What an informative discussion!
I'ld like to suggest a slight change in perspective though.
Derek wrote:"It seems as if an NR event takes precedence over everyything elese"
An NR is not a clinical event. It is
a sequence of events defined by
Respironics. The following is a quote from
respironics, presented so as to emphasise the sequence aspect of it. For the sake of clarity, I did drop 6 words from the last sentence.
1. At any pressure 8 cm H2O or higher we will make three pressure increases in response to a sustained string of events.
2. If there is no improvement after the third increase, (indicated by the persistence of events)
3. the pressure is dropped 2 cm and a constant pressure is held for several minutes.
4. If there is snoring noted during this period of constant pressure, which would indicate obstruction, we will increase pressure
5. … thus allowing for three more pressure increases.
Respironics' machine is basically instructed to go slow as it raises pressure above 8, and use snores as an indication of obstruction. The deal is, as Derek already noted: "You snore, we raise the pressure, because it indicates the obstruction we're here to remove. You don't snore, we wait".
If you look at the snore indicators, underneath loonlvrs two apenea sequences, you can see how they correlate with pressure raises. This is actually very clear in his first chart, starting at hour 4. I would guess that some of those snores (e.g. the one just near the 5th hour in chart two), coincided with the machine's instruction to lower pressure, and that instruction took precedence.
This of course raise the questions of
Timing: how long is "several minutes" of constant pressure, and is there a bottom limit in which snores will not affect it?
Precedence: when a snore and an instruction to lower pressure come together, is the snore given precedence (as is should be?)
Mechanics: once the machine starts lowering pressure , how soon can it switch to raising the pressure, because it got new("here's a snore, therefore this is an obstruction") information?