Need password for clinician for Transcend Mini Travel Cpap
Need password for clinician for Transcend Mini Travel Cpap
I am desperate for this as the company I bought it from will not help me. I have a collapsing trachea and now need the pressure changed— have to use this all day and Bipap at night.
CAN ANYONE GIVE ME THE PASSWORD—please—?
Jody
CAN ANYONE GIVE ME THE PASSWORD—please—?
Jody
- katestyles
- Posts: 610
- Joined: Sun Jan 06, 2019 9:08 am
Re: Need password for clinician for Transcend Mini Travel Cpap
bump
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: Bleep DreamPort CPAP Mask Solution |
Additional Comments: Back up mask - anything in the drawer |
- ChicagoGranny
- Posts: 14463
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
Re: Need password for clinician for Transcend Mini Travel Cpap
It seems you need to call Somnetics Customer Service 1-877-621-9629 --------->https://www.youtube.com/watch?v=FXj4A1eWgI4
- grayghost4
- Posts: 1554
- Joined: Wed Aug 20, 2014 6:52 pm
- Location: Norther Illinois
- Contact:
Re: Need password for clinician for Transcend Mini Travel Cpap
If you're not part of the solution you're just scumming up the bottom of the beaker!
Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual
Get the Clinicians manual here : http://apneaboard.com/adjust-cpap-press ... tup-manual
Re: Need password for clinician for Transcend Mini Travel Cpap
I tried that and got nowhere;(
Re: Need password for clinician for Transcend Mini Travel Cpap
Granny's Instructions seem pretty clear ...you need to contact customer service to get the password. https://www.youtube.com/watch?v=FXj4A1eWgI4
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
- ChicagoGranny
- Posts: 14463
- Joined: Sun Jan 29, 2012 1:43 pm
- Location: USA
- Wulfman...
- Posts: 6688
- Joined: Mon Sep 01, 2014 6:41 pm
- Location: Nearest fishing spot
Re: Need password for clinician for Transcend Mini Travel Cpap
search.php?keywords=transcend+and+passw ... 614cff9951
The default password always was......"juniper"
Den
.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05
Re: Need password for clinician for Transcend Mini Travel Cpap
In case someone needs this in the future...
We used the info in this thread and another to piece together how to find and use the software needed to change the settings in the Transcend Auto (rev C) machine.
This is the link where we found the user manuals, the clinician manuals, the software and software user manuals (just scroll down to find): https://mytranscend.com/customer-care/user-guides/.
Even though this link says it is the software for the Transcend Mini, it worked for the Transcend Auto (rev C) that I was given: https://transcend.minicpapsoftware.com/. (This was the first software link in the list of three on left side of screen.)
This is the link for the software user manual. It shows the machine I have (Transcend Auto, Rev C). https://mytranscend.com/wp-content/uplo ... -rev-c.pdf HOWEVER, the screens shown in this manual do not match the screens in the software we used.
The software we used (https://mytranscend.com/wp-content/uplo ... -Guide.pdf) says Transcend Micro on it, but it does match the screens of the software we used. More importantly, this is the software that worked for us and the password was still juniper (all lowercase).
With the software and password, we were able to adjust the settings on the Transcend Auto.
Cheers,
Erin
We used the info in this thread and another to piece together how to find and use the software needed to change the settings in the Transcend Auto (rev C) machine.
This is the link where we found the user manuals, the clinician manuals, the software and software user manuals (just scroll down to find): https://mytranscend.com/customer-care/user-guides/.
Even though this link says it is the software for the Transcend Mini, it worked for the Transcend Auto (rev C) that I was given: https://transcend.minicpapsoftware.com/. (This was the first software link in the list of three on left side of screen.)
This is the link for the software user manual. It shows the machine I have (Transcend Auto, Rev C). https://mytranscend.com/wp-content/uplo ... -rev-c.pdf HOWEVER, the screens shown in this manual do not match the screens in the software we used.
The software we used (https://mytranscend.com/wp-content/uplo ... -Guide.pdf) says Transcend Micro on it, but it does match the screens of the software we used. More importantly, this is the software that worked for us and the password was still juniper (all lowercase).
With the software and password, we were able to adjust the settings on the Transcend Auto.
Cheers,
Erin
- chunkyfrog
- Posts: 34391
- Joined: Mon Jul 12, 2010 5:10 pm
- Location: Nebraska--I am sworn to keep the secret of this paradise.
Re: Need password for clinician for Transcend Mini Travel Cpap
juniper
(I did not say it first--just publishing the large print edition.)
(I did not say it first--just publishing the large print edition.)
_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear |
Additional Comments: Airsense 10 Autoset for Her |
Last edited by chunkyfrog on Wed Feb 08, 2023 10:05 pm, edited 2 times in total.
Re: Need password for clinician for Transcend Mini Travel Cpap
Thank you, thank you, thank you!!! I just received a Transcend Micro today, and this information allowed me to change it from a constant-pressure of 10 to an auto-pressure of 10-20. Just what I needed.
fshrchik wrote: ↑Sat Oct 01, 2022 3:11 pmIn case someone needs this in the future...
We used the info in this thread and another to piece together how to find and use the software needed to change the settings in the Transcend Auto (rev C) machine.
This is the link where we found the user manuals, the clinician manuals, the software and software user manuals (just scroll down to find): https://mytranscend.com/customer-care/user-guides/.
Even though this link says it is the software for the Transcend Mini, it worked for the Transcend Auto (rev C) that I was given: https://transcend.minicpapsoftware.com/. (This was the first software link in the list of three on left side of screen.)
This is the link for the software user manual. It shows the machine I have (Transcend Auto, Rev C). https://mytranscend.com/wp-content/uplo ... -rev-c.pdf HOWEVER, the screens shown in this manual do not match the screens in the software we used.
The software we used (https://mytranscend.com/wp-content/uplo ... -Guide.pdf) says Transcend Micro on it, but it does match the screens of the software we used. More importantly, this is the software that worked for us and the password was still juniper (all lowercase).
With the software and password, we were able to adjust the settings on the Transcend Auto.
Cheers,
Erin
_________________
Machine: Somnetics Transcend Micro Travel Machine |
Mask: SleepWeaver Advance Small Soft Cloth Nasal CPAP Mask with Headgear |
Additional Comments: Pressure = 10.5 - 13.5 (Wisp > Quattro FX for Her > SleepWeaver Anew > SleepWeaver Advanced) |
Sleep Study AHI = 70.5
Re: Need password for clinician for Transcend Mini Travel Cpap
I am about to purchase the new Transcend Micro Auto Travel CPAP (the newest one that is the size of a baseball). I want to confirm before I purchase this that the clinical Setting password is still "juniper".
_________________
Machine: Somnetics Transcend Micro Travel Machine |
Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Re: Need password for clinician for Transcend Mini Travel Cpap
I returned the Transcend Mini so I can't verify for certain about the password, however...
In May 2019 Wulfman posted: The default password always was......"juniper"
In October 2022 fshrchik posted: ... the password was still juniper (all lowercase).
My guess would be the password is still "juniper".
Buy your cpap from a vendor who offers a return window, so you are able to return it if needed.
In May 2019 Wulfman posted: The default password always was......"juniper"
In October 2022 fshrchik posted: ... the password was still juniper (all lowercase).
My guess would be the password is still "juniper".
Buy your cpap from a vendor who offers a return window, so you are able to return it if needed.
_________________
Machine: Somnetics Transcend Micro Travel Machine |
Mask: SleepWeaver Advance Small Soft Cloth Nasal CPAP Mask with Headgear |
Additional Comments: Pressure = 10.5 - 13.5 (Wisp > Quattro FX for Her > SleepWeaver Anew > SleepWeaver Advanced) |
Sleep Study AHI = 70.5
Re: Need password for clinician for Transcend Mini Travel Cpap
Did you return the Micro also? I see you signature block says you have the "Machine: Somnetics Transcend Micro Travel Machine"star444 wrote: ↑Tue Feb 07, 2023 9:24 pmI returned the Transcend Mini so I can't verify for certain about the password, however...
In May 2019 Wulfman posted: The default password always was......"juniper"
In October 2022 fshrchik posted: ... the password was still juniper (all lowercase).
My guess would be the password is still "juniper".
Buy your cpap from a vendor who offers a return window, so you are able to return it if needed.
This is the machine I am interested in:
https://www.cpap.com/productpage/somnet ... ravel-cpap
_________________
Machine: Somnetics Transcend Micro Travel Machine |
Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |
Re: Need password for clinician for Transcend Mini Travel Cpap
I received the new machine Somnetics Transcend Micro Travel (baseball) today and can confirm the password is still juniper.wpbear wrote: ↑Wed Feb 08, 2023 7:23 pmDid you return the Micro also? I see you signature block says you have the "Machine: Somnetics Transcend Micro Travel Machine"star444 wrote: ↑Tue Feb 07, 2023 9:24 pmI returned the Transcend Mini so I can't verify for certain about the password, however...
In May 2019 Wulfman posted: The default password always was......"juniper"
In October 2022 fshrchik posted: ... the password was still juniper (all lowercase).
My guess would be the password is still "juniper".
Buy your cpap from a vendor who offers a return window, so you are able to return it if needed.
This is the machine I am interested in:
https://www.cpap.com/productpage/somnet ... ravel-cpap
_________________
Machine: Somnetics Transcend Micro Travel Machine |
Mask: AirFit™ N30i Nasal CPAP Mask with Headgear Starter Pack |