I have been on a CPAP machine for a little over a year and have noticed that I swallow air while using it. I have tried diff. sleeping positions and had the unit
adjusted down a little to try to prevent swallowing air. Still having issues with swallowing air.
Has anyone had similar problems and switched to BIPAP and had good results?
I use Resmed 10 and full mask.
CPAP Vs BIPAP Swallowing air
Re: CPAP Vs BIPAP Swallowing air
Air swallowing seems to be pressure on inhale related problem so it is unlikely that switching to a BiPap but keeping the same inhale pressure would have much effect. The lower pressure on exhale might reduce the air swallowing during that part of the cycle but that is just speculation. You still would have the inhale cycle issue.
To get more information, try searching for Aerophagia as a topic as there has been quite a bit of discussion on the topic.
To get more information, try searching for Aerophagia as a topic as there has been quite a bit of discussion on the topic.
_________________
| Mask: Oracle HC452 Oral CPAP Mask |
| Humidifier: DreamStation Heated Humidifier |
| Additional Comments: EverFlo Q 3.0 Liters O2 PR DSX900 ASV |
Oracle 452 Lessons Learned Updated
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
DSX900 AutoSV with HC150 extra humidifier and Hibernite heated hose
Settings: EPAP Min-10.0, EPAP Max-17, PS Min-3, PS Max-10, Max Pressure-20, Rate-Auto, Biflex-1.
Sleepyhead and Encore Pro 2.21.
- ChicagoGranny
- Posts: 15487
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: CPAP Vs BIPAP Swallowing air
Check out the wiki on aerophagia - wiki/index.php/Aerophagia
Re: CPAP Vs BIPAP Swallowing air
What are your current pressure settings now?
Are you using EPR? If so, at what setting?
If you aren't using EPR....turn it on because it will create a bilevel situation similar to a bilevel (bipap is a Respironics marketing term) machine.
Are you using EPR? If so, at what setting?
If you aren't using EPR....turn it on because it will create a bilevel situation similar to a bilevel (bipap is a Respironics marketing term) machine.