Cleaning the Resmed S9 Autoset?
Cleaning the Resmed S9 Autoset?
Hello all, I have a conundrum...
In recent years I've developed a recurring chest infection (bronchitis). It will hang on for 5-6 weeks then slowly improves but typically returns within months. I've seen my GP and I've been on antibiotics, cortisone and pretty much everything else in an attempt to knock it on the head. I was due to have a hip replacement recently but the op was called off due to the state of my chest.
Now what's all this doing on a CPAP forum you might ask. Well I have had a Resmed S9 Autoset for about 8 years now and it's never been repaired or been in for professional maintenance. I change the filter in the rear and clean around the inlet about 3 or 4 times per year but internally it's never been cleaned. Being at my wits end about the allergy like condition that develops with the bronchitis (stuffy nose, scratchy eyes etc.) I was wondering if it might be possible some nasty bacteria might have built up inside the S9. Has anyone heard of anything like that?
The other thing is that I rarely clean the main hose that feeds the headset tubing, is that a problem? I clean the P10 nasal pillow daily and the headset every couple of weeks? I'm probably clutching at straws with this and it's probably not related to my chest problem at all but I thought I'd ask.
Thanks and Happy New year.
In recent years I've developed a recurring chest infection (bronchitis). It will hang on for 5-6 weeks then slowly improves but typically returns within months. I've seen my GP and I've been on antibiotics, cortisone and pretty much everything else in an attempt to knock it on the head. I was due to have a hip replacement recently but the op was called off due to the state of my chest.
Now what's all this doing on a CPAP forum you might ask. Well I have had a Resmed S9 Autoset for about 8 years now and it's never been repaired or been in for professional maintenance. I change the filter in the rear and clean around the inlet about 3 or 4 times per year but internally it's never been cleaned. Being at my wits end about the allergy like condition that develops with the bronchitis (stuffy nose, scratchy eyes etc.) I was wondering if it might be possible some nasty bacteria might have built up inside the S9. Has anyone heard of anything like that?
The other thing is that I rarely clean the main hose that feeds the headset tubing, is that a problem? I clean the P10 nasal pillow daily and the headset every couple of weeks? I'm probably clutching at straws with this and it's probably not related to my chest problem at all but I thought I'd ask.
Thanks and Happy New year.
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Re: Cleaning the Resmed S9 Autoset?
Hi, sorry about your problems - hope you find an answer. First - there is nothing to 'clean' inside your machine and nothing there that would contribute to your ongoing trouble. Really. Very many of us haven't cleaned our hoses for years, decades even, so unless you see something obvious inside, relax.
The thing however, is that these machines usually are considered replaceable by the insce. co's (at least in the U.S.) after 5 yrs, tho' may last twice as long... but yours may be ready for recycling now and at best might be kept as a backup.
Speaking of straws, however, are you using your humidifier at a high setting (or even using it at all) when your local climate may be humid enough not to need as much or as high a setting and possibly contributing to (not necessarily causing) your resp. problems? I rarely use mine at all, others only for parts of the year, and if I were to use it when not needed, I would develop congestion, etc. Maybe something to consider depending on where you live. btw - any chance of mould in your home - it can certainly cause your problems too.
The thing however, is that these machines usually are considered replaceable by the insce. co's (at least in the U.S.) after 5 yrs, tho' may last twice as long... but yours may be ready for recycling now and at best might be kept as a backup.
Speaking of straws, however, are you using your humidifier at a high setting (or even using it at all) when your local climate may be humid enough not to need as much or as high a setting and possibly contributing to (not necessarily causing) your resp. problems? I rarely use mine at all, others only for parts of the year, and if I were to use it when not needed, I would develop congestion, etc. Maybe something to consider depending on where you live. btw - any chance of mould in your home - it can certainly cause your problems too.
Last edited by Julie on Wed Dec 26, 2018 10:19 pm, edited 1 time in total.
Re: Cleaning the Resmed S9 Autoset?
Short answer: No.crookneez wrote: ↑Wed Dec 26, 2018 9:29 pm\. Well I have had a Resmed S9 Autoset for about 8 years now and it's never been repaired or been in for professional maintenance. I change the filter in the rear and clean around the inlet about 3 or 4 times per year but internally it's never been cleaned. Being at my wits end about the allergy like condition that develops with the bronchitis (stuffy nose, scratchy eyes etc.) I was wondering if it might be possible some nasty bacteria might have built up inside the S9. Has anyone heard of anything like that?
Longer answer: Noooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooooo
I've been in the S9, there's nothing in there that would encourage bacteria to grow on... a little foam, a lot of hard plastics, and some silicone. In short, nothing that needs cleaning.
If you want to go ahead and take it apart, though: https://imgur.com/a/S0ojG
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Cleaning the Resmed S9 Autoset?
Thanks Julie (and palerider), I guessed as much about internal cleaning. The S9 doesn't seem to be designed to be opened so I assumed... "Julie wrote: ↑Wed Dec 26, 2018 9:59 pmThe thing however, is that these machines usually are considered replaceable by the insce. co's (at least in the U.S.) after 5 yrs, tho' may last twice as long... but yours may be ready for recycling now and at best might be kept as a backup.
Speaking of straws, however, are you using your humidifier at a high setting (or even using it at all) when your local climate may be humid enough not to need as much or as high a setting and possibly contributing to (not necessarily causing) your resp. problems? I rarely use mine at all, others only for parts of the year, and if I were to use it when not needed, I would develop congestion, etc. Maybe something to consider depending on where you live. btw - any chance of mould in your home - it can certainly cause your problems too.
Mine might even be older than 8 years, I can't remember. I've considered replacing it but I don't like the look of the newer machines. I definitely don't want the S10 with the built in humidifier. For traveling the S9 without humidifier is perfect.
Julie, I never use the humidifier. It has never been used since the day I got the machine and I don't seem to have a problem with it.
Back to the drawing board i guess.
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Re: Cleaning the Resmed S9 Autoset?
The S9 series started coming out in early 2010, fwiw.crookneez wrote: ↑Wed Dec 26, 2018 10:33 pmMine might even be older than 8 years, I can't remember. I've considered replacing it but I don't like the look of the newer machines. I definitely don't want the S10 with the built in humidifier. For traveling the S9 without humidifier is perfect.
Julie, I never use the humidifier. It has never been used since the day I got the machine and I don't seem to have a problem with it.
Back to the drawing board i guess.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Cleaning the Resmed S9 Autoset?
Okay, thanks Palerider.palerider wrote: ↑Wed Dec 26, 2018 10:52 pmThe S9 series started coming out in early 2010, fwiw.crookneez wrote: ↑Wed Dec 26, 2018 10:33 pmMine might even be older than 8 years, I can't remember. I've considered replacing it but I don't like the look of the newer machines. I definitely don't want the S10 with the built in humidifier. For traveling the S9 without humidifier is perfect.
Julie, I never use the humidifier. It has never been used since the day I got the machine and I don't seem to have a problem with it.
Back to the drawing board i guess.
_________________
Machine: AirSense™ 10 CPAP Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
- zoocrewphoto
- Posts: 3732
- Joined: Mon Apr 30, 2012 10:34 pm
- Location: Seatac, WA
Re: Cleaning the Resmed S9 Autoset?
This is probably a long shot, but is there any chance you are being exposed to a chemical that is causing you to get sick? Have you seen any pattern to when you get sick?
I started cpap in April of 2012. In June, I got sick for a few weeks. Then I got sick in July, and the cough lasted 2 months. I was very frustrated,and my doctor was useless. All she did was prescribe cough syrup with codeine. I was coughing day and night. Then I got sick again in October. My dad commented that I got sick after every cat show (I photograph cats, and I was also showing a cat in the household pet class for fun). I responded not ALL the cat shows and listed the ones where I did not get sick. I then realized those were all shows where I rode with somebody else and did not take the family van. Hmm. The van had a chemical spill (liquid fertilizer) that smelled icky. When was that? May. My dad cleaned it up, but the stuff had gone under the passenger seat. Every time I went for a road trip in the van, I came home sick. My dad removed the chair and deep cleaned it. I never got sick after a trip again. That van became MY van in 2014. I drive it everyday.
Anyway, it took awhile to figure out the pattern and realize that I was reacting to a chemical spill that we thought we had cleaned up. Even though I couldn't smell it anymore, I was still very sensitive to it.
I started cpap in April of 2012. In June, I got sick for a few weeks. Then I got sick in July, and the cough lasted 2 months. I was very frustrated,and my doctor was useless. All she did was prescribe cough syrup with codeine. I was coughing day and night. Then I got sick again in October. My dad commented that I got sick after every cat show (I photograph cats, and I was also showing a cat in the household pet class for fun). I responded not ALL the cat shows and listed the ones where I did not get sick. I then realized those were all shows where I rode with somebody else and did not take the family van. Hmm. The van had a chemical spill (liquid fertilizer) that smelled icky. When was that? May. My dad cleaned it up, but the stuff had gone under the passenger seat. Every time I went for a road trip in the van, I came home sick. My dad removed the chair and deep cleaned it. I never got sick after a trip again. That van became MY van in 2014. I drive it everyday.
Anyway, it took awhile to figure out the pattern and realize that I was reacting to a chemical spill that we thought we had cleaned up. Even though I couldn't smell it anymore, I was still very sensitive to it.
_________________
Mask: Quattro™ FX Full Face CPAP Mask with Headgear |
Humidifier: S9™ Series H5i™ Heated Humidifier with Climate Control |
Additional Comments: Resmed S9 autoset pressure range 11-17 |
Who would have thought it would be this challenging to sleep and breathe at the same time?
-
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Cleaning the Resmed S9 Autoset?
There is a side cover for the S10 series for when you do not want to use the humidifier. It makes the machine more compact for traveling too.
https://www.cpap.com/productpage/resmed ... side-cover
https://www.cpap.com/productpage/resmed ... side-cover
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
Re: Cleaning the Resmed S9 Autoset?
I don't see any mention of an S10 in the postprodigyplace wrote: ↑Thu Dec 27, 2018 5:53 amThere is a side cover for the S10 series for when you do not want to use the humidifier. It makes the machine more compact for traveling too.
https://www.cpap.com/productpage/resmed ... side-cover
_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier |
Mask: Fisher & Paykel Vitera Full Face Mask with Headgear (S, M, or L Cushion) |
Additional Comments: Back up is S9 Autoset...... |
- Sheriff Buford
- Posts: 4101
- Joined: Mon Aug 09, 2010 8:01 am
- Location: Kingwood, Texas
Re: Cleaning the Resmed S9 Autoset?
You didn’t refer to your mask. How often do you clean your mask and what method do you use to clean it?
Sheriff
Sheriff
_________________
Machine: AirSense 11 Autoset |
Mask: Mirage Quattro™ Full Face CPAP Mask with Headgear |
-
- Posts: 1795
- Joined: Mon Mar 27, 2017 5:45 pm
- Location: Central Virginia
Re: Cleaning the Resmed S9 Autoset?
Here is a quote from their second post.LSAT wrote: ↑Thu Dec 27, 2018 6:26 amI don't see any mention of an S10 in the postprodigyplace wrote: ↑Thu Dec 27, 2018 5:53 amThere is a side cover for the S10 series for when you do not want to use the humidifier. It makes the machine more compact for traveling too.
https://www.cpap.com/productpage/resmed ... side-cover
I've considered replacing it but I don't like the look of the newer machines. I definitely don't want the S10 with the built in humidifier. For traveling the S9 without humidifier is perfect.
_________________
Machine: AirSense 11 Autoset |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
Additional Comments: ClimateLineAir Heated Tube & Sleepyhead software |
Please visit our sponsor, CPAP.com at https://www.cpap.com/ for all your CPAP needs.
Re: Cleaning the Resmed S9 Autoset?
There's no such thing as a Resmed S10.
Get OSCAR
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.
Re: Cleaning the Resmed S9 Autoset?
I have an S9 Resmed. I weekly run a basin of hot water and a bit of Dawn detergent and put the hose and all parts of my mask as well as the water receptacle for the humidifier in it and soak it for a while. I then thoroughly rinse all in cool water. I hang the hose to dry and lay the rest to dry on a clean towel. I do have a second hose and mask so I can interchange them. I change the filter about once a month. I wash the nose piece (I use Dreamwear undermost cushion) EVERY DAY, and if I forget I use CPAP wipes. I have COPD so try and be careful not to get a URI.
Re: Cleaning the Resmed S9 Autoset?
if i may be so bold as to make a suggestion? unscented baby wipe will do the same cleaning as specific cpap wipes and are cheaper.
_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier |
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear |
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg
-
- Posts: 1
- Joined: Sat Nov 23, 2024 4:37 am
Re: Cleaning the Resmed S9 Autoset?
I have an airsense 10. I've recently discovered blackened areas and small blackish spots on and around and inside the entrance to the passageways they sit on. Inside the machine.
It looks unmistakably like spots of black mould.
The company from whom I rent the machine are pulling out an the stops to get the machine back off me but I want them to test it for mould and they are refusing to directly acknowledge my concerns.
Any thoughts?
It looks unmistakably like spots of black mould.
The company from whom I rent the machine are pulling out an the stops to get the machine back off me but I want them to test it for mould and they are refusing to directly acknowledge my concerns.
Any thoughts?