RB wrote:Once again does anyone know where to find a complete listing of the error codes for the Remstar Auto? Not just one at a time but the complete list.
Thank you
Yeah.....they're in the "Service Manual" for the machines. The manuals are in .pdf format.
But.....just for YOU.....I've converted that section to text format and am posting the complete list for the Auto and Pro 2. Be sure to copy them to your computer so I don't have to do this AGAIN!
REMstar Auto & Pro 2 error codes:
Error Code - Description - Probable Cause - Corrective Action
E-0 NO_ERROR •Reserved •N/A
E-1 ERR_SOFTWARE •Reserved •Replace Main PCA
E-2 ERR_PROGRAM_CRC_FAILURE •Defective FLASH chip •Replace Main PCA
E-3 ERR_EXTERNAL_RAM_FAILURE •Defective RAM chip •Replace Main PCA
E-4 ERR_FIQ_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-5 ERR_FIQ_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-6 ERR_NEST_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-7 ERR_NEST_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-8 ERR_IRQ_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-9 ERR_IRQ_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-10 ERR_TMR_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-11 ERR_TMR_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-12 ERR_SVC_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-13 ERR_SVC_STACK_UNDERFLOW •Defective RAM chip •Replace Main PCA
E-14 ERR_THREAD_STACK_OVERFLOW •Defective RAM chip •Replace Main PCA
E-15 ERR_ARM_UNDEFINED_INSTRUCTION_EXCEPTION •Defective FLASH chip •Replace Main PCA
E-16 ERR_ARM_SWI_EXCEPTION •Defective FLASH chip •Replace Main PCA
E-17 ERR_ARM_PREFETCH_EXCEPTION •Defective FLASH chip •Replace Main PCA
E-18 ERR_ARM_ABORT_EXCEPTION •Defective RAM chip •Replace Main PCA
E-19 ERR_ARM_RESERVED_EXCEPTION •Defective FLASH chip •Defective microprocessor •Replace Main PCA •Replace Main PCA
E-20 ERR_AIC_DEFAULT_HANDLER •Defective FLASH chip •Replace Main PCA
E-21 ERR_AIC_SPURIOUS_HANDLER •Defective FLASH or RAM chip •Replace Main PCA
E-22 ERR_CAL_DATA_CRC_FAILURE •Improper calibration •Defective FLASH chip •Recalibrate •Replace Main PCA
E-23 ERR_INVALID_DRIFT_INDEX_AT_CAL •Improper calibration •Occluded or restricted pressure tube •Defective flow sensor •Recalibrate •Replace pressure tube •Replace Main PCA
E-24 ERR_UNRECOGNIZED_CAL_VERSION_BY_SW •Improper calibration •Defective FLASH chip •Recalibrate •Replace Main PCA
E-25 ERR_DRIFT_FLOW_TOO_HIGH •Occluded or restricted pressure tube •Defective flow sensor •Replace Flow Tube •Replace Main PCA
E-26 ERR_UI_DISPLAY_FAILURE •Defective LCD driver chip •Replace Main PCA
E-27 ERR_DRIFT_PRESSURE_TOO_HIGH •Occluded or restricted pressure tube •Defective pressure sensors •Replace pressure tubing •Replace Main PCA
E-28 ERR_CAL_DATA_BLANK •Improper calibration •Defective FLASH chip •Recalibrate •Replace Main PCA
E-29 ERR_DRIFT_SECONDARY_PRESSURE_TOO_HIGH •Occluded or restricted pressure tube •Defective pressure sensors •Replace pressure tubing •Replace Main PCA
E-30 ERR_INVALID_NVDATA_SCHEMA •Normal for first time power up •Defective EEPROM •Remove & restore power •Replace Main PCA
E-31 ERR_INVALID_NVDATA_LOCK •Defective EEPROM •Replace Main PCA
E-32 ERR_INVALID_NVDATA_CHECKSUM •Normal for first time power up •Defective EEPROM •Remove & restore power •Replace Main PCA
E-33 ERR_NVDATA_STORAGE_ERROR •Defective EEPROM •Replace Main PCA
E-34 ERR_NVDATA_QUEUE_FULL •Defective EEPROM •Replace Main PCA
E-35 ERR_NVDATA_READBACK_FAILURE •Defective EEPROM •Replace Main PCA
E-36 ERR_NVDATA_RANGE_FAILURE •Defective EEPROM •Replace Main PCA
E-37 ERR_CORRUPT_RTC_VALUE •Defective microprocessor •Replace Main PCA
E-38 ERR_RTC_NOT_TICKING •Defective microprocessor •Defective crystal •Replace Main PCA •Replace Main PCA
E-39 ERR_UI_QUEUE_FULL •Defective PCA •Replace Main PCA
E-40 ERR_INVALID_BIST_TEST_CALL •Invalid RASP command •Reissue RASP command
E-41 ERR_BIST_THREAD_LOCKED •Defective PCA •Replace Main PCA
E-42 ERR_ALARM_VOLTAGE_FAILURE •Defective power supply •Replace Main PCA
E-43 ERR_12VREF_FAILURE •Defective power supply •Replace Main PCA
E-44 ERR_5VREF_FAILURE •Defective power supply •Replace Main PCA
E-45 ERR_VBULK_FAILURE •Defective power supply •Replace Main PCA
E-46 ERR_NEG15VREF_FAILURE •Defective power supply •Replace Main PCA
E-47 ERR_FL12_SENSOR_RAILED •Occluded or restricted pressure tube •Defective flow sensor •Replace flow tube •Replace Main PCA
E-48 ERR_PV2_SENSOR_RAILED •Occluded or restricted pressure tube •Defective pressure sensor •Replace pressure tube •Replace Main PCA
E-49 ERR_FLASH_FAILURE •Defective FLASH chip •Replace Main PCA
E-50 ERR_UNRECOGNIZED_BOARD_REV •Needs new software •Defective PCA •Replace Main PCA •Replace Main PCA
E-51 ERR_MOTOR_FAILURE •Defective motor drive chip •Replace Main PCA
E-52 ERR_BLOWER_RPM_FAILED_TOLERANCE •Stalled or locked blower •Replace Blower •Replace Main PCA
E-53 ERR_BLOWER_OFF_CURRENT_HIGH •Stalled or locked motor •Defective motor drive chip •Replace Blower •Replace Main PCA
E-54 ERR_BLOWER_ON_CURRENT_HIGH •Stalled or locked motor •Defective motor drive chip •Replace Blower •Replace Main PCA
E-55 ERR_PRINTF_BUFFER_OVERFLOW •Defective PCA •Replace Main PCA
E-56 ERR_THERAPY_EVENT_QUEUE_FULL •Defective PCA •Replace Main PCA
E-57 ERR_KEYPAD_STUCK_KEY •Verify all keys protrude through top •Defective Keypad •Defective pad on PCA •Align keys with top •Replace Keypad •Replace Main PCA
E-58 ERR_UNUSED_58 •Reserved
E-59 ERR_UNUSED_59 •Reserved
E-60 ERR_UNUSED_60 •Reserved
E-61 ERR_UNUSED_61 •Reserved
E-62 ERR_UNUSED_62 •Reserved
E-63 ERR_WATCHDOG_FAILURE •Defective microprocessor •Replace Main PCA
E-64 ERR_UNEXPECTED_WATCHDOG •Defective FLASH chip •Defective RAM chip •Replace Main PCA •Replace Main PCA
E-65 ERR_UNUSED_65 •Reserved
E-66 ERR_RTOS_INIT_FAILURE •Defective PCA •Replace Main PCA
E-67 ERR_SERIAL_QUEUE_FULL •Defective PCA •Replace Main PCA
E-68 ERR_SNORE_SAMPLING_THREAD_LOCKED •Defective PCA •Replace Main PCA
E-69 ERR_MAIN_EXECUTION_THREAD_LOCKED •Defective PCA •Replace Main PCA
E-70 ERR_INTERNAL_RAM_FAILURE •Defective microprocessor •Replace Main PCA
E-71 ERR_EVENT_LOG_QUEUE_FULL •Defective PCA •Replace Main PCA
E-72 ERR_TREND_DIV_BY_ZERO •Defective RAM chip •Replace Main PCA
E-73 ERR_ROLLOVER_CASE •Defective RAM chip •Replace Main PCA
E-74 ERR_WRONG_DISPLAY_TYPE •Defective PCA •Replace Main PCA
E-75 ERR_ALARM_CAPACITOR_FAILURE •Defective PCA •Replace Main PCA
E-76 ERR_OVER_PRESSURE •Occluded or restricted tube •Defective PCA •Replace Pressure tubing •Replace Main PCA
E-77 ERR_UNDER_PRESSURE •Occluded or restricted tube •Defective PCA •Replace Pressure tubing •Replace Main PCA
E-78 ERR_OVER_PRESSURE_REG •Occluded or restricted tube •Defective PCA •Replace Pressure tubing •Replace Main PCA
E-79 ERR_UNDER_PRESSURE_REG •Occluded or restricted tube •Defective PCA •Replace Pressure tubing •Replace Main PCA
E-80 ERR_ROTOR_LOCKED •Locked Motor •Defective PCA •Replace Motor •Replace PCA
E-81 ERR_MOTOR_EDGE_MISSED •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-82 ERR_MOTOR_EDGE_EXTRA •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-83 ERR_MOTOR_OVERSPEED •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-84 ERR_MOTOR_BRAKE •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-85 ERR_MOTOR_CURRENT •Defective Motor •Defective PCA •Replace Motor •Replace PCA
E-86 ERR_INVALID_ERROR_CODE •Reserved
6.5 SmartCard Error Codes - Code Label - User Action
C-1 SMARTCARD_12C_WRITE_FAIL Card inserted upside-down. Remove card and reinsert properly. Otherwise, card is corrupt and should be reprogrammed.
C-2 SMARTCARD_BAD_CARD_TYPE Card is not supported by this device. Otherwise, card is corrupt and should be reprogrammed.
C-3 SMARTCARD_BAD_CIB_VERSION Card is corrupt and should be reprogrammed.
C-4 SMARTCARD_SCRIPT_DURING_PROVIDER_MODE Remove card. Exit Provider Mode screens. Then reinsert card.
C-5 SMARTCARD_PRESENT_IN_CAL_MODE Remove card. Exit Calibration Mode. Then reinsert card.
C-6 SMARTCARD_BAD_CIB_CHECKSUM Card is corrupt and should be reprogrammed.
C-7 SMARTCARD_NOT_PRESENT
C-8 SMARTCARD_12C_READ_FAIL Card inserted upside down. Remove card and reinsert properly. Otherwise, card is corrupt and should be reprogrammed.
C-9 SMARTCARD_WRITE_CONFIRM_FAIL Card is corrupt and should be reprogrammed. Otherwise, replace card.
C-10 SMARTCARD_BAD_RI_BRAND Card is corrupt and should be reprogrammed.
C-100 SMARTCARD_BAD_CPB_CHECKSUM Card is corrupt and should be reprogrammed.
C-101 SMARTCARD_BAD_CPB_VERSION Card is corrupt and should be reprogrammed.
C-102 SMARTCARD_INVALID_UNIT
C-103 SMARTCARD_UNKNOWN_CPB_FORMAT Card contains a prescription format that is not supported. Perhaps card is intended for another device.
C-104 SMARTCARD_UNKNOWN_CPB_PRESCRIPTION Card contains a prescription type that is not supported. Perhaps card is intended for another device.
C-105 SMARTCARD_SCRIPT_PARM_RANGE Card contains prescription settings that are outside of this device’s range. Perhaps card is intended for another device.
C-106 SMARTCARD_BAD_CPB_SCRIPT_READY Card is corrupt and should be reprogrammed.
C-107 SMARTCARD_BAD_CPB_START_ADDR Card is corrupt and should be reprogrammed.
C-108 SMARTCARD_BAD_CPB_END_ADDR Card is corrupt and should be reprogrammed.
C-109 SMARTCARD_BAD_CPB_ERASE_PERMISSION Card is corrupt and should be reprogrammed.
C-200 SMARTCARD_BAD_CDCB_VERSION Card is corrupt and should be reprogrammed.
C-201 SMARTCARD_BAD_CDB_VERSION Card contains a storage format that is not supported. Perhaps card is intended for another device.
C-202 SMARTCARD_BAD_CDCB_CHECKSUM Card is corrupt and should be reprogrammed.
C-203 SMARTCARD_BAD_CDCB_CONTROL_BLOCKS Card is corrupt and should be reprogrammed.
C-204 SMARTCARD_BAD_CDCB_START_ADDR Card is corrupt and should be reprogrammed.
C-205 SMARTCARD_BAD_CDCB_END_ADDR Card is corrupt and should be reprogrammed.
C-206 SMARTCARD_BAD_CDB_START_ADDR Card is corrupt and should be reprogrammed.
C-207 SMARTCARD_BAD_CDB_END_ADDR Card is corrupt and should be reprogrammed.
C-300 SMARTCARD_BAD_UIDB_VERSION Card is corrupt and should be reprogrammed.
C-301 SMARTCARD_BAD_UIDB_CHECKSUM Card is corrupt and should be reprogrammed.
C-302 SMARTCARD_BAD_UIDB_START_ADDR Card is corrupt and should be reprogrammed.
C-303 SMARTCARD_BAD_UIDB_END_ADDR Card is corrupt and should be reprogrammed.
C-400 SMARTCARD_BAD_DDCB_VERSION Card is corrupt and should be reprogrammed.
C-401 SMARTCARD_BAD_DDCB_CHECKSUM Card is corrupt and should be reprogrammed.
C-402 SMARTCARD_BAD_DDCB_PROFILE_CONTROL_BLOCKS Card is corrupt and should be reprogrammed.
C-403 SMARTCARD_BAD_DDCB_HISTOGRAM_CONTROL_BLOCKS Card is corrupt and should be reprogrammed.
C-404 SMARTCARD_BAD_DDCB_FOSQ_COMPL_CONTROL_BLOCKS Card is corrupt and should be reprogrammed.
C-500 SMARTCARD_FILE_OPEN_FAIL Card is corrupt and should be reprogrammed.
C-501 SMARTCARD_FILE_TELL_FAIL Card is corrupt and should be reprogrammed.
C-502 SMARTCARD_FILE_SEEK_FAIL Card is corrupt and should be reprogrammed.
C-503 SMARTCARD_FILE_READ_FAIL Card is corrupt and should be reprogrammed.
C-504 SMARTCARD_FILE_WRITE_FAIL Card is corrupt and should be reprogrammed.
C-505 SMARTCARD_FILE_GET_PTR_FAIL Card is corrupt and should be reprogrammed.
C-506 SMARTCARD_FILE_SET_PTR_FAIL Card is corrupt and should be reprogrammed.