Page 2 of 2

Re: Things...

Posted: Thu May 14, 2020 4:44 pm
by Pugsy
ASV machines are much more aggressive and when they "do their thing" it can take quite a bit of adjusting to get used to.

Are you spending much time with the mask and machine on and you are awake?

I have one suggestion to try for now....your PS setting...let's limit the max PS and stop the rapid wide fluctuations.
Set the minimum PS to 3 which is where it is now and set the max PS to 5.
No other changes at this time.

Re: Things...

Posted: Thu May 14, 2020 5:16 pm
by DMA97701
Ok...I'll try that for tonight. Thanx again.

Re: Things...

Posted: Thu May 14, 2020 11:00 pm
by DMA97701
Pugsy wrote:
Thu May 14, 2020 4:44 pm
Set the minimum PS to 3 which is where it is now and set the max PS to 5.
No other changes at this time.
Well, darn. It won't let me set the max PS down below a difference of 5 between the min PS and the max PS. Therefore, if my min PS is 3, my max PS can go no lower than 8. So, if I want to set the max PS to 5, the min PS has to be 0. I'm going to leave it at 3 (min) and 8 (max) for tonight. Thanx again.

Re: Things...

Posted: Fri May 15, 2020 9:00 am
by DMA97701
So, last night was the longest I've slept so far. Sadly, I woke up in the middle of the night due to back pain, but I didn't have any of that feeling of too much pressure on the mask and breathing (exhaling) was much easier. I'm going to keep these settings for a while and see how it goes. Thanx for the assistance!
:D

Re: Things...

Posted: Mon May 18, 2020 9:41 pm
by DMA97701
DMA97701 wrote:
Fri May 15, 2020 9:00 am
So, last night was the longest I've slept so far. Sadly, I woke up in the middle of the night due to back pain, but I didn't have any of that feeling of too much pressure on the mask and breathing (exhaling) was much easier. I'm going to keep these settings for a while and see how it goes. Thanx for the assistance!
:D
Some progress is better than none, right?
Image