I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
Iamknoware
Posts: 5
Joined: Sun Dec 29, 2019 9:46 am

I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Iamknoware » Sun Dec 29, 2019 7:04 pm

Insurance has yet get back to me (Its 2 months...), so I just bought one off Direct Medical (https://www.directhomemedical.com/dsx50 ... onics.html). I didn't even know there were different types of machine until they told my prescription was for a machine with Auto setting.

My sleep center I went to wanted $75-$150 to set up my new device. Nah.

Well I think I'm done setting up, but I'm not 100% sure. I went through my clinical settings on my old machine F&P Icon+ and basically copied and paste? Any critical information would be greatly appreciated.

I already went through the 3 part series from this YT channel
https://m.youtube.com/watch?v=xM5fY53Q0Vw

https://m.youtube.com/watch?v=MYtcyviA-ic

https://m.youtube.com/watch?v=0BhrJPDq_5s

[UPDATED] I was recently introduced to this website and Oscar, so I hope this info can help with getting my unit set up properly. [UPDATED] screenshots https://imgur.com/a/dd4HTHa

So basically, I dont think i set it up properly, because I dont have that refreshed feeling I once did since my old machine. I am really trying to avoid going to sleep center for set up, but they want to charge me. My AHI is extremely high and Im trying to get that fixed.
Last edited by Iamknoware on Mon Dec 30, 2019 1:18 am, edited 2 times in total.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by palerider » Sun Dec 29, 2019 9:14 pm

Was there a question that I missed?

Please post screenshots, not photos of computer screens.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11284
Joined: Fri Jun 19, 2015 4:36 pm

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by zonker » Sun Dec 29, 2019 9:27 pm

Iamknoware wrote:
Sun Dec 29, 2019 7:04 pm
Insurance has yet get back to me (Its 2 months...), so I just bought one off Direct Medical (https://www.directhomemedical.com/dsx50 ... onics.html). I didn't even know there were different types of machine until they told my prescription was for a machine with Auto setting.

welcome to the zoo! and congrats on taking your therapy into your own hands.

but as palerider says, we need to see screenshots, not shots of your screen. :D

please see the links below for newbie posting info. then please post some oscar screenshots following the tips in the newbie post. it will say sleepyhead there, but same instructions work for oscar.

good luck!
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Okie bipap
Posts: 3567
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Okie bipap » Sun Dec 29, 2019 10:23 pm

To learn how to change the setting, you can get a copy of the Clinical Manual here:

https://www.apneaboard.com/adjust-cpap- ... tup-manual

You will find the directions for ordering the Clinical Manual about 2/3 down the page. Once you get the manual, it will tell you how to get into the clinician menu to set your pressure and turn on ramp if you want to use it. The default for the machine is 4 to 20, but many people find it hard to breath at a pressure of 4 cm. I recommend you set the low end to 6.

If you want to start using the machine before getting the manual, just plug it in and hook up your mask. The machine should already be set for a pressure range of 4 to 20. If using the humidifier, it is recommended to use distilled water. However, any drinkable water can be used. If you use tap water, you may get a build up of mineral deposits in the water chamber. If this happens, they can be removed with a little vinegar or lemon juice.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

Iamknoware
Posts: 5
Joined: Sun Dec 29, 2019 9:46 am

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Iamknoware » Mon Dec 30, 2019 1:13 am

palerider wrote:
Sun Dec 29, 2019 9:14 pm
Was there a question that I missed?

Please post screenshots, not photos of computer screens.
I updated the screenshots, https://imgur.com/a/dd4HTHa

Iamknoware
Posts: 5
Joined: Sun Dec 29, 2019 9:46 am

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Iamknoware » Mon Dec 30, 2019 1:16 am

Okie bipap wrote:
Sun Dec 29, 2019 10:23 pm
To learn how to change the setting, you can get a copy of the Clinical Manual here:

https://www.apneaboard.com/adjust-cpap- ... tup-manual

You will find the directions for ordering the Clinical Manual about 2/3 down the page. Once you get the manual, it will tell you how to get into the clinician menu to set your pressure and turn on ramp if you want to use it. The default for the machine is 4 to 20, but many people find it hard to breath at a pressure of 4 cm. I recommend you set the low end to 6.

If you want to start using the machine before getting the manual, just plug it in and hook up your mask. The machine should already be set for a pressure range of 4 to 20. If using the humidifier, it is recommended to use distilled water. However, any drinkable water can be used. If you use tap water, you may get a build up of mineral deposits in the water chamber. If this happens, they can be removed with a little vinegar or lemon juice.
The link doesnt work. But right now my AHI is extremely high, so im trying to figure how to get that low.

Iamknoware
Posts: 5
Joined: Sun Dec 29, 2019 9:46 am

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Iamknoware » Mon Dec 30, 2019 1:18 am

zonker wrote:
Sun Dec 29, 2019 9:27 pm
Iamknoware wrote:
Sun Dec 29, 2019 7:04 pm
Insurance has yet get back to me (Its 2 months...), so I just bought one off Direct Medical (https://www.directhomemedical.com/dsx50 ... onics.html). I didn't even know there were different types of machine until they told my prescription was for a machine with Auto setting.

welcome to the zoo! and congrats on taking your therapy into your own hands.

but as palerider says, we need to see screenshots, not shots of your screen. :D

please see the links below for newbie posting info. then please post some oscar screenshots following the tips in the newbie post. it will say sleepyhead there, but same instructions work for oscar.

good luck!
I updated the screenshots, https://imgur.com/a/dd4HTHa

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by palerider » Mon Dec 30, 2019 1:32 am

Iamknoware wrote:
Mon Dec 30, 2019 1:13 am
palerider wrote:
Sun Dec 29, 2019 9:14 pm
Was there a question that I missed?

Please post screenshots, not photos of computer screens.
I updated the screenshots, https://imgur.com/a/dd4HTHa
Jibbers Crabst! that's awful.

I'd prefer you do your screenshots like this:
http://cpaptalk.com/wiki/index.php/Oscar:organize

The one you did looked like it was a max of 9cm pressure, and you need more, min 10 max 20 for a start.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by palerider » Mon Dec 30, 2019 1:34 am

Iamknoware wrote:
Mon Dec 30, 2019 1:16 am
Okie bipap wrote:
Sun Dec 29, 2019 10:23 pm
To learn how to change the setting, you can get a copy of the Clinical Manual here:

https://www.apneaboard.com/adjust-cpap- ... tup-manual
.
The link doesnt work. But right now my AHI is extremely high, so im trying to figure how to get that low.
Correct link : https://www.apneaboard.com/adjust-cpap- ... tup-manual

I recommend having the manual, it explains what all the settings do, plus there's good info in there, I recommend a read through.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Jas_williams
Posts: 1120
Joined: Tue Dec 19, 2017 2:12 pm
Location: Somerset UK

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Jas_williams » Mon Dec 30, 2019 7:42 am

Did you have ez start turned on if so turn it off as it is limiting your pressure and you need more pressure now not in 13 days time

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: Using sleepyhead and a pressure of 6 - 21 Resmed S9 Adapt SV with a Bleep Sleep Mask

Iamknoware
Posts: 5
Joined: Sun Dec 29, 2019 9:46 am

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Iamknoware » Thu Jan 02, 2020 10:46 am

It's on at the moment. I have been trying different settings. Right now it's 7-18. My AHI was 73 last night. I'm thinking off tapping out and bringing the device to the docs and let them set it up.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by palerider » Thu Jan 02, 2020 10:51 am

Iamknoware wrote:
Thu Jan 02, 2020 10:46 am
It's on at the moment. I have been trying different settings. Right now it's 7-18. My AHI was 73 last night. I'm thinking off tapping out and bringing the device to the docs and let them set it up.
Ezstart is one of the stupid things Philips put on the machines.

Shoulda called it "choke-more"

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Last edited by palerider on Thu Jan 02, 2020 10:55 am, edited 1 time in total.
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65062
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Pugsy » Thu Jan 02, 2020 10:53 am

Please share an image of last night's detailed report.

I am a little late here but your first image looked like EZ Start was turned on and that's causing the machine to not be able to do what it is designed to do.

You DON'T want to be using EZ Start mode. Turn it off or be prepared for the machine to take over a month to maybe do a decent job of holding your airway open. It is NOT your friend at all. Stupidest thing Respironics ever offered on a machine.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4211
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: I bought a Philips Dreamstation Auto CPAP, just need some help setting it, properly. (Very high AHI w/ Oscar data)

Post by Dog Slobber » Thu Jan 02, 2020 11:02 am

Iamknoware wrote:
Thu Jan 02, 2020 10:46 am
It's on at the moment. I have been trying different settings. Right now it's 7-18. My AHI was 73 last night. I'm thinking off tapping out and bringing the device to the docs and let them set it up.
There isn't magic button or configuration he's going to set that's going to reduce your AHI from 73 to 3.
  • Set your device to the values suggested
  • Leave it alone for the night, OSCAR doesn't report properly if there are multiple configurations in a day
  • Present the screen shots as asked
  • Be meticulous about reporting the settings accurately (ie Pressures, EZ Start, flex)