CPAP BATTERIES
-
- Posts: 1
- Joined: Thu Sep 23, 2021 7:41 am
CPAP BATTERIES
I"m going to be traveling to isolated areas in Africa that will not have electricity at night. I'm currently using a Philips RESPIRONICS REMstar Pro. Question: Is their a machine that is better to use with batteries and are there batteries that are easy to charge and provide a reasonable useable length of time?
-
- Posts: 4
- Joined: Mon Sep 20, 2021 9:05 pm
Re: CPAP BATTERIES
I had a CPAP machine go crazy and then kaput that was on a battery back up. Tech support said using a UPS with their machine was a no no. I think it is because most inverters produce almost square waves instead of a pure sine wave. The machine was a Philips Respironics REMstar Auto. That said, my replacement machine which is also a Philips Respironics REMstar Auto has been on a battery backup and I use it on batteries while camping with out any further issues.
For battery I use a Greenworks 288Wh 80V battery. Greenworks also makes a power inverter to go with that battery. It will last me at least two nights. It's big, its heavy, and the charger is not small either. You might want to look into one of the smaller medical batteries on a place like cpap.com and a solar setup to charge it. If you are going to be flying, there are restrictions on how large of a battery you can bring on board. Keep an eye out for that also.
For battery I use a Greenworks 288Wh 80V battery. Greenworks also makes a power inverter to go with that battery. It will last me at least two nights. It's big, its heavy, and the charger is not small either. You might want to look into one of the smaller medical batteries on a place like cpap.com and a solar setup to charge it. If you are going to be flying, there are restrictions on how large of a battery you can bring on board. Keep an eye out for that also.
Re: CPAP BATTERIES
Modern (made in the last 11 years) CPAPsdo not care about modified sine wave power, they have the same kind of power supplies that the computers you use the UPSs on have, the first thing they do is turn that modified square wave into DC anyway.RocketBeaglesSam wrote: ↑Thu Sep 23, 2021 12:22 pmI had a CPAP machine go crazy and then kaput that was on a battery back up. Tech support said using a UPS with their machine was a no no. I think it is because most inverters produce almost square waves instead of a pure sine wave.
That says nothing about what machine is is, since Respironics slapped "remstar" on everything up till the dreamstations came along.RocketBeaglesSam wrote: ↑Thu Sep 23, 2021 12:22 pmThe machine was a Philips Respironics REMstar Auto. That said, my replacement machine which is also a Philips Respironics REMstar Auto has been on a battery backup and I use it on batteries while camping with out any further issues.
Using any sort of inverter will drastically decrease the runtime, because they're not very efficient.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: CPAP BATTERIES
We'd need to know which "remstar pro" machine you're talking about.RichSleepWell!! wrote: ↑Thu Sep 23, 2021 8:09 amI"m going to be traveling to isolated areas in Africa that will not have electricity at night. I'm currently using a Philips RESPIRONICS REMstar Pro. Question: Is their a machine that is better to use with batteries and are there batteries that are easy to charge and provide a reasonable useable length of time?
look on the bottom for a model or REF number and tell us what it is.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
- lcunningham8332
- Posts: 2
- Joined: Sun Oct 10, 2021 4:37 pm
Re: CPAP BATTERIES
I recently purchased the F&P Sleepstyle Auto. This is the first cpap ive had that i couldnt run on a regular cpap battery. I have NO IDEA what i need to buy for a battery backup as we lose the power often in the winter, in my area. Can someone tell me exactly what I need to look for? Name, link etc.. i have absolutly no clue, and am being managed by my Family Practice Dr and not a sleep company.
Thanks for you your help~!
Thanks for you your help~!
- ChicagoGranny
- Posts: 14471
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: CPAP BATTERIES
Here is the F&P battery guide - https://static1.squarespace.com/static/ ... +Guide.pdflcunningham8332 wrote: ↑Sun Oct 10, 2021 5:11 pmI recently purchased the F&P Sleepstyle Auto. This is the first cpap ive had that i couldnt run on a regular cpap battery. I have NO IDEA what i need to buy for a battery backup as we lose the power often in the winter, in my area. Can someone tell me exactly what I need to look for? Name, link etc.. i have absolutly no clue, and am being managed by my Family Practice Dr and not a sleep company.
Thanks for you your help~!
You want a mobility scooter battery, not a marine battery.
- lcunningham8332
- Posts: 2
- Joined: Sun Oct 10, 2021 4:37 pm
Re: CPAP BATTERIES
Thanks, i have seen that info before, but I dont know how to decipher it. I dont understand what that means.ChicagoGranny wrote: ↑Sun Oct 10, 2021 5:22 pmHere is the F&P battery guide - https://static1.squarespace.com/static/ ... +Guide.pdflcunningham8332 wrote: ↑Sun Oct 10, 2021 5:11 pmI recently purchased the F&P Sleepstyle Auto. This is the first cpap ive had that i couldnt run on a regular cpap battery. I have NO IDEA what i need to buy for a battery backup as we lose the power often in the winter, in my area. Can someone tell me exactly what I need to look for? Name, link etc.. i have absolutly no clue, and am being managed by my Family Practice Dr and not a sleep company.
Thanks for you your help~!
You want a mobility scooter battery, not a marine battery.
Re: CPAP BATTERIES
What part is the problem?
_________________
Mask: Ultra Mirage™ Full Face CPAP Mask with Headgear |
Humidifier: IntelliPAP Integrated Heated Humidifier |
- ChicagoGranny
- Posts: 14471
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: CPAP BATTERIES
Please specify the model number. There are at least three different models. In fact, you should put the machine name and model number and mask model in the signature field of your profile. It will then show with each of your posts and comments.
How many nights must the battery last? Will you have an opportunity to charge it during the day?lcunningham8332 wrote: ↑Sun Oct 10, 2021 5:40 pmi have seen that info before, but I dont know how to decipher it. I dont understand what that means.
Is it correct that you will only be using it at home where size/weight are not as big a factor as when using a battery for certain kinds of travel?
Re: CPAP BATTERIES
I just bought a GOLABS 60W portable power station
https://www.amazon.ca/gp/product/B08NSY ... UTF8&psc=1
I used it last night with my Resmed AirMini and it worked very well. I used less than 30% of the battery power last night, so I should be able to get 3 nights sleep on one battery charge, not bad.
https://www.amazon.ca/gp/product/B08NSY ... UTF8&psc=1
I used it last night with my Resmed AirMini and it worked very well. I used less than 30% of the battery power last night, so I should be able to get 3 nights sleep on one battery charge, not bad.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ N20 Nasal CPAP Mask with Headgear |
-
- Posts: 508
- Joined: Sat Feb 13, 2021 2:40 am
Re: CPAP BATTERIES
For battery longevity, it’s better not to cycle that deeply.
ResMed Lumis Tx - VAuto mode
Re: CPAP BATTERIES
That's a LiFePO4 lithium battery. It's probably more tolerant of deep discharge. It would probably still outlast most of us if occasionally discharged to 0%
Edit: also holds a charge on the shelf better than other types. A nice feature.
Edit: also holds a charge on the shelf better than other types. A nice feature.
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Oracle HC452 Oral CPAP Mask |
Re: CPAP BATTERIES
so, on amazon.com, this unit is $199.00 and then can deduct $70 with coupon.
(prices american, of course.)
so, folk who know batteries, is this something a total battery ignoramus like me could/should buy?
https://www.amazon.com/GOLABS-Portable- ... 247&sr=8-5
(prices american, of course.)
so, folk who know batteries, is this something a total battery ignoramus like me could/should buy?
https://www.amazon.com/GOLABS-Portable- ... 247&sr=8-5
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
- loggerhead12
- Posts: 560
- Joined: Fri Sep 01, 2017 4:00 pm
Re: CPAP BATTERIES
I use about 80 watt hours per night, so It wouldn't last me three days. Seems easy enough to use, though.zonker wrote: ↑Fri Oct 15, 2021 3:47 pmso, on amazon.com, this unit is $199.00 and then can deduct $70 with coupon.
(prices american, of course.)
so, folk who know batteries, is this something a total battery ignoramus like me could/should buy?
https://www.amazon.com/GOLABS-Portable- ... 247&sr=8-5
What kind of cable plugs into those 12v outputs? You'd need something that would plug in there and have a cigarette/accessory plug on the other end so the Resmed converter would plug into it.
Re: CPAP BATTERIES
oh. i'm not looking for any kind of camping with the unit. just a backup should the power go out here.loggerhead12 wrote: ↑Fri Oct 15, 2021 9:30 pm
I use about 80 watt hours per night, so It wouldn't last me three days. Seems easy enough to use, though.
What kind of cable plugs into those 12v outputs? You'd need something that would plug in there and have a cigarette/accessory plug on the other end so the Resmed converter would plug into it.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg