Re: Do asv machines eliminate centrals?
Posted: Sat Jul 06, 2019 6:02 pm
by zonker
Chapalauser wrote: ↑Sat Jul 06, 2019 5:31 pm
I have attached a sleepyhead screenshot
I am having a problem attaching a screenshot
okay. at the bottom of your text, there is a tab that says ATTACHMENT. click that.
next click the box that reads "Add files"
pick the screenshot you want to attach.
double click it.
you'll see the name of your file and next to it "place inline". click that.
screenshot-20190629-114643.png
viola! your screenshot should appear as mine has above.
Re: Do asv machines eliminate centrals?
Posted: Sat Jul 06, 2019 6:40 pm
by Chapalauser
Here is a recent screenshot. any comments would be appreciated

- screenshot-20190706-174536.png (240.2 KiB) Viewed 2763 times
Re: Do asv machines eliminate centrals?
Posted: Sat Jul 06, 2019 6:56 pm
by palerider
Chapalauser wrote: ↑Sat Jul 06, 2019 6:40 pm
Here is a recent screenshot. any comments would be appreciatedscreenshot-20190706-174536.png
I would lower your PS, try bumping the minEPAP to 6 and drop PS to 2, and see what happens.