Posted: Tue Aug 01, 2006 4:57 am
Lotta stuff here, but briefly (HAH!):
There were criteria/recommendations that were given us at ASV start-up:
Indications
CHF NYHF Class III/IV with LVEF <40%
+/- Atrial fibrillation
pCO2 <38 mmHg
CSR by history or PSG
Cyclic hypoxia
Contraindications:
Chronic hypoventilation
Moderate-to-severe COPD (pCO2 >45 mmHg)
Restrictive thoracic or neuromuscular disease
We have added:
Central-oriented AHI refractory to traditional pressure therapy
to address CSBD.
The concept of using ASV for everything is VERY recent, hence my earlier comments, which I still stand by. There's "approved uses", "anecdotal evidence", and "hybrid mask" philosophy.
Now, you can point to triggers, and you can look at the components that perpetuate periodic breathing. The perpetuators compose a short list, namely, the 3 "gains" (plant, feedback, and controller) and circulatory delay (yeah, I'm gonna die with that one). The way I see it, fix CT and you fix half the problem. Fix a gain and you fix all the problem. And in my mind (such that it is), ASV works not because of the effect it has on the apnea, but what happens in the resulting hyperpnea. The "loop" isn't a constant, it oscillates.
OLT, the concept of ASV has similarly been around a while, under a variety of names. It might be a good idea to start out with the basic concepts of ASV, and then we can apply this to what these "servo" machines pick and choose:
Adaptive Support Ventilation
• If no spontaneous effort to breathe, machine delivers required minute ventilation (VE) as pressure control, comprised of pressure support (PS) and rate (f).
• If patient starts to breathe spontaneously, machine reduces f and lowers PS to keep VE above set minimum.
• If spontaneous tidal volume (VT) is > target and f <target, PS is reduced and f is increased.
• If VT > target and f > target, PS is lowered and f is reduced.
• If VT < target and f > target, PS is increased and f is lowered.
• If both VT and f are < target, machine increases f and PS.
Gotta run, a trophy striper is calling me ("SAAAAG, SAAAAAG!")
SAG
Actually, we're a "center" rather than a "clinic." We're picky about that.SWS wrote: 1) what your clinic's patient criteria is for your ASV trial
There were criteria/recommendations that were given us at ASV start-up:
Indications
CHF NYHF Class III/IV with LVEF <40%
+/- Atrial fibrillation
pCO2 <38 mmHg
CSR by history or PSG
Cyclic hypoxia
Contraindications:
Chronic hypoventilation
Moderate-to-severe COPD (pCO2 >45 mmHg)
Restrictive thoracic or neuromuscular disease
We have added:
Central-oriented AHI refractory to traditional pressure therapy
to address CSBD.
The concept of using ASV for everything is VERY recent, hence my earlier comments, which I still stand by. There's "approved uses", "anecdotal evidence", and "hybrid mask" philosophy.
It's a center!! Anyway, ya gotta learn to walk before you can run. The real question is as long as CS-2 and BiPAP-SV have been out, why are almost all the reports still anecdotal? Why didn't they include CS-2 in the CANPAP trial? That would have been the ideal study. I mean, even if SAG is dumber than dirt, what's the hold-up with everybody else?2) why your clinic's ASV patient-trial criteria is implicitly such an extremely narrow subset of patients targeted by Resmed for ASV?
This sounds like an NMI argument coming down the river. And that's one leaky boat.The reason I suspect there are likely several CSA etiologies relates to a multitude of ways in which the human respiratory drive can be dysregulated if genetic diversity is simply parameter-remiss in one of many possible regulating mechanisms. Mother nature's rule is that if it is a physiologic mechanism, it is virtually guaranteed to fail in a certain percentage of humanity. Take any chemoreceptor, and neurological process, any organ and devise a physiologically reasonable way for it to fail. That reasonable model of physiologic failure is almost guaranteed to manifest in some epidemiological percentage IMHO. If you can devise multiple highly reasonable physiologic respiratory drive failure scenarios, there is a better than fair chance more than one CSA etiology exists. Haven't noticed too many physiologic traits that don't manifest with both diversity and failure.
Now, you can point to triggers, and you can look at the components that perpetuate periodic breathing. The perpetuators compose a short list, namely, the 3 "gains" (plant, feedback, and controller) and circulatory delay (yeah, I'm gonna die with that one). The way I see it, fix CT and you fix half the problem. Fix a gain and you fix all the problem. And in my mind (such that it is), ASV works not because of the effect it has on the apnea, but what happens in the resulting hyperpnea. The "loop" isn't a constant, it oscillates.
OLT, the concept of ASV has similarly been around a while, under a variety of names. It might be a good idea to start out with the basic concepts of ASV, and then we can apply this to what these "servo" machines pick and choose:
Adaptive Support Ventilation
• If no spontaneous effort to breathe, machine delivers required minute ventilation (VE) as pressure control, comprised of pressure support (PS) and rate (f).
• If patient starts to breathe spontaneously, machine reduces f and lowers PS to keep VE above set minimum.
• If spontaneous tidal volume (VT) is > target and f <target, PS is reduced and f is increased.
• If VT > target and f > target, PS is lowered and f is reduced.
• If VT < target and f > target, PS is increased and f is lowered.
• If both VT and f are < target, machine increases f and PS.
Gotta run, a trophy striper is calling me ("SAAAAG, SAAAAAG!")
SAG