New CPAP user, about 2 weeks, ribs sore , throat sore.

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
zonker
Posts: 11332
Joined: Fri Jun 19, 2015 4:36 pm

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by zonker » Sat Jan 26, 2019 11:26 am

palerider wrote:
Fri Jan 25, 2019 5:00 pm

please read here to see what info the experts want and how they want the info formatted-

viewtopic/t88983/Pugsys-PointersSleepyH ... Hints.html
Good luck with that, was already given that, and ignored it.
[/quote]

i fully understand what you are saying, my friend. but some of us(certainly ME when i first got here!) need a bit more hand holding. let's see how the op does on his next try!
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by palerider » Sat Jan 26, 2019 12:45 pm

zonker wrote:
Sat Jan 26, 2019 11:26 am
palerider wrote:
Fri Jan 25, 2019 5:00 pm
zonker wrote:
Sat Jan 26, 2019 11:26 am
please read here to see what info the experts want and how they want the info formatted-

viewtopic/t88983/Pugsys-PointersSleepyH ... Hints.html
Good luck with that, was already given that, and ignored it.
i fully understand what you are saying, my friend. but some of us(certainly ME when i first got here!) need a bit more hand holding. let's see how the op does on his next try!
I wish you the best of luck in your endeavor.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
zonker
Posts: 11332
Joined: Fri Jun 19, 2015 4:36 pm

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by zonker » Sat Jan 26, 2019 3:11 pm

palerider wrote:
Sat Jan 26, 2019 12:45 pm
I wish you the best of luck in your endeavor.
hat-tip.gif
:wink:
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
TheWeehawky
Posts: 48
Joined: Mon Jan 07, 2019 1:02 pm
Location: Gilmer, TX

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by TheWeehawky » Sat Jan 26, 2019 5:17 pm

zonker wrote:
Sat Jan 26, 2019 11:23 am
TheWeehawky wrote:
Fri Jan 25, 2019 6:37 pm
zonker wrote:
Fri Jan 25, 2019 4:27 pm
TheWeehawky wrote:
Fri Jan 25, 2019 12:02 pm
Here is some of the data off the SDCard.

https://ibb.co/Zz06Xv6
https://ibb.co/f8fNGvp
please read here to see what info the experts want and how they want the info formatted-

viewtopic/t88983/Pugsys-PointersSleepyH ... Hints.html
After following that I see viewtopic/t158560/How-to-post-images-for-review.html .


Then see viewtopic/t103468/Need-help-with-screen-shots.html

It still doesn't fit on one page , 2k/1080p .
okay, try this example-

screenshot-10-25-18.png

note that it only has, from top to bottom, event flags, flow rate, pressure, leak rate and flow limit. you have to move around the various charts to arrange them this way, then take the screenshot.

please be advised that the experts here aren't out to just push you around. they really do need to see your charts this way in order lo lock onto what things you may need to change in order to optimize your therapy.

good luck!
I don't see Pressure, but Time At Pressure.. Flow Limit isn't there either.

Leaving flags, leak rate, snore, resp rate, tidal volume, ispt time , AHI, exp. time , time at pressure.

_________________
MachineMaskHumidifier

User avatar
zonker
Posts: 11332
Joined: Fri Jun 19, 2015 4:36 pm

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by zonker » Sat Jan 26, 2019 7:12 pm

TheWeehawky wrote:
Sat Jan 26, 2019 5:17 pm

I don't see Pressure, but Time At Pressure.. Flow Limit isn't there either.

Leaving flags, leak rate, snore, resp rate, tidal volume, ispt time , AHI, exp. time , time at pressure.
then i bow deeply to you and withdraw. i can only report what sleepyhead shows on my charts. i don't use your machine, so i can't speak to what you see.

someone will be along shortly to help you organize your charts.

good luck!
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
TheWeehawky
Posts: 48
Joined: Mon Jan 07, 2019 1:02 pm
Location: Gilmer, TX

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by TheWeehawky » Sat Jan 26, 2019 7:18 pm

zonker wrote:
Sat Jan 26, 2019 7:12 pm
TheWeehawky wrote:
Sat Jan 26, 2019 5:17 pm

I don't see Pressure, but Time At Pressure.. Flow Limit isn't there either.

Leaving flags, leak rate, snore, resp rate, tidal volume, ispt time , AHI, exp. time , time at pressure.
then i bow deeply to you and withdraw. i can only report what sleepyhead shows on my charts. i don't use your machine, so i can't speak to what you see.

someone will be along shortly to help you organize your charts.

good luck!

_________________
MachineMaskHumidifier
Attachments
ddd.png
ddd.png (200.6 KiB) Viewed 431 times

User avatar
zonker
Posts: 11332
Joined: Fri Jun 19, 2015 4:36 pm

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by zonker » Sat Jan 26, 2019 7:39 pm

hey, you're getting closer!

i know that in your first post of your screenshot, it looked like one of the events was labelled "pressure" but it's so squished i'm not sure.

anyway, keep trying. it took me the LONGEST time to get it right.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
palerider
Posts: 32299
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by palerider » Sat Jan 26, 2019 7:50 pm

zonker wrote:
Sat Jan 26, 2019 7:12 pm
TheWeehawky wrote:
Sat Jan 26, 2019 5:17 pm

I don't see Pressure, but Time At Pressure.. Flow Limit isn't there either.

Leaving flags, leak rate, snore, resp rate, tidal volume, ispt time , AHI, exp. time , time at pressure.
then i bow deeply to you and withdraw. i can only report what sleepyhead shows on my charts. i don't use your machine, so i can't speak to what you see.
But, you've seen plenty Dreamstation reports, and they have Flags, flow, pressure, leaks and snore.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Pugsy
Posts: 65115
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by Pugsy » Sat Jan 26, 2019 8:22 pm

For starters...you have turned off the second session...look down under the session information...see the green box and the red box??
Click on the red box and change it to green.
If you are going to have anything red...make it that top session of only 9 minutes.

Now look at the bottom of the graphs (lower right corner) for the little arrow where it says "Graphs"...click on that arrow and when the menu expands scroll down through all the choices and any that are red..click on it and change it to green.
Do the same thing with the Flag choices lower left corner...anything red click on it and change it to green.

Your machine does indeed have these graphs
Events
Flow Rate
Pressure
Leak

It won't have Flow limitation graphs or snore graphs because those graphs are only found on ResMed machine reports.
Respironics machines don't do the graphs because Flow Limitations and snores are included in the top Events graph instead.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

User avatar
TheWeehawky
Posts: 48
Joined: Mon Jan 07, 2019 1:02 pm
Location: Gilmer, TX

Re: New CPAP user, about 2 weeks, ribs sore , throat sore.

Post by TheWeehawky » Sat Jan 26, 2019 8:56 pm

Pugsy wrote:
Sat Jan 26, 2019 8:22 pm
For starters...you have turned off the second session...look down under the session information...see the green box and the red box??
Click on the red box and change it to green.
If you are going to have anything red...make it that top session of only 9 minutes.

Now look at the bottom of the graphs (lower right corner) for the little arrow where it says "Graphs"...click on that arrow and when the menu expands scroll down through all the choices and any that are red..click on it and change it to green.
Do the same thing with the Flag choices lower left corner...anything red click on it and change it to green.

Your machine does indeed have these graphs
Events
Flow Rate
Pressure
Leak

It won't have Flow limitation graphs or snore graphs because those graphs are only found on ResMed machine reports.
Respironics machines don't do the graphs because Flow Limitations and snores are included in the top Events graph instead.

_________________