Posted: Wed Oct 04, 2006 12:56 am
That makes sense, setting it to maybe 15cm Max, did not have trouble with the 7cm Min (that is what my Fisher & Paykel was set at and I am used to that).
Hey...wait a minute...its 3AM what are you doing up? Don't tell me others don't get that euphoric 8 hours solid sleep using CPAP..... j/k.
[quote="CollegeGirl"]Snoredog,
I'm going to have to respectfully disagree here. A range of just 1 cm over titrated pressure is not enough when starting out on an auto, and when there is the possibility of a pressure issue (and from what he describes, re: waking up because of breathing issues), there may very well be. At the very least I would set the range for, say, 5 to 10, or maybe even 5 to 12.
RH, if you can't breathe at 5 cms, you can up that bottom number to 6 or 7, whichever is most comfortable for breathing. I would suggest giving it a wider range, though - you need to find out what's going on, and that's going to be hard if the auto doesn't have "room to breathe," so to speak.
Just my two cents.
--CG
Hey...wait a minute...its 3AM what are you doing up? Don't tell me others don't get that euphoric 8 hours solid sleep using CPAP..... j/k.
[quote="CollegeGirl"]Snoredog,
I'm going to have to respectfully disagree here. A range of just 1 cm over titrated pressure is not enough when starting out on an auto, and when there is the possibility of a pressure issue (and from what he describes, re: waking up because of breathing issues), there may very well be. At the very least I would set the range for, say, 5 to 10, or maybe even 5 to 12.
RH, if you can't breathe at 5 cms, you can up that bottom number to 6 or 7, whichever is most comfortable for breathing. I would suggest giving it a wider range, though - you need to find out what's going on, and that's going to be hard if the auto doesn't have "room to breathe," so to speak.
Just my two cents.
--CG