New Resmed Airsense 10 making wheezing noise on inhaling

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
oreooreo
Posts: 5
Joined: Fri Jun 15, 2018 7:56 am

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by oreooreo » Sun Dec 23, 2018 10:02 am

I am using a climate line hose, now I understand why I don't see the tube setting.

The Buzzing noise is so loud that it even bothers my wife. The buzzing noise is strange some nights it's louder and worse than others. Last night it was really loud and my wife asked me to not sleep with the machine because she coudln't take the buzzing.

I got the new machine in November and I have dealt with it for two months.

Does anyone happen to have any ideas how to get rid of the buzzing noise. I can't believe resmed has a machine with this defect and I am at my wits end. I must use the machine 4 hours a night to comply with the insurance and have them pay for it.

If there is no way to get rid of the buzzing then I am calling Apria to replace the machine.

_________________
MaskHumidifier

User avatar
Pugsy
Posts: 64026
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by Pugsy » Sun Dec 23, 2018 10:19 am

oreooreo wrote:
Sun Dec 23, 2018 10:02 am
I am using a climate line hose, now I understand why I don't see the tube setting.

The Buzzing noise is so loud that it even bothers my wife. The buzzing noise is strange some nights it's louder and worse than others. Last night it was really loud and my wife asked me to not sleep with the machine because she coudln't take the buzzing.

I got the new machine in November and I have dealt with it for two months.

Does anyone happen to have any ideas how to get rid of the buzzing noise. I can't believe resmed has a machine with this defect and I am at my wits end. I must use the machine 4 hours a night to comply with the insurance and have them pay for it.

If there is no way to get rid of the buzzing then I am calling Apria to replace the machine.

Call your supplier and tell them you have a noisy lemon. It happens. Not common but it happens.
When a spouse hears unacceptable machine noises there’s something wrong with the machine and you can’t fix it. Spouses shouldn’t be hearing anything related to machine noise.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

AHI88
Posts: 55
Joined: Sun Nov 25, 2018 4:16 am

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by AHI88 » Mon Dec 24, 2018 11:10 am

Reading this thread I've got exactly the problem. The concern that my first machine did this, it's was replaced, the second machine does the same, unfortunately I've given up the fight and just accepting that this is the machine Noise under normal operation.

Sadly this inhaling whinning anoys my wife too

Very sad.

_________________
MachineMask

paulantognoni
Posts: 1
Joined: Thu Dec 27, 2018 11:20 am

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by paulantognoni » Thu Dec 27, 2018 11:33 am

I was told the whining sound which I also have with my machine at 5000 hours is the blower starting to go...

MichiganMike
Posts: 1
Joined: Mon Dec 31, 2018 9:08 pm

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by MichiganMike » Mon Dec 31, 2018 9:21 pm

:shock: I have now had two Resmed 10 Airsense machines and both of them make the same wheezing sound described by others. The medical supply offered to give me a third and I declined. I am back to using my Resmed S9 unit and am returning the new machine back to the supplier. One of the Med supplier reps noted that the Resmed 10 model does make the wheezing sound on inhale and that she could not understand why the Airsense wouldn’t be quieter since it is technologically advanced? My wife couldn’t sleep hearing this sound, she was agitated by it and I found myself trying to adjust my breathing to help her and this turned into a much bigger problem than I anticipated. My pressure was set at 13 via my sleep study and was just hoping to get a new and quieter machine. Resmed needs to hear these complaints and needs to do something to correct the problem.

StudioGeek
Posts: 1
Joined: Thu Mar 07, 2019 5:37 am

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by StudioGeek » Thu Mar 07, 2019 5:43 am

It’s 3:30am in Los Angeles and I too have a wheezy whale ResMed AirSence 10 APAP Machine. This is day one for me with this machine with over 10 years on the AirMed S9 which I love and thankfully still have operating to revert back to.

I’m hoping this is a bad unit and not the AirSence10’s base operating volume. The machine shouldn’t be making a wheeze every time I breathe. Ridiculous! Back to my provider it’s going tomorrow.

ccat9
Posts: 1
Joined: Fri Mar 08, 2019 3:47 pm

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by ccat9 » Fri Mar 08, 2019 4:03 pm

Intro: Had the Resmed S9 for 10yrs (?) and loved it, but it's going. Switched to Airsense 10 in Dec 17 and in 3 months I have tried 3 of them, all faulty. Machine whine noise! Well, the third one has been sent away to be checked for faults. I have had nothing but trouble with this version. Looks like a fourth try is not an option, Resmed will refund my money and write me off.

Note I have air leak problems, full face mask... seems the new machine is too sensitive for me.

You can reduce noise in the following ways: switch off EPR, switch off SmartStart, make sure hose setting standard/slimline is correct. For whistling check connections, and if it's water it'll clear after 10mins or so.

If machine sounds like a vacuum cleaner (my first try, both blower and motor were bust) then clearly, you have a lemon. But I'm finding with Airsense 10 there's whining and it's coming up the hose so earplugs won't keep the noise out. Oh, I sleep with earplugs, my home has noise issues. Anyway, too much noise for me to sleep.

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by palerider » Fri Mar 08, 2019 4:21 pm

ccat9 wrote:
Fri Mar 08, 2019 4:03 pm
, switch off SmartStart,
Smartstart has nothing to do with it.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

cpaperer
Posts: 1
Joined: Tue Nov 24, 2015 11:33 am

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by cpaperer » Tue Mar 12, 2019 2:21 pm

I had a quiet machine for years, and it started wheezing yesterday. I was annoyed it was not letting me fall asleep so I tried some percussive maintenance, before I remembered that I had "washed" the humidifier because I forgot to fill it and running it dry gets it ganky. When you open up the humidifier, there's a silicone based seal between the top and the bottom. If that seal is not set correctly, air leaks out and causes that wheezing. Open up the humidifier, jiggle it a bit, make sure the bit over the vent hole on the inside is seated properly and then try again. This fix worked for me.

User avatar
tbprojex
Posts: 5
Joined: Thu Oct 24, 2019 6:47 pm
Location: Australasia

Re: Resmed Airsense 10 making breathing noise.

Post by tbprojex » Fri Sep 04, 2020 7:08 pm

I'm also now hearing my own breathing from the device after a year or so not hearing a thing ?
Either my hearing is getting better? or more likely something has change on my Airsence 10.
What is the Tube setting on the ClimateLineAir hose supposed to be Standard or slimline ?
Thanks

Edit: Made the mistake of checking the Tube setting without the ClimateLineAir plugged in "doh"
Okie bipap is 100% correct :D

_________________
MachineMask
Additional Comments: ClimateLineAir Heated Tube
Last edited by tbprojex on Sat Sep 05, 2020 9:41 pm, edited 2 times in total.

User avatar
Okie bipap
Posts: 3553
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by Okie bipap » Fri Sep 04, 2020 8:45 pm

With the climateline hose, the machine will recognize it when you plug it in and automatically make any necessary adjustments.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

Dog Tired
Posts: 72
Joined: Sat Dec 12, 2009 7:41 pm
Location: The Land Downunder (Australia)

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by Dog Tired » Fri Mar 19, 2021 9:40 am

cpaperer wrote:
Tue Mar 12, 2019 2:21 pm
I remembered that I had "washed" the humidifier because I forgot to fill it and running it dry gets it ganky. When you open up the humidifier, there's a silicone based seal between the top and the bottom. If that seal is not set correctly, air leaks out and causes that wheezing. Open up the humidifier, jiggle it a bit, make sure the bit over the vent hole on the inside is seated properly and then try again. This fix worked for me.
As it did for me as well, at 2 in the morning, after several hours of it driving me nuts. I was having visions of the unit’s impending failure and getting ready to dig out my old S9.

So thank you for the very valuable tip - worked like a charm.

Now to try and get back to sleep...

_________________
MachineMask

Roger12
Posts: 1
Joined: Tue Jun 15, 2021 8:23 pm

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by Roger12 » Tue Jun 15, 2021 8:28 pm

My AirSense 10 just started making this noise last night, extremely noisy, keeping both my wife and I awake. (Which is why I searched for and found this thread.) As I looked through the different items, and tried setting it lower, I noticed it was much noisier on the nightstand. In hindsight, we just got new nightstands, and the new nightstand is resonant at just the right frequency and was acting as a sounding board and amplifying the sound significantly. Placing a towel under the unit helped significantly. I may still set it on a lower platform, (on the floor/carpet helped too, but I don't want it drawing in all that dust,) but I thought I'd share that discovery, in hopes that it maybe helps someone else. Good luck!

Roger

User avatar
Okie bipap
Posts: 3553
Joined: Thu Oct 15, 2015 4:14 pm
Location: Central Oklahoma

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by Okie bipap » Wed Jun 16, 2021 2:14 pm

I have placed a thick mouse pad under each of our machines. This makes a great vibration barrier between the machine and the shelf it sits on.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Mask: Evora Full Face Mask - Fitpack
Additional Comments: IPAP 20-25, ps 4, OSCAR software
Growing old is mandatory, but growing up is optional.

User avatar
thesleepermustawake
Posts: 72
Joined: Sun Apr 13, 2014 6:20 am
Location: UK

Re: New Resmed Airsense 10 making wheezing noise on inhaling

Post by thesleepermustawake » Sun Jun 27, 2021 8:26 am

I have the same problem, my machine gets very noisy, and prevents me and my partner from sleeping, ironic since the whole point of it is to help you sleep better. I'll call resmed tomorrow to see if there's anything they're willing to do about it