmed Airsense 10 Autoset
Resmed F20 (but trying the P10 now)

User avatar
Pugsy
Posts: 59303
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: How long should hose remain heated?

Post by Pugsy » Wed Oct 13, 2021 11:33 am

Probably one of 2 things or maybe a combination of both.

1...these heaters cycle a bit....on/off for a short period of time not unlike the furnace or A/C of your house
2...our brains adjust to a new normal as the night goes on kike when the pressure we start the night with feels like a hurricane and it's a gentle breeze in the mornings....the warmer air seems markedly warmer at the beginning of the night but after a few hours it's the new normal and we just don't notice it as much.
Mask: Bleep no headgear mask.. https://bleepsleep.com/
Machine: AirCurve 10 VAuto

I may have to RISE but I refuse to SHINE.

User avatar
onward60
Posts: 136
Joined: Tue Sep 14, 2021 1:48 pm
Location: Midwest, USA

Re: How long should hose remain heated?

Post by onward60 » Wed Oct 13, 2021 12:44 pm

CyberSpiff wrote:
Wed Oct 13, 2021 10:48 am
Just got a heated hose from my DME and have to say after 7+ months of the non-heated variety I’m enjoying it. One thing I noticed is that it does not seem to maintain a constant temp all night as I can feel the air on the rare occasion I wake up not being the same temp as when I was falling asleep. I’m assuming that is how it works, but wanted to check with those who have been using one for a while.
Curious about what you like better. Is it the lack of rain out? The temperature? Both? Also, what hose did you choose? Thanks!
New to CPAP in Sept. 2021. N20 and MAD caused dental pain. Currently trying the Resmed Airfit F20 (medium). ResMed Air 10 Autoset for her machine.

My private messages might be turned off.

User avatar
CyberSpiff
Posts: 19
Joined: Tue Sep 21, 2021 1:47 pm

Re: How long should hose remain heated?

Post by CyberSpiff » Wed Oct 13, 2021 3:02 pm

onward60 wrote:
Wed Oct 13, 2021 12:44 pm
Curious about what you like better. Is it the lack of rain out? The temperature? Both? Also, what hose did you choose? Thanks!
I have a Resmed Airsense 10 Autoset so they sent me the ClimateLine heated hose. Just plugged it in and set everything to auto. I’ve only used it a few nights but the warm moist air is very soothing (once it gets up to temp). I have not previously had issues with rainout with the non-heated hose and no issues with this one so far.
Amateurs guess, then apologize afterwards.

Resmed Airsense 10 Autoset
Resmed F20 (but trying the P10 now)

User avatar
onward60
Posts: 136
Joined: Tue Sep 14, 2021 1:48 pm
Location: Midwest, USA

Re: How long should hose remain heated?

Post by onward60 » Wed Oct 13, 2021 6:13 pm

CyberSpiff wrote:
Wed Oct 13, 2021 3:02 pm
onward60 wrote:
Wed Oct 13, 2021 12:44 pm
Curious about what you like better. Is it the lack of rain out? The temperature? Both? Also, what hose did you choose? Thanks!
I have a Resmed Airsense 10 Autoset so they sent me the ClimateLine heated hose. Just plugged it in and set everything to auto. I’ve only used it a few nights but the warm moist air is very soothing (once it gets up to temp). I have not previously had issues with rainout with the non-heated hose and no issues with this one so far.
How is the smell? I have seen many complaints about the smell of the heated hose.
New to CPAP in Sept. 2021. N20 and MAD caused dental pain. Currently trying the Resmed Airfit F20 (medium). ResMed Air 10 Autoset for her machine.

My private messages might be turned off.

User avatar
CyberSpiff
Posts: 19
Joined: Tue Sep 21, 2021 1:47 pm

Re: How long should hose remain heated?

Post by CyberSpiff » Wed Oct 13, 2021 6:22 pm

onward60 wrote:
Wed Oct 13, 2021 6:13 pm
How is the smell? I have seen many complaints about the smell of the heated hose.
I have not noticed any smell so far. Based on what I’ve read in this forum I did wash it in cold water and dish soap, rinsed it well, and let it dry overnight before using.
Amateurs guess, then apologize afterwards.

Resmed Airsense 10 Autoset
Resmed F20 (but trying the P10 now)

User avatar
onward60
Posts: 136
Joined: Tue Sep 14, 2021 1:48 pm
Location: Midwest, USA

Re: How long should hose remain heated?

Post by onward60 » Wed Oct 13, 2021 6:45 pm

CyberSpiff wrote:
Wed Oct 13, 2021 6:22 pm
onward60 wrote:
Wed Oct 13, 2021 6:13 pm
How is the smell? I have seen many complaints about the smell of the heated hose.
I have not noticed any smell so far. Based on what I’ve read in this forum I did wash it in cold water and dish soap, rinsed it well, and let it dry overnight before using.
Thank you. That is very good news!
New to CPAP in Sept. 2021. N20 and MAD caused dental pain. Currently trying the Resmed Airfit F20 (medium). ResMed Air 10 Autoset for her machine.

My private messages might be turned off.

User avatar
CyberSpiff
Posts: 19
Joined: Tue Sep 21, 2021 1:47 pm

Re: How long should hose remain heated?

Post by CyberSpiff » Thu Oct 14, 2021 7:17 am

Pugsy wrote:
Wed Oct 13, 2021 11:33 am
Probably one of 2 things or maybe a combination of both.

1...these heaters cycle a bit....on/off for a short period of time not unlike the furnace or A/C of your house
2...our brains adjust to a new normal as the night goes on kike when the pressure we start the night with feels like a hurricane and it's a gentle breeze in the mornings....the warmer air seems markedly warmer at the beginning of the night but after a few hours it's the new normal and we just don't notice it as much.
Thank you for the suggestions. I didn’t consider the cycling of heat nor how I likely adjust to it. It does not seem to wake me up either way, so not concerned, just curious.
Amateurs guess, then apologize afterwards.

Resmed Airsense 10 Autoset
Resmed F20 (but trying the P10 now)