pressure difference with new mask?
-
- Posts: 28
- Joined: Fri Oct 08, 2021 3:13 am
Re: pressure difference with new mask?
[/quote]
Maybe they need to be adjusted now. Are you reviewing your daily AHI and leak level that shows on the machine display?
[/quote]
I am reviewing them daily. I am getting good numbers generally…AHI between 2 and 4, leaks usually less than 12L/min. I am going to install OSCAR today and hopefully get a lot more data than my airsense 11 gives me….I only get time used, leaks, AHI, Total AI and CI. No other report settings with this machine…not even in the clinical menu.
What do you suggest in terms of adjustments? I was thinking of starting off with changing the AutoSet to Standard versus Soft.
Maybe they need to be adjusted now. Are you reviewing your daily AHI and leak level that shows on the machine display?
[/quote]
I am reviewing them daily. I am getting good numbers generally…AHI between 2 and 4, leaks usually less than 12L/min. I am going to install OSCAR today and hopefully get a lot more data than my airsense 11 gives me….I only get time used, leaks, AHI, Total AI and CI. No other report settings with this machine…not even in the clinical menu.
What do you suggest in terms of adjustments? I was thinking of starting off with changing the AutoSet to Standard versus Soft.
Mask: Airfit F30i
Machine: Airsense 11
New York
Machine: Airsense 11
New York
Re: pressure difference with new mask?
Post Oscar, then we can see what's what.
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
- ChicagoGranny
- Posts: 14469
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: pressure difference with new mask?
If you mean changing this setting when you switch to a nasal mask, I wouldn't do it. Standard practice is to change only one variable at a time. If two are changed and the results are different, you likely won't know which of the two variables caused the different results.Redstormsju wrote: ↑Mon Oct 18, 2021 2:47 amI was thinking of starting off with changing the AutoSet to Standard versus Soft.
I would leave the machine settings the same when starting the new mask. The settings can be changed later if they prove to be suboptimal.
BTW, it is sometimes said that a higher pressure is needed with a FFM versus a nasal mask. Actually, not a change in mask, but mouth-breathing may result in higher pressure needs. In humans, when the mouth opens, the mandible moves backward slightly. This tends to narrow the airway.
These are good results for a newbie (and many oldies ). I wouldn't change any settings at this point. OSCAR may tell us more.Redstormsju wrote: ↑Mon Oct 18, 2021 2:47 amI am getting good numbers generally…AHI between 2 and 4, leaks usually less than 12L/min.
-
- Posts: 28
- Joined: Fri Oct 08, 2021 3:13 am
Re: pressure difference with new mask?
These are good results for a newbie (and many oldies ). I wouldn't change any settings at this point. OSCAR may tell us more.Redstormsju wrote: ↑Mon Oct 18, 2021 2:47 amI am getting good numbers generally…AHI between 2 and 4, leaks usually less than 12L/min.
[/quote]
Thank you so much.
Mask: Airfit F30i
Machine: Airsense 11
New York
Machine: Airsense 11
New York
-
- Posts: 28
- Joined: Fri Oct 08, 2021 3:13 am
Re: pressure difference with new mask?
What graphs should I have up on OSCAR for posting on this forum? I see there are many that I can scroll through on OSCAR.These are good results for a newbie (and many oldies ). I wouldn't change any settings at this point. OSCAR may tell us more.
Mask: Airfit F30i
Machine: Airsense 11
New York
Machine: Airsense 11
New York
Re: pressure difference with new mask?
Take a look here. Start with the basics.Redstormsju wrote: ↑Tue Oct 19, 2021 6:39 pmWhat graphs should I have up on OSCAR for posting on this forum?
viewtopic/t158560/How-to-post-images-for-review.html
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/ |
I may have to RISE but I refuse to SHINE.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.
Re: pressure difference with new mask?
forum friend grumpy here has this to say-Redstormsju wrote: ↑Tue Oct 19, 2021 6:39 pmWhat graphs should I have up on OSCAR for posting on this forum? I see there are many that I can scroll through on OSCAR.These are good results for a newbie (and many oldies ). I wouldn't change any settings at this point. OSCAR may tell us more.
https://www.cpaptalk.com/viewtopic.php? ... 9#p1390059
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 28
- Joined: Fri Oct 08, 2021 3:13 am
Re: pressure difference with new mask?
Hello all, these screenshots are from last nights sleep. I see the numbers are good, but I am not quite confident in how to read the graphs. I’ve been reading from the help page but still am not quite sure how to interpret the graphs. Any insight is appreciated.
- Attachments
-
- A663A53D-18ED-465D-8A13-ACEBE8D39C25.png (229.62 KiB) Viewed 692 times
-
- CD70FF8B-B3FA-4607-9F57-77BC2E12731F.png (237.98 KiB) Viewed 692 times
-
- 8EB56541-FCD5-43D4-B915-687693CF4D6B.png (228.35 KiB) Viewed 692 times
Mask: Airfit F30i
Machine: Airsense 11
New York
Machine: Airsense 11
New York
- ChicagoGranny
- Posts: 14469
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: pressure difference with new mask?
Can you be more specific about what you don't understand?Redstormsju wrote: ↑Sat Oct 23, 2021 8:36 ambut still am not quite sure how to interpret the graphs. Any insight is appreciated.
- ChicagoGranny
- Posts: 14469
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: pressure difference with new mask?
These are not in the recommended format.
-
- Posts: 28
- Joined: Fri Oct 08, 2021 3:13 am
Re: pressure difference with new mask?
I’m not sure how to read/interpret the AHI graph, Minute vent, Insp. Time, Exp. Time.Can you be more specific about what you don't understand?
And the pressure graph, I assume inhale and exhale pressures?
Mask: Airfit F30i
Machine: Airsense 11
New York
Machine: Airsense 11
New York
-
- Posts: 28
- Joined: Fri Oct 08, 2021 3:13 am
Re: pressure difference with new mask?
What is the recommended format?
These are not in the recommended format.
Mask: Airfit F30i
Machine: Airsense 11
New York
Machine: Airsense 11
New York
Re: pressure difference with new mask?
The AHI graph is useless,Redstormsju wrote: ↑Sat Oct 23, 2021 9:01 amI’m not sure how to read/interpret the AHI graph, Minute vent, Insp. Time, Exp. Time.Can you be more specific about what you don't understand?
MV is your average total respiration volume over a minute.
Exp time, how long your exhalation is
Ignore all of them.
Yes.Redstormsju wrote: ↑Sat Oct 23, 2021 9:01 amAnd the pressure graph, I assume inhale and exhale pressures?
As explained in the "READ THIS BEFORE POSTING" post at the top of the forum, we only need six graphs 99% of the time.
flags
flow rate
pressure (not mask pressure)
leaks
snores
flow limitation.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: pressure difference with new mask?
https://www.cpaptalk.com/wiki/index.php/Oscar:organizeRedstormsju wrote: ↑Sat Oct 23, 2021 9:04 amWhat is the recommended format?
These are not in the recommended format.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
-
- Posts: 28
- Joined: Fri Oct 08, 2021 3:13 am
Re: pressure difference with new mask?
Not sure how my attachments are not the recommended format. According what I read on the link provided, the format is png image file and that is what my images are. I took screenshots directly from the Take Screenshot option on Oscar, I saved the image to my desktop and just changed the file name to the date of the data. I then used the attachment option in the forum page to attach the pictures to my post. The only thing I did notice was that when they attached to the post, they attached as one large image versus the three individual images I attached while creating g the post.
Mask: Airfit F30i
Machine: Airsense 11
New York
Machine: Airsense 11
New York