Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 4:09 pm
by Vennechris
After two sleep studies showing AHI of 34 CA and 1 OA, I was prescribed the Aircurve 10 ASV, EPAP: 4, Min PS: 6, Max PS: 12.
After 8 months of consistent ASV usage, 8+ hours per night sleep with machine, with consistent AHI 0.0 - 0.4. I am still extremely fatigued, even more so than before treatment. Leak levels are very low. Flow limitation spikes occur throughout the night, but Sleep doc says they are meaningless, sleep doc is out of ideas and has no further help to offer. Consistently waking up exhausted, in a fog, with red glassy eyes, worsening memory problems, headaches. Adding Nuvigil makes slogging through the day bearable, but this is taking a toll. Any help would be greatly appreciated.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 4:15 pm
by palerider
Vennechris wrote:After two sleep studies showing AHI of 34 CA and 1 OA, I was prescribed the Aircurve 10 ASV, EPAP: 4, Min PS: 6, Max PS: 12.
After 8 months of consistent ASV usage, 8+ hours per night sleep with machine, with consistent AHI 0.0 - 0.4. I am still extremely fatigued, even more so than before treatment. Leak levels are very low. Flow limitation spikes occur throughout the night, but Sleep doc says they are meaningless, sleep doc is out of ideas and has no further help to offer. Consistently waking up exhausted, in a fog, with red glassy eyes, worsening memory problems, headaches. Adding Nuvigil makes slogging through the day bearable, but this is taking a toll. Any help would be greatly appreciated.
download sleepyhead, post some charts.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 5:29 pm
by Vennechris
Image

Image

Image

Image

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 5:53 pm
by JDS74
Check your PM.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 6:00 pm
by Sleeprider
Vennechris, obviously there is a disconnect between AHI results and how you feel. see sI ome noticeable flow limitation, and might suggest an increase in EPAP min. Outside of the apnea treatment, have you had workups for testosterone and other possible sources? Are there any other health complications that might contribute to how you feel?

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 6:29 pm
by Vennechris
No other health issues outside of the centrals. Have had nearly a full blood workup, (no vitamin tests), thyroid check, heart echo, etc. Going to see neurologist next week to exclude that. Running out of options, was hoping you folks might see something I don't. Will try raising EPAP to knock out flow limitations.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 7:16 pm
by palerider
I second the notion of raising the epap a bit, if it were me, I'd raise the maxps some, since you appear to be maxing out on that at times.

I'd like to see your minute vent trace.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 7:17 pm
by palerider
Sleeprider wrote:Vennechris, obviously there is a disconnect between AHI results and how you feel. see sI ome noticeable flow limitation, and might suggest an increase in EPAP min. Outside of the apnea treatment, have you had workups for testosterone and other possible sources? Are there any other health complications that might contribute to how you feel?
crazy flow limitations traces seem to be a feature of the resmed ASV machines, I haven't ever seen an explanation for it.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 7:50 pm
by Vennechris
Image

Image

Here is the minute vent

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 8:03 pm
by palerider
Vennechris wrote:Here is the minute vent
it looks ok, I was wanting to see if there were drops in it at any points.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Feb 04, 2016 8:41 pm
by Sleeprider
I have to admit, I'm a bit flummoxed. Even your tidal volume looks great. Hope the increased EPAP helps a bit.

Re: Still extremely fatigued on ASV, help?

Posted: Fri Feb 05, 2016 5:33 am
by Guest
Do any ASV users out there feel that the pressure changes prevent them from reaching a deeper level of sleep?

Re: Still extremely fatigued on ASV, help?

Posted: Fri Feb 05, 2016 2:05 pm
by palerider
Guest wrote:Do any ASV users out there feel that the pressure changes prevent them from reaching a deeper level of sleep?
not as much as having a ton of centrals would.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Mar 18, 2021 6:01 am
by chrisj
Hi Vennechris,

Did you ever get to the bottom of this? Seems a little like my experience too.

Cheers

Vennechris wrote:
Thu Feb 04, 2016 4:09 pm
After two sleep studies showing AHI of 34 CA and 1 OA, I was prescribed the Aircurve 10 ASV, EPAP: 4, Min PS: 6, Max PS: 12.
After 8 months of consistent ASV usage, 8+ hours per night sleep with machine, with consistent AHI 0.0 - 0.4. I am still extremely fatigued, even more so than before treatment. Leak levels are very low. Flow limitation spikes occur throughout the night, but Sleep doc says they are meaningless, sleep doc is out of ideas and has no further help to offer. Consistently waking up exhausted, in a fog, with red glassy eyes, worsening memory problems, headaches. Adding Nuvigil makes slogging through the day bearable, but this is taking a toll. Any help would be greatly appreciated.

Re: Still extremely fatigued on ASV, help?

Posted: Thu Mar 18, 2021 7:56 am
by Julie
They haven't been here since '16... I wouldn't hold my breath.