Do I have mouth leak? (w/ detailed treatment log)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Beat OSA
Posts: 29
Joined: Mon Feb 15, 2021 7:51 pm
Location: Asia

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by Beat OSA » Fri Oct 01, 2021 5:56 pm

Pugsy wrote:
Thu Sep 30, 2021 6:39 pm
Chipmunk cheeks can happen even at lower pressures...actually at any pressure but you could certainly try reducing the pressure and see what happens.

Bear in mind that when you reduce the pressure then the pressure may not be able to hold the airway open and prevent airway collapses so you may see an increase in the AHI should that happen.

Full face masks come with their own leak baggage...up to you if you want to go down that road. They don't necessarily stop chipmunk cheeks either but will allow you to simple release the pressure by opening your mouth without losing therapy pressure. Might also still wake you up though which of course is unwanted because it messes with sleep quality.

Not my job to make decisions for you.
That's your job.
My job is only to explain the options available and pros and cons to those options.

I do almost always go for the cheapest and easiest experiment to be trying though. Worst thing that could happen is you learn it wasn't a good option and you move on to a different experiment. I always learned something though from all my experiments. Sometimes all I learned was what a stupid idea I tried. :lol:
When I first started cpap therapy I was doing some mouth breathing out of habit....so I opted to tape my mouth because it was cheap easy experiment to try. Now I have never had a problem with chipmunk cheeks and I routine saw pressures in the mid teens (still do sometimes). I quit taping after about 3 or 4 months....broke the mouth opening habit.
Every now and then I might see evidence of very brief mouth opening....I ignore it since it doesn't wake me up and it never lasts long enough deep in large leak territory for it to really have a negative impact on my therapy.

Getting to sleep and staying asleep and getting enough hours of sleep has always been my primary goal because without sleep none of the good numbers mean much.
Thank you Pugsy.
I tried chin strap last night. It's a home-made one made from one of my ankle straps. I decided not to use tape this time.
I do not recall having chipmunk cheek last night. But I do recall occasionally I was waked up and had to open my lips to let the air out.
Not surprisingly the OSCAR showed major leaks last night.
The homemade strap did not work, I ordered another one online and will give it a try.

Pic OSCAR (Homemade strap. No tape)
截屏2021-10-02 上午7.56.32.png
截屏2021-10-02 上午7.56.32.png (529.19 KiB) Viewed 913 times
Machine:
Resmed Airsense 10 Autoset For Her Plus (Since 2021/09/24 )
Philips Dreamstation (Retired) (In-service 2021/2/1 - 2021/9/23)
Mask:
Resmed P10

User avatar
jimbud
Posts: 1454
Joined: Sat Dec 01, 2018 8:19 pm

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by jimbud » Fri Oct 01, 2021 6:58 pm

Beat OSA wrote:
Fri Oct 01, 2021 5:56 pm
hank you Pugsy.
I tried chin strap last night. It's a home-made one made from one of my ankle straps. I decided not to use tape this time.
I do not recall having chipmunk cheek last night. But I do recall occasionally I was waked up and had to open my lips to let the air out.
Not surprisingly the OSCAR showed major leaks last night.
The homemade strap did not work, I ordered another one online and will give it a try.
You are a mouth breather.

Just a chin strap will not work. Even with your mouth closed you will still breath through your teeth.
Tape your mouth or get a full face mask while you try to train yourself to eliminate mouth breathing.

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

User avatar
Beat OSA
Posts: 29
Joined: Mon Feb 15, 2021 7:51 pm
Location: Asia

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by Beat OSA » Sat Oct 02, 2021 8:00 pm

jimbud wrote:
Fri Oct 01, 2021 6:58 pm
Beat OSA wrote:
Fri Oct 01, 2021 5:56 pm
hank you Pugsy.
I tried chin strap last night. It's a home-made one made from one of my ankle straps. I decided not to use tape this time.
I do not recall having chipmunk cheek last night. But I do recall occasionally I was waked up and had to open my lips to let the air out.
Not surprisingly the OSCAR showed major leaks last night.
The homemade strap did not work, I ordered another one online and will give it a try.
You are a mouth breather.

Just a chin strap will not work. Even with your mouth closed you will still breath through your teeth.
Tape your mouth or get a full face mask while you try to train yourself to eliminate mouth breathing.

JPB
Thank you Jimbud.

Last night I tried to sleep with strap + tape + CPAP w/ P10 mask.

Good news: Best night probably in a month! I can only recall one time I waked up at midnight found my cheeks are puffing up under the strap. Perhaps when I get my new strap it will get better.

Bad news: I can not think of sleeping w/ strap and tape for the rest of my life. It's not compatible with my job I occasionally have to do the night shift, I'm a medical doctor. But I used to think I can not live with CPAP so who knows. Eventually, I need to train to keep the back of my throat shut while sleeping like I used to. Or perhaps try a full face mask if that works as good as strap and tape. I still suspect it has sth to do with my weight loss. There is no mouth leak when I was a bigger guy. Hopefully, my body will adjust to the weight change gradually.

Pic OSCAR from last night (Strap + tape + CPAP w/ P10 mask)
截屏2021-10-03 上午9.59.49.png
截屏2021-10-03 上午9.59.49.png (544.45 KiB) Viewed 868 times
Machine:
Resmed Airsense 10 Autoset For Her Plus (Since 2021/09/24 )
Philips Dreamstation (Retired) (In-service 2021/2/1 - 2021/9/23)
Mask:
Resmed P10

User avatar
Julie
Posts: 19910
Joined: Tue Feb 28, 2006 12:58 pm

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by Julie » Sun Oct 03, 2021 6:25 am

I don't understand the logic of keeping the back of your throat closed - completely counterproductive to the whole point of Cpap... or did I miss something along the way?
Last edited by Julie on Sun Oct 03, 2021 1:59 pm, edited 1 time in total.

User avatar
Pugsy
Posts: 64020
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by Pugsy » Sun Oct 03, 2021 6:51 am

Julie wrote:
Sun Oct 03, 2021 6:25 am
I don't understand the logic of keeping the back of your throat closed - completely 180 counterproductive to the whole point of Cpap... or did I miss something along the way?
I think referring to closing the oral cavity/back of mouth so that the air pressure doesn't sneak into the mouth causing chipmunk cheeks.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Miss Emerita
Posts: 3444
Joined: Sun Nov 04, 2018 8:07 pm

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by Miss Emerita » Sun Oct 03, 2021 12:07 pm

Congratulations on having a better night! Don't try to think too far into the future; with time, all kinds of routines can become, well, routine. Or not; you'll see. Meanwhile, keep working on training your tongue. Many people here have had good luck with that.
Oscar software is available at https://www.sleepfiles.com/OSCAR/

User avatar
Beat OSA
Posts: 29
Joined: Mon Feb 15, 2021 7:51 pm
Location: Asia

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by Beat OSA » Sun Oct 03, 2021 10:23 pm

Julie wrote:
Sun Oct 03, 2021 6:25 am
I don't understand the logic of keeping the back of your throat closed - completely counterproductive to the whole point of Cpap... or did I miss something along the way?
Sorry for the wording. I meant the back of my oral cavity. I have an issue with the chipmunk cheek when using CPAP.
Machine:
Resmed Airsense 10 Autoset For Her Plus (Since 2021/09/24 )
Philips Dreamstation (Retired) (In-service 2021/2/1 - 2021/9/23)
Mask:
Resmed P10

User avatar
Beat OSA
Posts: 29
Joined: Mon Feb 15, 2021 7:51 pm
Location: Asia

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by Beat OSA » Sun Oct 03, 2021 10:29 pm

Miss Emerita wrote:
Sun Oct 03, 2021 12:07 pm
Congratulations on having a better night! Don't try to think too far into the future; with time, all kinds of routines can become, well, routine. Or not; you'll see. Meanwhile, keep working on training your tongue. Many people here have had good luck with that.
Thank you Miss Emerita. Last night I still waked up several times. I suppose the strap I'm using now could only partly prevent the chipmunk cheek at best. And even though there is no leak on OSCAR, I could still be waked up by either the ill-fitted strap or the sensation of my cheek puffing up. I'm still waiting for my new chin strap and hopefully, it will do a better job. But I think the ultimate solution would be training my tongue well enough to prevent the air from going into my oral cavity, with or without a strap and tape.
Machine:
Resmed Airsense 10 Autoset For Her Plus (Since 2021/09/24 )
Philips Dreamstation (Retired) (In-service 2021/2/1 - 2021/9/23)
Mask:
Resmed P10

User avatar
Beat OSA
Posts: 29
Joined: Mon Feb 15, 2021 7:51 pm
Location: Asia

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by Beat OSA » Thu Oct 07, 2021 7:18 pm

It's so weird.
We had a party last night. And I end up totally drunk. And for the first time in a year, I forgot to use my CPAP machine.
I still had the problem of early waking up at around 5:30 am, only this time I felt very refreshed. (Except for the hangover of course). And I felt great throughout the day.

It's so weird. My osa is cured?
Machine:
Resmed Airsense 10 Autoset For Her Plus (Since 2021/09/24 )
Philips Dreamstation (Retired) (In-service 2021/2/1 - 2021/9/23)
Mask:
Resmed P10

User avatar
jimbud
Posts: 1454
Joined: Sat Dec 01, 2018 8:19 pm

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by jimbud » Thu Oct 07, 2021 7:51 pm

Beat OSA wrote:
Thu Oct 07, 2021 7:18 pm

I end up totally drunk.
I felt great throughout the day. (Except for the hangover of course).

My osa is cured?
Beat OSA wrote:
Sat Oct 02, 2021 8:00 pm
I'm a medical doctor.
Not much of one, Huh.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.

GrumpyHere
Posts: 508
Joined: Sat Feb 13, 2021 2:40 am

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by GrumpyHere » Thu Oct 07, 2021 9:05 pm

Beat OSA wrote:
Thu Oct 07, 2021 7:18 pm
It's so weird. My osa is cured?
Every night's sleep is different.

Occasionally, I go with only 2-3 hours of sleep and feel ok the next day.
The short sleep session is just a nap.

Can't blame you for the wishful thinking though.
CPAP is a PITA.
Don't give up on trying to make it work just because of a lucky fluke.
ResMed Lumis Tx - VAuto mode

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by zonker » Thu Oct 07, 2021 10:04 pm

GrumpyHere wrote:
Thu Oct 07, 2021 9:05 pm

CPAP is a PITA.
if it's a PITA, you have it hooked up wrong.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by zonker » Thu Oct 07, 2021 10:05 pm

Beat OSA wrote:
Thu Oct 07, 2021 7:18 pm


It's so weird. My osa is cured?
yup. you've found the answer.just get drunk every night.
:lol: :lol: :lol:
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

GrumpyHere
Posts: 508
Joined: Sat Feb 13, 2021 2:40 am

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by GrumpyHere » Thu Oct 07, 2021 11:47 pm

zonker wrote:
Thu Oct 07, 2021 10:04 pm

if it's a PITA, you have it hooked up wrong.
:shock: :shock: :shock:

Given the torrent of farts on a bad day, it does feel like it's been hooked up that way :? :? :?
ResMed Lumis Tx - VAuto mode

User avatar
jimbud
Posts: 1454
Joined: Sat Dec 01, 2018 8:19 pm

Re: Do I have mouth leak? (w/ detailed treatment log)

Post by jimbud » Fri Oct 08, 2021 8:40 am

GrumpyHere wrote:
Thu Oct 07, 2021 9:05 pm
CPAP is a PITA.
You know what was a PITA for me?
I'll tell you anyway.

My life living with Apnea before therapy. Knowing something was wrong, but not knowing what.
No doctor recognizing my symptoms and steering me toward therapy. Until my health took a dive and I ended up in the hospital. My cardiologist sent me to get checked out with a home monitor. Severe OSA.

The therapy is a God Send, not a PITA. For me.

YMMV. :D

JPB

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Sleep on a Buckwheat Hull Pillow.