Post by BlueDragon » Sat Aug 28, 2021 12:06 pm

OSCAR v 1.3.0-Beta2 is here and we need beta testers especially for newly supported machines.

If you have one of the newly supported machines or if you would just like to help beta test the next version of OSCAR, we would like to hear from you.

All beta testing files and support are managed on an OSCAR Beta Test forum on Apneaboard.com. Just join apneaboard and send a PM to Gideon, tell him about your beta test interest, and ask to be added to the beta test forum. We ask that all beta test reports be posted on that forum as the entire team watches it regularly. Reports posted here might get missed. Please PM me if I miss something here.

This version of OSCAR contains support for the following new devices.
. [new] Add preliminary support for ResMed AirSense 11 CPAP machines.
. [new] Add support for Fisher & Paykel SleepStyle CPAP machines.
. [new] Add support for DeVilbiss BLUE (DV6x) CPAP machines.

and in the near future the new Philips Respironics DS2 (we are actively working on it).

Your task is to exercise OSCAR and identify issues that we have missed. Please post what you find.
Be sure to state what version of OSCAR you are using as Beta and pre-release versions can change rapidly.

We do take all issues seriously so as much as possible. To help us,
. Say What you tested
. What went wrong
. How to repeat the results (if you can)
. Post Screenshots for clarity

As always, feel free to post therapy questions in the forum but do state what version you are using.

Changes and fixes in OSCAR v1.3.0
Portions of OSCAR are © 2019-2021 by The OSCAR Team
. [new] Add preliminary support for ResMed AirSense 11 CPAP machines.
. [new] Add support for Fisher & Paykel SleepStyle CPAP machines.
. [new] Add support for DeVilbiss BLUE (DV6x) CPAP machines.
. [new] Additional Philips Respironics devices tested and fully supported:
-- DreamStation Go Auto (500G120)
-- DreamStation Auto CPAP with A-Flex (500X140)
-- DreamStation Auto BiPAP (700X130)
-- DreamStation BiPAP AVAPS 30 (1130X200)

. [new] Add support for DreamStation Go humidifier Target Time setting.
. [new] Detect DreamStation 2 cards, which are still unsupported.
. [new] Add Bulgarian translation; update other languages.
. [new] Improve Somnopose import options.
. [new] Purge Current Selected Day allows purge of each machine type separately
. [new] Multi-file import for non-CPAP loaders (Somnopose, Viatom, Zeo, Dreem)
. [new] Weight, BMI and Zombie history appear in statistics
. [new] Add Turkish signal names to RedMed loader.
. [new] Add Traditional Chinese to languages available.
. [fix] Correct calculation of average leak rate on Welcome page.
. [fix] Correct installation of non-English Release Notes on Windows.
. [fix] About/Credits page now offers Google translations to other languages.
. [fix] User first and last name in Profile window may now have UTF-8 (non-ASCII) characters.
. [fix] Fix crash and other problems when disabling an oximeter session on Daily page when a bookmark was present.
. [fix] Fix rare problem of OSCAR crashing with unusual Journal file.
. [fix] ResMed loader no longer rejects data from an earlier timezone or DST.
. [fix] Newly entered notes no longer lost when importing new day or purging oximetry data.
. [fix] Purge currently selected day no longer deletes bookmarks for that day.
. [fix] Remove warning from Chromebook when importing from previously used local folder.
. [fix] Update link to Contec drivers.
. [fix] Fix display problems for short duration events.
. [fix] Statistics headings will now be 99.5% or Max, depending on machine type and preference settings.
. [fix] Mark exported Journal backup file as UTF-8.
. [fix] Improve error message when unable to access OSCAR database.
. [fix] Stop skipping the first 40 seconds of ResMed low-rate pressure data.
. [fix] Correct Total Time and AHI in CSV Export when non-CPAP devices are used.
. [fix] Fix value display and bookmark behavior with clock drift.
. [fix] Ignore old sessions should not impact existing data.
. [fix] Fix occasional misordering of indexes on the Daily page.
. [fix] Add Unclassified Apneas to the Statistics page.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
Last edited by BlueDragon on Sat Aug 28, 2021 1:04 pm, edited 1 time in total.
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
Pugsy
Posts: 64012
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by Pugsy » Sat Aug 28, 2021 1:01 pm

Made this a sticky for a while. Exact timeline is fluid for now.
I won't be participating though. So don't go asking me questions about it. :lol:

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
kbh209
Posts: 339
Joined: Wed Dec 30, 2015 9:12 pm

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by kbh209 » Tue Aug 31, 2021 12:59 am

@BlueDragon: I thought the DS2 was a brick due to Philips locking down the clinician menu? Has there been a change?
Phillips Dreamstation 2 Auto CPAP with Humidifier **

User avatar
BlueDragon
Posts: 541
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by BlueDragon » Tue Aug 31, 2021 1:50 am

"Brick" is too strong a term for the DS2. If the clinician menu is locked, it can be unlocked. If you are referring to the data on the SD card being encrypted, the OSCAR Team is still confident that the encryption problem can be solved.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

AnnaGray
Posts: 1
Joined: Mon Sep 13, 2021 11:04 am

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by AnnaGray » Mon Sep 13, 2021 11:07 am

Hm, I look at this and I don't even know what to say)))

vilnius
Posts: 19
Joined: Mon Sep 20, 2021 5:01 pm

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by vilnius » Mon Sep 20, 2021 8:57 pm

This is not about the Beta version but I was told to post here after asking on a new thread.

I bought a used Fisher & Paykel Icon+ Auto and when I plug in the SmartStick (flash USB drive) into my laptop and import the data to OSCAR, it shows the current year as 2042 and the data is wrong (0 AHIs, lower leak rates than display, etc. Only correct data is the pressure changes.) There is some data in there from 2012, which I think is when the machine was last used by the previous owner, and it all seems correct. What's going on and how do I fix it?

Here is a screenshot

Image

User avatar
BlueDragon
Posts: 541
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by BlueDragon » Mon Sep 20, 2021 9:08 pm

I have not looked at Icon++ processing, so I'll need to look at your InfoUSB card contents. Can you make a zip of the card contents and upload it to my Dropbox folder? Please include your forum name in the zip file name so we know who the file is from.

Do you have access to F&P's InfoSmart Web site? If you do, what does the date look like there? I wonder if your machine has lost track of the year? Can you set the date on your machine, and is it correct?

I've noticed that with the SleepStyle machine there is no apparent way to change the date (only the time). The machine I have been using for testing is 34 days behind the current day, and I've found no way to reset that date.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

vilnius
Posts: 19
Joined: Mon Sep 20, 2021 5:01 pm

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by vilnius » Mon Sep 20, 2021 9:30 pm

BlueDragon wrote:
Mon Sep 20, 2021 9:08 pm
I have not looked at Icon++ processing, so I'll need to look at your InfoUSB card contents. Can you make a zip of the card contents and upload it to my Dropbox folder? Please include your forum name in the zip file name so we know who the file is from.

Do you have access to F&P's InfoSmart Web site? If you do, what does the date look like there? I wonder if your machine has lost track of the year? Can you set the date on your machine, and is it correct?

I've noticed that with the SleepStyle machine there is no apparent way to change the date (only the time). The machine I have been using for testing is 34 days behind the current day, and I've found no way to reset that date.
Thanks for the quick response. I just uploaded a zip of my card's contents to your Dropbox.

I don't have access to the F&P's InfoSmart Web site. I can only set the time on my machine. I've looked in the manual and couldn't find anything about changing the date... I would be fine with a date error but why is it not recording AHIs and other data either?

User avatar
BlueDragon
Posts: 541
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by BlueDragon » Mon Sep 20, 2021 9:49 pm

I took a quick look at your data and I'm not going to have an immediate response for you. It looks like I may have to do some reverse engineering on your data to figure out what is going on.

My priority right now has to be fixing some issues on the SleepStyle loader for the beta testing cycle we are in.

When I try to read your data into either OSCAR 1.2.0 or the 1.3.0 beta, OSCAR just hangs. I don't know why that is happening, and it doesn't seem to be the same thing happening for you, which is strange.

Are you sure you have sent me the entire contents of the card? I noticed that the file dates in the FPHCARE\ICON\17089316 directory are all 1/12/2007. The higher level directories are all 9/20/2021. So it seems like there's no new data? Or maybe it's been tacked on to the end of existing files. If so, I'll have to do some digging into the files which will take some time.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

vilnius
Posts: 19
Joined: Mon Sep 20, 2021 5:01 pm

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by vilnius » Mon Sep 20, 2021 10:03 pm

BlueDragon wrote:
Mon Sep 20, 2021 9:49 pm
I took a quick look at your data and I'm not going to have an immediate response for you. It looks like I may have to do some reverse engineering on your data to figure out what is going on.

My priority right now has to be fixing some issues on the SleepStyle loader for the beta testing cycle we are in.

When I try to read your data into either OSCAR 1.2.0 or the 1.3.0 beta, OSCAR just hangs. I don't know why that is happening, and it doesn't seem to be the same thing happening for you, which is strange.

Are you sure you have sent me the entire contents of the card? I noticed that the file dates in the FPHCARE\ICON\17089316 directory are all 1/12/2007. The higher level directories are all 9/20/2021. So it seems like there's no new data? Or maybe it's been tacked on to the end of existing files. If so, I'll have to do some digging into the files which will take some time.
When I try to import the data into Oscar 1.2 it takes a while for me as well and if I click on it the app gives the Windows Not Responding message.

I think I sent you everything. I clicked on Help > Troubleshooting > Create zip of CPAP card data. I just tried it again and the zip is of the same size as the one I uploaded (19.8 MB).

I am not sure about the dates. Only data I can see in Oscar is from 2011, 2012 and 2042.

Thanks for looking. Take your time.

PS: Here are some more screenshots https://imgur.com/a/sDBr9Z2

User avatar
BlueDragon
Posts: 541
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by BlueDragon » Mon Sep 20, 2021 10:08 pm

It appears the date in the machine is bad. Here is a list of sessions found. Notice how the date jumps from 2012 to 2042, and the 2042 dates probably correspond to your use. The month and day are wrong too.

My speculation is that a battery inside the machine died when it wasn't used and the machine's internal clock, and possibly other memory, lost its date setting.

Code: Select all

Icon++ SUM Session ts 1352871426 "11/13/2012 22:37:06"
Icon++ SUM Session ts 1353468084 "11/20/2012 20:21:24"
Icon++ SUM Session ts 1354506494 "12/02/2012 20:48:14"

Change occurs here ===>
Icon++ SUM Session ts 2298983710 "11/07/2042 07:35:10"
Icon++ SUM Session ts 2298985380 "11/07/2042 08:03:00"
Icon++ SUM Session ts 2298990516 "11/07/2042 09:28:36"
Icon++ SUM Session ts 2298996228 "11/07/2042 11:03:48"
Icon++ SUM Session ts 2299006702 "11/07/2042 13:58:22"
Icon++ SUM Session ts 2297473560 "10/20/2042 20:06:00"
Icon++ SUM Session ts 2297514378 "10/21/2042 07:26:18"
Icon++ SUM Session ts 2297515906 "10/21/2042 07:51:46"
Icon++ SUM Session ts 2297522022 "10/21/2042 09:33:42"
Icon++ SUM Session ts 2297522972 "10/21/2042 09:49:32"
Icon++ SUM Session ts 2297538410 "10/21/2042 14:06:50"
Icon++ SUM Session ts 2297541654 "10/21/2042 15:00:54"
Icon++ SUM Session ts 2297580216 "10/22/2042 01:43:36"
Icon++ SUM Session ts 2297591358 "10/22/2042 04:49:18"
Icon++ SUM Session ts 2297614404 "10/22/2042 11:13:24"
Icon++ SUM Session ts 2297669598 "10/23/2042 02:33:18"
Icon++ SUM Session ts 2297698496 "10/23/2042 10:34:56"
Icon++ SUM Session ts 2297708068 "10/23/2042 13:14:28"
Icon++ SUM Session ts 2297713348 "10/23/2042 14:42:28"
... etc...

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

User avatar
BlueDragon
Posts: 541
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by BlueDragon » Mon Sep 20, 2021 11:22 pm

I think that, for now, we have to call the Icon+ "unsupported." It's going to take some work to solve the problems. We would probably have to rework the loader, which is rarely an easy or quick job. More people use new machines rather than ones from the past, so supporting new machines helps more CPAP users.

I wonder if the Icon+ firmware might have been updated sometime since 2012? Do you recall seeing any messages about updates on your machine? It looks to me like we might be seeing data, from the same machine, in two different formats.

UPDATE: I no longer think a firmware update is an issue.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
Last edited by BlueDragon on Tue Sep 21, 2021 11:34 am, edited 1 time in total.
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.

vilnius
Posts: 19
Joined: Mon Sep 20, 2021 5:01 pm

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by vilnius » Tue Sep 21, 2021 3:13 am

BlueDragon wrote:
Mon Sep 20, 2021 11:22 pm
I think that, for now, we have to call the Icon+ "unsupported." It's going to take some work to solve the problems. We would probably have to rework the loader, which is rarely an easy or quick job. More people use new machines rather than ones from the past, so supporting new machines helps more CPAP users.

I wonder if the Icon+ firmware might have been updated sometime since 2012? Do you recall seeing any messages about updates on your machine? It looks to me like we might be seeing data, from the same machine, in two different formats.
I didn't see anything about updates on the display or when I plugged in the SmartStick. Would wiping out the data and starting with a blank card help? Is that even possible? I'm a total newb both to CPAP machines and OSCAR

dataq1
Posts: 820
Joined: Sat Dec 24, 2005 8:09 am
Location: Northeast Ohio

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by dataq1 » Sat Oct 02, 2021 7:38 pm

1) It is not clear to me how to download the 1.3 Oscar Version, details please
2) I am using a Respironics Dreamstation Autopap ( 500x110) but my wife has a Resmed Airsense 11. Will I need to use two different computers or will the v 1.3 support both data systems (with different profiles of course)?
"THE INFORMATION PROVIDED ON CPAPTALK.COM IS NOT INTENDED NOR RECOMMENDED AS A SUBSTITUTE FOR PROFESSIONAL MEDICAL ADVICE."

User avatar
BlueDragon
Posts: 541
Joined: Sun May 27, 2018 10:20 am
Location: Sonoran Desert

Re: OSCAR 1.3.0 Beta: AirSense 11, SleepStyle, DeVilbiss

Post by BlueDragon » Sat Oct 02, 2021 8:28 pm

Thank you for your interest in helping with beta testing OSCAR. Please see https://www.apneaboard.com/wiki/index.p ... _Beta_Test for information about beta testing and the skills we are looking for in beta testers.

We are not actively soliciting more beta testers at this time, as we are on our second beta and, hopefully, getting closer to release. OSCAR will advise you when a new release is available.

And, yes, two different profiles is what you will need to support two people with different CPAP machines.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Brevida™ Nasal Pillow CPAP Mask with Headgear
Additional Comments: Dx Mar 2018 (AHI=24, RDI=54; AHI=73 supine). Started APAP June 2018, VAuto Aug 2020.
OSCAR 1.5.1 is the latest release.
OSCAR Team
ResMed AirCurve 10 VAuto, F&P Brevida.
FlashAir SD and FlashPap for data transfer.