Page 3 of 4

Re: Hey guys

Posted: Wed Jun 10, 2020 10:05 am
by Juggsy75
Pugsy wrote: ↑
Wed Jun 10, 2020 9:52 am
That entire segment you show appears to be arousal related breathing to me. Can't see the breathing pattern immediately prior to that segment but if that were my report I would chalk those flagged events up to arousal related and thus they don't count.

A few weeks ago I had a night with AHI of 9.4. Unusual for me for sure so I looked more closely and I manually counted and looked at each flagged event. 90% of those flagged events were so obviously arousal/awake related that a blind man could see it.
It also happened to be a night where I slept poorly due to back pain. I had done some yard work outside that I paid for in the middle of the night. :lol: I over did things. I knew I woke up often because I hurt like hell.

Now sometimes we do have real asleep events that cause the subsequent arousal breathing and flagging. So sometimes a real event ends up causing a lot of false positive events. So if we see that happening a lot then obviously we want to reduce those primary real events and the best way to do that...more baseline pressure or allow the machine to auto adjust if the machine is capable.
My own AHI is very rarely more than 25% real events. The bulk of my AHI is always arousal related and not related to real events causing the arousals so I don't see the need to use more pressure.
Looking through mine a lot of mine seem to be arousel based, I do find I wake a lot during night and Toss and turn, also I sigh a lot even in day not sure why, unless I’m doing same during sleep aswell🤷🏻‍♂️🤷🏻‍♂️........ so it would seem my score of over 5 is probably considerably less, as quite a lot were the snore ones aswell, I felt really good this morning, felt like had good sleep, no vivid dreams or waking to turn over

Re: Hey guys

Posted: Wed Jun 10, 2020 10:14 am
by Pugsy
Let your body be your guide and try not to dwell too much on the numbers.
If you feel decent...take it and be happy. Numbers don't mean much but how you feel means a lot more.
The best we can do with the false positive numbers thing is to try to figure out why we toss and turn and see what we can do to improve our sleep quality in general.
The machine does it's part with our sleep quality related to the airway itself but it can't help bad sleep that isn't related to airway issues and the list of potential culprits that mess with our sleep is miles long.

For me as long as the "real" AHI numbers, meaning I was asleep for sure when they happened, are below 2.0 then I am good with the OSA side of my therapy. I often have to manually adjust an elevated AHI to remove those false positive arousal related flags because they aren't related to the OSA side of my sleep. I let my body be my guide as best I can.

Re: Hey guys

Posted: Wed Jun 10, 2020 11:00 am
by Juggsy75
Yes, I just am more focused I think due to my afib, if keeping osa under control helps that then great stuff, wish I knew how to engage the semi auto mode my nurse spoke about to see if it helps

Re: Hey guys

Posted: Wed Jun 10, 2020 11:06 am
by Pugsy
Juggsy75 wrote: ↑
Wed Jun 10, 2020 11:00 am
wish I knew how to engage the semi auto mode my nurse spoke about to see if it helps
You have to go into the clinical menu setup area and there should be an auto trial mode option. You just change to that mode.
If you don't know how to get into the clinic menu setup area it is explained in the clinical/provider menu which you can get here via email by requesting it. Just follow the instructions.
The manual you would request is the Respironics PR System One 60 Series REMstar Pro with C-flex+
https://www.apneaboard.com/adjust-cpap- ... tup-manual
It's easy to get to...just a little combination of button pushing.

Re: Hey guys

Posted: Mon Jun 15, 2020 6:25 am
by Juggsy75
hi again all, so last few nights have been very bad, to point of wondering if my machine is actually right or not, night before last was AHI of 4.8 which i wasnt happy about, then last night was worse, ill give you a bit of backstory to it.

So we were in bed and heard people being loud outside around 11.30pm, so i went out to ask as my wife was up for work at 6am could they keep it down or move away from house, anyway the man took exception and wanted to fight me, eventually they left and i went to bed, i awoke at around 3:40am to use toilet came back and checked how id slept.....to my horror id only been sleeping 4 hours but my AHI was 14.1.......my score without cpap at diagnosis was only 14.5........so i went back to sleep woke around 8am and AHI was 7.4 for night still very high for me, my wife thinks maybe the anxiety of what happened may have played a part, she also said i was quite restless last night, tossed and turned a bit,

so ive included my graphs and some close ups of some hypopneas to see if theyre real or machine related as im struggling to tell,

Re: Hey guys

Posted: Mon Jun 15, 2020 6:26 am
by Juggsy75
theres also a period aroiund 4:20am where the pressure and leak rate seem to drop right off the scale

Re: Hey guys

Posted: Tue Jun 16, 2020 6:38 am
by Juggsy75
Upped my pressure again last night to 8 to see how that goes ahi this morning was 2.9, I’m wondering whether to request a different machine

Re: Hey guys

Posted: Thu Jun 18, 2020 9:40 am
by Juggsy75
Any thoughts anyone???

Re: Hey guys

Posted: Thu Jun 18, 2020 10:58 am
by Pugsy
I would have liked to see the report with the minimum of 8.0 along with comments about how you feel and how you slept.
That's what I was waiting for. I don't really do much or say much with a number only to go on.

As for the missing graph lines. I have no idea except maybe a corrupt data file.

Re: Hey guys

Posted: Fri Jun 19, 2020 9:20 am
by Juggsy75
Pugsy wrote: ↑
Thu Jun 18, 2020 10:58 am
I would have liked to see the report with the minimum of 8.0 along with comments about how you feel and how you slept.
That's what I was waiting for. I don't really do much or say much with a number only to go on.

As for the missing graph lines. I have no idea except maybe a corrupt data file.
apologies, so heres last 3 nights at pressure set to eight, i just seem to be all over place, to point where im anxious going to bed and first thing i think of when i wake is the numbers, i generally feel im having good sleeps....i usually work shifts so never same pattern of sleep but im currently on furlough so have really got into a good pattern of sleep, however im noticing im having really really vivid dreams which i didnt used to have. last night was 6.6 but i fely ok this morning

Re: Hey guys

Posted: Fri Jun 19, 2020 11:13 am
by Pugsy
When you remember dreams it isn't a good thing. It means you woke up during a dream or right at the end of the dream so remembering a lot of dreams means you woke up a lot and it's the waking up a lot that is the problem.
Now why the wake up happened...that's an unknown. It could be airway related or it might be something else.

3 nights and 2 of them were decent enough nights. The one ugly night I don't know if those flagged events are false positive arousal/awake related events or real asleep apnea related events. When we wake up a lot and toss and turn a lot the irregular awake breathing will often get flagged as some sort of apnea event but if we aren't asleep they don't count. The machine doesn't know though because it only measures air flow. It has no way to know if we are asleep or not.

So...was the ugly AHI night on the 18th a night where you didn't sleep or great and the AHI mainly false positive flagged events (sometimes called SWJ or sleep/wake/junk) or did something special happen to make the airway more difficult to hold open...like maybe more on your back or something like that???
We don't know.
You can learn to evaluate the flow rate to get sort of an idea if you were asleep when those flagged events happened. Watch all the videos here.
http://freecpapadvice.com/sleepyhead-free-software
Takes a lot of watching and evaluating and sometimes even then it's hard to tell for sure but you can usually get a close enough idea to get you in the ball park.

Since 2 out of 3 nights were decent....continue with the 8 cm pressure for 4 more nights so we have a full week at this pressure and lets see if you have more good nights or more bad nights.

Try to figure out why you have some nights where you don't sleep so great if you can. Keep a log of how you slept and feel along with anything that might be a factor. Late meals...alcohol....stress...etc.

Re: Hey guys

Posted: Fri Jun 19, 2020 1:22 pm
by Juggsy75
Pugsy wrote: ↑
Fri Jun 19, 2020 11:13 am
When you remember dreams it isn't a good thing. It means you woke up during a dream or right at the end of the dream so remembering a lot of dreams means you woke up a lot and it's the waking up a lot that is the problem.
Now why the wake up happened...that's an unknown. It could be airway related or it might be something else.

3 nights and 2 of them were decent enough nights. The one ugly night I don't know if those flagged events are false positive arousal/awake related events or real asleep apnea related events. When we wake up a lot and toss and turn a lot the irregular awake breathing will often get flagged as some sort of apnea event but if we aren't asleep they don't count. The machine doesn't know though because it only measures air flow. It has no way to know if we are asleep or not.

So...was the ugly AHI night on the 18th a night where you didn't sleep or great and the AHI mainly false positive flagged events (sometimes called SWJ or sleep/wake/junk) or did something special happen to make the airway more difficult to hold open...like maybe more on your back or something like that???
We don't know.
You can learn to evaluate the flow rate to get sort of an idea if you were asleep when those flagged events happened. Watch all the videos here.
http://freecpapadvice.com/sleepyhead-free-software
Takes a lot of watching and evaluating and sometimes even then it's hard to tell for sure but you can usually get a close enough idea to get you in the ball park.

Since 2 out of 3 nights were decent....continue with the 8 cm pressure for 4 more nights so we have a full week at this pressure and lets see if you have more good nights or more bad nights.

Try to figure out why you have some nights where you don't sleep so great if you can. Keep a log of how you slept and feel along with anything that might be a factor. Late meals...alcohol....stress...etc.
I’m stressed at the numbers at moment, my dreams are incredibly vivid and lifelike, I’ve watched the videos and I think some are false but not 100% on them really

Re: Hey guys

Posted: Fri Jun 19, 2020 2:18 pm
by Pugsy
Try to quit stressing over numbers. It will just make things worse and there's nothing you can do about water under the bridge anyway.
We all have bad nights occasionally. Heck a couple of weeks ago I had a 9.4 AHI night and I have been on cpap over 11 years now.
Crap happens...even to cpap veterans.

Some years ago I was doing some fixed pressure experiments and I picked a setting that gave me good "numbers" and decided to keep it for 6 weeks. Watched the numbers go up and down and sometimes a lot up but I knew that I got some good numbers once and I felt I could do it again. Mtore and more the good nights started to out number the bad nights and at the end of 6 weeks and I never changed a thing my AHI was 50% less consistently than when I started at the beginning of the experiment. The bad nights became extremely rare.

But....like I said we can still have some bad nights...and mine are always related to awakening/arousal breathing false positives or at least the bulk of them.
I might have 40 flagged events during an 8 hour session and only 5 of those events are real.
So AHI of 5.0....really is a 1.6 AHI

Re: Hey guys

Posted: Fri Jun 19, 2020 2:50 pm
by Juggsy75
Pugsy wrote: ↑
Fri Jun 19, 2020 2:18 pm
Try to quit stressing over numbers. It will just make things worse and there's nothing you can do about water under the bridge anyway.
We all have bad nights occasionally. Heck a couple of weeks ago I had a 9.4 AHI night and I have been on cpap over 11 years now.
Crap happens...even to cpap veterans.

Some years ago I was doing some fixed pressure experiments and I picked a setting that gave me good "numbers" and decided to keep it for 6 weeks. Watched the numbers go up and down and sometimes a lot up but I knew that I got some good numbers once and I felt I could do it again. Mtore and more the good nights started to out number the bad nights and at the end of 6 weeks and I never changed a thing my AHI was 50% less consistently than when I started at the beginning of the experiment. The bad nights became extremely rare.

But....like I said we can still have some bad nights...and mine are always related to awakening/arousal breathing false positives or at least the bulk of them.
I might have 40 flagged events during an 8 hour session and only 5 of those events are real.
So AHI of 5.0....really is a 1.6 AHI
I can tell the obstructive sir centrals that are false I think as there’s a lot of jumpy lines before them not steady rhythm, it’s the hypopneas I’m having trouble seeing are one or another

Re: Hey guys

Posted: Fri Jun 19, 2020 5:22 pm
by Pugsy
Hyponeas aren't always so easy to tell but if you look at the breathing prior to the flag it should give you a good idea.

Also remember....you can have a real asleep event as the first flag and then that event causes the arousal and subsequent false positive flags are related to the first event causing the arousal.
I don't have a good example of that handy but I will go see if I can find one. Won't be easy though because I don't have many real asleep events at all. The bulk of mine are all SWJ false positives due to pain and tossing and turning a lot.

But I do have these handy.

These 2 flagged events below...probably asleep real events. Looks like sleep before and after the flags.
Image

This one below...all SWJ false positives.
Image

This one below the asleep time is circled in red...then arousal breathing comes after and everything flagged after is false positives.
Image
And a more zoomed out view of the red circled asleep breathing and what happened after the arousal...everything else is arousal breathing.
Image