Re: Wireless SD Card Data-transfer for ResScan / SH
Posted: Tue Feb 09, 2016 5:14 pm
Oltremare wrote:E. reinitializing the FlashAir SD card When Necessary (e.g., after a power outage)
Follow the on-screen prompts to Initialize the card. When instructed, safely remove the SD card as follows: Open Computer, right click the SD card's drive, select Eject and remove the card (multiple times if prompted). When you see "Initialize succeeded," click OK, leave the card in the card reader and close the FlashAirTool. Perform Part A above steps. Sometimes the card gets initialized properly even though "Initialize succeeded" is not displayed.
Papit Thanks for your info.
I still have a doubt: the point "E" of your instructions, it says that if you disconnect the power supply must be reset.
I understood right?
I, every morning, I disconnect my AIRSENSE from the mains and do not think I would want every day to reset my FlashAir!
No easier way to this thing?
Thanks so much
There is an easier way. Leave your machine powered On all the time. There's no need to turn power off to your CPAP. Normal operation of your CPAP machine when you wake up in the morning is to simply turn off the blower, not the entire machine. Don't turn off the machine power or pull the power plug out of the machine or the wall socket. We call that leaving the machine in its "Ready" state which means leave it powered on with the blower off when you're not sleeping. Then you won't need to reinitialize the FlashAir card daily.
Go to a pet shop and ask if they have something to coat your power cable with something s/he won't want to bite on. Thus far I haven't heard of any problems with pets biting power cables.