Auto Shutoff Stops Working After Mask & Hose Replacement

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
CPAPblows
Posts: 15
Joined: Fri Sep 03, 2010 9:50 pm

Auto Shutoff Stops Working After Mask & Hose Replacement

Post by CPAPblows » Wed Jun 12, 2019 9:54 am

So how do I restore the auto shutoff function on my Resmed Airsense 10 Autoset? It no longer works since I installed replacement gear from Apria.

NoOnesPerfect
Posts: 163
Joined: Thu May 09, 2019 7:29 am
Location: Washington, USA

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by NoOnesPerfect » Wed Jun 12, 2019 9:56 am

Are the replacement mask and hose the same type as you had previously? Different size hoses and different mask styles will likely require setting changes

User avatar
Dog Slobber
Posts: 4323
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by Dog Slobber » Wed Jun 12, 2019 10:19 am

Verify SmartStart is still turned on (Clinical Mode)
Verify Mask type is set correctly (Full, Nasal or Pillows)
Verify Hose is set correctly, if not using ClimateLine (Standard, or SlimLine)(Clinical Mode)
Battery Backup: EcoFlow Delta 2

User avatar
LSAT
Posts: 13399
Joined: Sun Nov 16, 2008 10:11 am
Location: SE Wisconsin

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by LSAT » Wed Jun 12, 2019 12:57 pm

CPAPblows wrote:
Wed Jun 12, 2019 9:54 am
So how do I restore the auto shutoff function on my Resmed Airsense 10 Autoset? It no longer works since I installed replacement gear from Apria.
Auto start and auto shutoff work together. If start works so should shutoff. Change in gear should have made no difference.

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by palerider » Wed Jun 12, 2019 2:17 pm

LSAT wrote:
Wed Jun 12, 2019 12:57 pm
CPAPblows wrote:
Wed Jun 12, 2019 9:54 am
So how do I restore the auto shutoff function on my Resmed Airsense 10 Autoset? It no longer works since I installed replacement gear from Apria.
Auto start and auto shutoff work together. If start works so should shutoff. Change in gear should have made no difference.
Only, auto off is sensitive to the resistance of the circuit, if the flow resistance is too high, then auto off will think it's just a big leak, and not mask off. Even something like small pillows can be enough to fool it, not to mention a tube type being wrong.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by chunkyfrog » Wed Jun 12, 2019 9:49 pm

Aliens.
Keep your pajamas on--especially the bottoms. :shock:

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by zonker » Wed Jun 12, 2019 10:08 pm

chunkyfrog wrote:
Wed Jun 12, 2019 9:49 pm
Aliens.
Keep your pajamas on--especially the bottoms. :shock:

squint.gif

surely, you aren't afraid of probes?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34544
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nowhere special--this year in particular.

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by chunkyfrog » Wed Jun 12, 2019 10:25 pm

Aren't you?
The uninvited kind, of course!

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

User avatar
CPAPblows
Posts: 15
Joined: Fri Sep 03, 2010 9:50 pm

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by CPAPblows » Wed Jun 12, 2019 11:16 pm

The mask, elbow, and hose were the only items replaced. As far as I could tell, they were IDENTICAL to the previous items. Hoses were both slimline. I tried another elbow, and the still had the same symptoms. The new mask was placed open side up on the unit, as it was before the items were replaced. A similar unit did the same thing for a while after an equipment update, but then started shutting off as before. I tried toggling the autostart off and on, but that didn't help.

Could it be that I have to reboot the CPAP unit? Can't believe this is such an uncommon or elusive problem?!

User avatar
RonS
Posts: 272
Joined: Wed Apr 16, 2008 4:43 am

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by RonS » Thu Jun 13, 2019 1:13 am

Have you run mask fit?

Probably won't fix the problem, but you'll feel like you tried something else. ;-)

jsmit86
Posts: 265
Joined: Wed Mar 20, 2019 2:43 pm
Location: Conroe, TX (Suburban Houston TX)

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by jsmit86 » Thu Jun 13, 2019 6:18 am

CPAPblows wrote:
Wed Jun 12, 2019 11:16 pm
The mask, elbow, and hose were the only items replaced. As far as I could tell, they were IDENTICAL to the previous items. Hoses were both slimline. I tried another elbow, and the still had the same symptoms. The new mask was placed open side up on the unit, as it was before the items were replaced. A similar unit did the same thing for a while after an equipment update, but then started shutting off as before. I tried toggling the autostart off and on, but that didn't help.

Could it be that I have to reboot the CPAP unit? Can't believe this is such an uncommon or elusive problem?!
This could be a coincidence.
I just had a similar event on my same unit. I had a tear in a hose, so I replaced it.
At that point auto off stopped working, so I assumed it was related to the hose. I tried different hoses, verified all of the settings as noted in the above posts.
The auto on worked, but the auto off was not working.

The last nights I used the machine, I woke up gasping for breath, so I called the DME who told me to bring the machine in, as something was amiss.
His thought was that likely one of the sensors had failed. I am awaiting a repair now. You probably want to get the unit checked out to be sure.

_________________
Machine: AirSense 11 Autoset
Mask: AirTouch™ F20 Full Face CPAP Mask with Headgear
Additional Comments: also AirFit F20 mask. For travel HDM Z2 Auto CPAP, using AirFit or AirTouch F20 mask. Former Airsense 10 Autoset
Resmed Airsense 11 Autoset - Former Airsense 10 Autoset
Resmed Airtouch F20 Foam Mask
Travel unit HDM Z2 Auto

User avatar
Pugsy
Posts: 65293
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by Pugsy » Thu Jun 13, 2019 6:22 am

Try unplugging from the wall socket electric source and then plug back in.
This will force the machine software to "reboot" so to speak. Similar to what we have to do when computers get a cobweb in them and act squirrelly.

And no...not a common problem when all we do is change hoses or masks.

Then after the reboot see if auto off works...2 tests....one with the mask attached to the long hose and the second if it doesn't work with you removing the mask while it is blowing (unhook it from the long hose) and see if it stops then.
If it doesn't stop with no mask attached then one of the sensors in the machine itself may be ailing.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
Dog Slobber
Posts: 4323
Joined: Thu Feb 15, 2018 2:05 pm
Location: Ontario, Canada

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by Dog Slobber » Thu Jun 13, 2019 8:46 am

Almost always, whenever you suspect something is wrong with an electronic device, unplug it, wait a bit then plug it back in.

Unless it's a life support system, then only do it if you're named as a beneficiary on the insurance.
Battery Backup: EcoFlow Delta 2

User avatar
zonker
Posts: 11433
Joined: Fri Jun 19, 2015 4:36 pm

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by zonker » Thu Jun 13, 2019 10:33 am

Dog Slobber wrote:
Thu Jun 13, 2019 8:46 am
Almost always, whenever you suspect something is wrong with an electronic device, unplug it, wait a bit then plug it back in.

Unless it's a life support system, then only do it if you're named as a beneficiary on the insurance.
no,no,no! you get someone ELSE to do it.

sheesh, haven't you ever watched 70's detective shows?
"Age is not an accomplishment and youth is not a sin"-Robert A. Heinlein
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
Pugsy
Posts: 65293
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: Auto Shutoff Stops Working After Mask & Hose Replacement

Post by Pugsy » Thu Jun 13, 2019 10:55 am

Q. I have stopped therapy, but my device is still blowing air.

A. After therapy has stopped, your device will blow a small amount of air to avoid condensation in the machine and air tube. This will stop automatically after 20 minutes.
Don't mistake reduction in air flow for cool down mode function for not stopping.
Normal function of auto off is to....
Remove mask/hose and after 5 to 10 seconds the air flow gets reduced significantly but doesn't totally stop.
This is called cool down mode.
So it's normal for it to not stop totally but the air flow should drop significantly....when I have the problem with the machine and the P10 mask when I remove the mask...the air flow never reduces at all...it never enters cool down mode.

The reduction in air flow is significant and really can't be missed...it is a really low level of air flow moving through the hose...barely blowing. Totally unlike regular air flow even if a person was using 4 cm...because it's a lot less than 4 cm during cool down mode.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.