Re: Why have prescription maximums (at least initially)?
Posted: Sun Oct 30, 2016 8:17 am
PEF,
I'm going to write two responses to what you've said in your last couple of posts. This one is about the "technical" aspects of your therapy and your sleep. The second one will focus on your answers to Sludge's questions and less on the technical stuff.
According to Pugsy, your Flow Limitation graph on the Resmed has quite a bit of activity in it and you also mention UARS as a problem.
Is UARS an official diagnosis? Or are you putting things together from Sleep Interrupted and noticing that your pattern matches what Dr. Park says about UARS?
Now some responses to specific things you've written:
Yes, the S9 doesn't record RERAs and the Flow Limitation graph is more difficult to interpret than the Flow Limitation Tick marks on the DreamStation. But the point of therapy is to get some sleep. And if the DreamStation doesn't feel right, then it's NOT the machine for you.
And because I think you ought to be using the S9 rather than the DreamStation, in this post I'm not going to comment on the DreamStation for the most part. If there is some reason that makes you want to revist the question of whether to use the DreamStation, we can revisit that later. But you've made it clear that you and your brain prefer the S9 AutoSet and I see no good reason to argue with your brain and body's preference.
Now as Pugsy said: Flow limitations seems to be what's driving the S9 to increase your pressure to the max setting each night. I understand this problem well: Flow limitations and the PR search algorithm (which is looking for "baby flow limitations") is what drives my own IPAP to literally sit at max IPAP pressure for 50% or more of the night on most nights.
The question is whether the flow limitations that are showing up in the S9 Flow Limitation graph are "real" in the sense of being caused by an airway that is in danger of collapsing or whether they're caused by something else---something that does not actually respond to additional pressure by getting "better". In my case, I have evidence, gained the hard way, that no matter where I set my max IPAP, the IPAP is going to go up to max pressure. And if the max IPAP is less than about 12 or 12cm, it's likely to stay at max IPAP for a large chunk of the night without making any real difference in my Flow Limitation Index and making me feel a whole lot worse in terms of aerophagia.
Some things that would help me tease out just how bad your flow limitations actually are:
1) A picture of the daily data for the whole night from a bad night on the S9, where you get to define "bad night" in terms that are most important to you, that shows the following in this order:
3) A zoomed in picture of the data from the same bad night on the S9 where you zoom in a part of the graph where the Flow Limitation graph is as "boring" as possible---in other words, a place where you think you were asleep AND the machine is scoring no or very little flow limitations.
If you want spend a bit more time, you could also post the same data from a bad night on the DreamStation where the zooms would be for the part of the graph with the most densely packed FL ticks and a part of the graph with no FL ticks.
The point of all this is to see if we can establish what your normal breathing pattern looks like on the S9. And whether there is a significant visual change that is clearly apparent when the machine decides to start increasing pressure based on activity in the FL graph.
We all have our own circadian rhythm, and some of us just don't get sleepy at a "normal" time. Left to its own devices, most people's bodies will settle down into a roughly 24-hour cycle where they get sleepy at roughly the same time every 24-hours. And while most people get sleepy towards the beginning of the night, not everybody does. For example, left to its own devices, my body would prefer to sleep from 3:30 or 4:00am to about 10:30 or 11:00am, but that doesn't quite work with my day job as a college professor.
So some questions for you:
1) If you could sleep whenever your body and mind naturally got sleepy, when would your sleep window be? In other words, when would you naturally fall asleep without the medication and when would you wake up?
2) When do you need or want your bedtime to be? In other words, when do you need your body to be asleep in order to do the things you need to do in your life?
3) What medication are you taking to help with your sleep onset insomnia? When do you take it? How often do you take it? How well does it work when you take it?
Indeed, it may be that LEAKS are what you need to be focusing your attention on rather than the FL, RERA, and UARS stuff. How often do you have a "really good night" in terms of LEAKS and how often do you have a "really bad night" in terms of LEAKS?
Also it is not a surprise to me that you have more problems overall on nights where it takes you 3-4 hours to fall soundly asleep. My guess is that you probably also have more problems with aerophagia on nights where it takes you 3-4 hours to fall soundly asleep.
Some questions:
1) On the nights when it takes you 3-4 hours to fall asleep, are you lying in bed with the mask on fighting with the machine? If so, you ought to just get out of bed and go do something enjoyable to get your mind off the machine and go back to bed when you start to get a wee bit sleepy.
2) Do you have more aerophagia problems when it takes you 3-4 hours to fall asleep vs. a night where you are lucky enough to fall asleep much quicker?
3) What was your preferred sleep position before you started CPAP? Have you had serious problems trying to make that preferred sleep position work with CPAP?
4) Have you considered running the hose under the covers and next to your body rather than hanging it overhead?
5) What kinds of things have you tried to fix the LEAK problem beyond switching masks?
Think of it this way: If you were using straight CPAP, the pressure curve would be a straight, horizontal line at your (max) pressure for 100% of the night.
And, as you noted, increasing the pressure just so that you're not hugging your max pressure setting did NOT do anything positive for your sleep.
It's time to stop focusing so much on how good/bad the APAP data looks on paper and more on how good/bad you feel in the morning. If you feel decent enough in the morning and your pressure curve is at your max pressure for 50% of the night, why should you give a fig about the pressure graph? Particularly since your current settings are effectively treating the OSA and they are more than likely treating the UARS stuff "good enough"
Some things to think about: The more you worry about your sleep in terms of what's going "wrong" each night, the more likely you are to get "stuck" in very light Stage 1/Stage 2 sleep for very long periods during the night. And not getting past Stage 2 sleep into Stage 3 and REM is going to show up in terms of how lousy you feel the next day.
If you are equally comfortable with and without EPR, then turning EPR off makes sense: It's one less variable your body has to deal with. With EPR off you don't run the risk of feeling like the machine is trying to control your breathing if it has trouble tracking your breathing OR if you are very sensitive to when the pressure is increased during the exhalation-to-inhalation stage of your normal breathing pattern.
That said, some people with aerophagia problems find EPR or bilevel useful because it does minimize the amount of air being blown down your throat during exhalation. For me, the real fix for my very severe aerophagia involved three things: (1)a switch to BiPAP (from an S9 AutoSet with EPR = 3), (2) cognitive behavior therapy to drastically shorten my latency to sleep, and (3) time. You've been at this for a long time and you're still having pretty severe problems. So you might want to seriously consider whether shortening your latency to sleep would help your aerophagia as well as your total collection of sleep problems. And if the aerophagia has been pretty severe the entire time you've been on APAP, you might want to consider alternate xPAP strategies.
Which brings me to my last point in this post:
Have you ever considered trying straight CPAP instead of APAP? It could be that even with your exceptionally tight pressure range of 6.4--7.4 on the S9 that your body would be happier if you just set the thing at 7cm for the whole night. Then there would be one less "variable" that your body would have to deal with during the night. Your AHI would likely go up, but since the S9 is reporting an AHI < 1.0 most nights, you've got room to allow the AHI to go up, particularly if you feel better in spite of a slightly higher AHI.
I'm going to write two responses to what you've said in your last couple of posts. This one is about the "technical" aspects of your therapy and your sleep. The second one will focus on your answers to Sludge's questions and less on the technical stuff.
According to Pugsy, your Flow Limitation graph on the Resmed has quite a bit of activity in it and you also mention UARS as a problem.
Is UARS an official diagnosis? Or are you putting things together from Sleep Interrupted and noticing that your pattern matches what Dr. Park says about UARS?
Now some responses to specific things you've written:
So the answer to the question of which machine to use is pretty simple: Use the S9 since you are more comfortable with it.PEF wrote: The machines FELT very different. I don't know what it was. I used the Dreamstation for maybe a total of one month, split into about 2 weeks each. Both times, I could not get used to it and did not sleep well. My brain cried out for the Resmed. I felt the Resmed was "with" me, sort of "knew" me. I felt a strange sense that the Dreamstation was not in sync with my mind.
Yes, the S9 doesn't record RERAs and the Flow Limitation graph is more difficult to interpret than the Flow Limitation Tick marks on the DreamStation. But the point of therapy is to get some sleep. And if the DreamStation doesn't feel right, then it's NOT the machine for you.
And because I think you ought to be using the S9 rather than the DreamStation, in this post I'm not going to comment on the DreamStation for the most part. If there is some reason that makes you want to revist the question of whether to use the DreamStation, we can revisit that later. But you've made it clear that you and your brain prefer the S9 AutoSet and I see no good reason to argue with your brain and body's preference.
As you already know, the S9 AutoSet does not record RERAs. I believe the S9 AutoSet for Her does record RERAs. But just like the PR DreamStations's RERAs, the "for Her" is inferring that a RERA probably occurred based on the shape of the wave flow. The PR machines are looking for evidence of a "recovery breath" or two following a sequence of flow limited breaths. I would assume that the "S9 AutoSet for Her" does a similar thing.1. I am not after 0 AHI, but I would like it to be below 5 on both machines. But I do believe you that the numbers on both machines are good. I do think that my OSA is treated, but I am not sure about my UARS. That is such a complicated subject. that is why I wonder so much about RERA's and FL's.
Now as Pugsy said: Flow limitations seems to be what's driving the S9 to increase your pressure to the max setting each night. I understand this problem well: Flow limitations and the PR search algorithm (which is looking for "baby flow limitations") is what drives my own IPAP to literally sit at max IPAP pressure for 50% or more of the night on most nights.
The question is whether the flow limitations that are showing up in the S9 Flow Limitation graph are "real" in the sense of being caused by an airway that is in danger of collapsing or whether they're caused by something else---something that does not actually respond to additional pressure by getting "better". In my case, I have evidence, gained the hard way, that no matter where I set my max IPAP, the IPAP is going to go up to max pressure. And if the max IPAP is less than about 12 or 12cm, it's likely to stay at max IPAP for a large chunk of the night without making any real difference in my Flow Limitation Index and making me feel a whole lot worse in terms of aerophagia.
Some things that would help me tease out just how bad your flow limitations actually are:
1) A picture of the daily data for the whole night from a bad night on the S9, where you get to define "bad night" in terms that are most important to you, that shows the following in this order:
- The event chart
The Flow Rate graph
The Flow Limitation graph
The Pressure graph
3) A zoomed in picture of the data from the same bad night on the S9 where you zoom in a part of the graph where the Flow Limitation graph is as "boring" as possible---in other words, a place where you think you were asleep AND the machine is scoring no or very little flow limitations.
If you want spend a bit more time, you could also post the same data from a bad night on the DreamStation where the zooms would be for the part of the graph with the most densely packed FL ticks and a part of the graph with no FL ticks.
The point of all this is to see if we can establish what your normal breathing pattern looks like on the S9. And whether there is a significant visual change that is clearly apparent when the machine decides to start increasing pressure based on activity in the FL graph.
Have you heard of delayed sleep phase?It cannot fix insomnia. My sleep onset insomnia has continued to be worse since I started CPAP. I don't know why because the mask and tape are very comfortable. I guess my brain just uses it as an excuse to do what it wants: to stay awake. I read Dr. Krakow's book and got a lot out of it. But it made me realize that I am not a normal person when it comes to sleep. For example, I just don't get sleepy at night. It takes medication to do that. In the morning, I never remember falling asleep. And if I do get sleepy, it only lasts a few minutes at most.
We all have our own circadian rhythm, and some of us just don't get sleepy at a "normal" time. Left to its own devices, most people's bodies will settle down into a roughly 24-hour cycle where they get sleepy at roughly the same time every 24-hours. And while most people get sleepy towards the beginning of the night, not everybody does. For example, left to its own devices, my body would prefer to sleep from 3:30 or 4:00am to about 10:30 or 11:00am, but that doesn't quite work with my day job as a college professor.
So some questions for you:
1) If you could sleep whenever your body and mind naturally got sleepy, when would your sleep window be? In other words, when would you naturally fall asleep without the medication and when would you wake up?
2) When do you need or want your bedtime to be? In other words, when do you need your body to be asleep in order to do the things you need to do in your life?
3) What medication are you taking to help with your sleep onset insomnia? When do you take it? How often do you take it? How well does it work when you take it?
So LEAKS continue to be a BIG issue for your CPAP therapy.I think the leak problem happens when the insomnia is bad and it takes me 3 or 4 hours to get to sleep, or more. I fall into a DEAD sleep, often grinding my face into the pillow and somehow rolling around on my hose (even with the hanger). My mouth will be dry in the morning and I will be vaguely aware of struggling a few times to keep my mask on and in place. On these nights I get about 4 to 5 hours at most. I usually don't feel great in the morning, but better that I would feel, pre-cpap. It is just frustrating to have a really good night with minimal leaks and the next spending 20% of the night in large leak territory.
When I can get to sleep within an hour, my mask stays in place, mouth leaks are minimal and I wake up feeling really good.
Indeed, it may be that LEAKS are what you need to be focusing your attention on rather than the FL, RERA, and UARS stuff. How often do you have a "really good night" in terms of LEAKS and how often do you have a "really bad night" in terms of LEAKS?
Also it is not a surprise to me that you have more problems overall on nights where it takes you 3-4 hours to fall soundly asleep. My guess is that you probably also have more problems with aerophagia on nights where it takes you 3-4 hours to fall soundly asleep.
Some questions:
1) On the nights when it takes you 3-4 hours to fall asleep, are you lying in bed with the mask on fighting with the machine? If so, you ought to just get out of bed and go do something enjoyable to get your mind off the machine and go back to bed when you start to get a wee bit sleepy.
2) Do you have more aerophagia problems when it takes you 3-4 hours to fall asleep vs. a night where you are lucky enough to fall asleep much quicker?
3) What was your preferred sleep position before you started CPAP? Have you had serious problems trying to make that preferred sleep position work with CPAP?
4) Have you considered running the hose under the covers and next to your body rather than hanging it overhead?
5) What kinds of things have you tried to fix the LEAK problem beyond switching masks?
Again, why are you obsessed with making the pressure graph "look better"?I have been obsessed with raising the pressure, I know that. I am bothered by the pressure graph hugging the top for most of the night. I made the graph look better by raising the pressure to max 10 one night, but did not sleep well, lots of leaks and stomach air. So I will just leave the pressure where I am most comfortable. I agree that comfort is most important. (emphasis added)
Think of it this way: If you were using straight CPAP, the pressure curve would be a straight, horizontal line at your (max) pressure for 100% of the night.
And, as you noted, increasing the pressure just so that you're not hugging your max pressure setting did NOT do anything positive for your sleep.
It's time to stop focusing so much on how good/bad the APAP data looks on paper and more on how good/bad you feel in the morning. If you feel decent enough in the morning and your pressure curve is at your max pressure for 50% of the night, why should you give a fig about the pressure graph? Particularly since your current settings are effectively treating the OSA and they are more than likely treating the UARS stuff "good enough"
Some things to think about: The more you worry about your sleep in terms of what's going "wrong" each night, the more likely you are to get "stuck" in very light Stage 1/Stage 2 sleep for very long periods during the night. And not getting past Stage 2 sleep into Stage 3 and REM is going to show up in terms of how lousy you feel the next day.
Go with COMFORT.To answer question #4, I have both Flex and EPR turned off. I did some experimenting, but noticed no difference in comfort. Do you have any specific advice?
If you are equally comfortable with and without EPR, then turning EPR off makes sense: It's one less variable your body has to deal with. With EPR off you don't run the risk of feeling like the machine is trying to control your breathing if it has trouble tracking your breathing OR if you are very sensitive to when the pressure is increased during the exhalation-to-inhalation stage of your normal breathing pattern.
That said, some people with aerophagia problems find EPR or bilevel useful because it does minimize the amount of air being blown down your throat during exhalation. For me, the real fix for my very severe aerophagia involved three things: (1)a switch to BiPAP (from an S9 AutoSet with EPR = 3), (2) cognitive behavior therapy to drastically shorten my latency to sleep, and (3) time. You've been at this for a long time and you're still having pretty severe problems. So you might want to seriously consider whether shortening your latency to sleep would help your aerophagia as well as your total collection of sleep problems. And if the aerophagia has been pretty severe the entire time you've been on APAP, you might want to consider alternate xPAP strategies.
Which brings me to my last point in this post:
Have you ever considered trying straight CPAP instead of APAP? It could be that even with your exceptionally tight pressure range of 6.4--7.4 on the S9 that your body would be happier if you just set the thing at 7cm for the whole night. Then there would be one less "variable" that your body would have to deal with during the night. Your AHI would likely go up, but since the S9 is reporting an AHI < 1.0 most nights, you've got room to allow the AHI to go up, particularly if you feel better in spite of a slightly higher AHI.
I'm glad to hear you're going to do a sleep log. Just keep it simple so it's not too much of a drag to complete each day.I am in the process of designing a sleep log, based on Robysue's recommendations. It is not as easy as I thought it would be. I hope it will be worth the effort.