ResMed AirSense 10 "Smart Start" Doesn't Shut Down

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Reuven Gruber » Tue Aug 18, 2015 1:27 pm

With smart start enabled the machine starts automatically but when I remove my mask (full face mask) the machine should automatically shut down but it keeps running. Any Ideas?

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Krelvin » Tue Aug 18, 2015 1:38 pm

Define running. It's it blowing lots of air or is it doing its cool down where it blows at a very low amount while the humidifier cools down. If the latter, it is supposed to do that. Even if you turn it off at the switch it will fyi the same.

If it continues to blow at a high rate something is not right.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin

User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by palerider » Tue Aug 18, 2015 1:41 pm

Reuven Gruber wrote:With smart start enabled the machine starts automatically but when I remove my mask (full face mask) the machine should automatically shut down but it keeps running. Any Ideas?
I've seen that happen when the mask and tubing settings aren't correct.

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Wulfman... » Tue Aug 18, 2015 1:48 pm

Reuven Gruber wrote:With smart start enabled the machine starts automatically but when I remove my mask (full face mask) the machine should automatically shut down but it keeps running. Any Ideas?
From the manual:

SmartStart is enabled, but the device does not automatically stop when the patient removes their mask
Incompatible mask being used. Only use equipment recommended by ResMed.

Contact ResMed or see http://www.resmed.com for more information.
If the patient is using a nasal pillows mask with set pressure less than 7 cm H2O, SmartStart will not work and should be
disabled.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Jay Aitchsee » Tue Aug 18, 2015 1:50 pm

Clinician's Manual wrote:SmartStart is enabled, but the device does not automatically stop when the patient removes their mask
  • Incompatible mask being used. Only use equipment recommended by ResMed.
    Contact ResMed or see http://www.resmed.com for more
    information.
    If the patient is using a nasal pillows mask with set pressure
    less than 7 cm H2O, SmartStart will not work and should be
    disabled.
Edit: I guess we covered that, didn't we Den?

I'm betting on this: If you are using the humidifier the blower will run for approx. 20 minutes (at a reduced rate) after quitting therapy to prevent condensation in the hose.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Reuven Gruber » Tue Aug 18, 2015 1:56 pm

Krelvin - - It blows hard. I let it go for 1 minute and 10 seconds and it was still blowing so I turned it off manually

Palerider - Settings for the mask and tubing are correct

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Reuven Gruber » Tue Aug 18, 2015 1:58 pm

Mask is ResMed F10 and tubing is ResMed Climateline

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
Pugsy
Posts: 64012
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Pugsy » Tue Aug 18, 2015 2:02 pm

What pressure do you normally use?
Next time it happens look on the LCD screen to see what pressure it is blowing at.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Jay Aitchsee
Posts: 2936
Joined: Sun May 22, 2011 12:47 pm
Location: Southwest Florida

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Jay Aitchsee » Tue Aug 18, 2015 2:06 pm

Reuven Gruber wrote:Mask is ResMed F10 and tubing is ResMed Climateline
Reuven, a while back you said you were using pillows (as it says in your profile). Are you sure the mask setting of the machine is what you are actually using?

Seriously, I would (and do) just use the on/off button.

_________________
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: S9 Auto, P10 mask, P=7.0, EPR3, ResScan 5.3, SleepyHead V1.B2, Windows 10, ZEO, CMS50F, Infrared Video

User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Reuven Gruber » Tue Aug 18, 2015 2:15 pm

My pressure settings are Min 4.0; Max 10.0

I used to use pillows but now I am using the full face mask - I will update my profile

I do turn it off manually but I really want to get the device working correctly as advertised

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
Pugsy
Posts: 64012
Joined: Thu May 14, 2009 9:31 am
Location: Missouri, USA

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Pugsy » Tue Aug 18, 2015 2:24 pm

Can you do an experiment while awake?
Change your minimum pressure from 4 cm to 6 or 7 cm...breathe enough for the machine to come on for maybe a minute...then remove the mask and see if the machine will turn off.

_________________
Machine: AirCurve™ 10 VAuto BiLevel Machine with HumidAir™ Heated Humidifier
Additional Comments: Mask Bleep Eclipse https://bleepsleep.com/the-eclipse/
I may have to RISE but I refuse to SHINE.

If you want to try the Eclipse mask and want a special promo code to get a little off the price...send me a private message.

User avatar
Reuven Gruber
Posts: 39
Joined: Tue Sep 16, 2014 3:29 am
Location: Israel

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Reuven Gruber » Tue Aug 18, 2015 2:34 pm

Pugsy
I set min pressure to 7, ran it for 2 minutes, took the mask off and it still didn't shut down

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: AirFit™ P10 Nasal Pillow CPAP Mask with Headgear
Additional Comments: Pressure limits 4.0 to 8.0, ResScan 5.7 Software, Ruby Chin Strap. Untreated AHI 32; With CPAP 1

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Wulfman... » Tue Aug 18, 2015 2:41 pm

Reuven Gruber wrote:Pugsy
I set min pressure to 7, ran it for 2 minutes, took the mask off and it still didn't shut down
The manual says "less than 7 cm". I suspect that it's "blowing hard" trying to keep up whatever pressure it was at when you took the mask off. Even if it was below 7 cm., it would seem to be blowing hard trying to maintain that pressure.


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Wulfman...
Posts: 6688
Joined: Mon Sep 01, 2014 6:41 pm
Location: Nearest fishing spot

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Wulfman... » Tue Aug 18, 2015 3:02 pm

Reuven Gruber wrote:My pressure settings are Min 4.0; Max 10.0

I used to use pillows but now I am using the full face mask - I will update my profile

I do turn it off manually but I really want to get the device working correctly as advertised
Is the mask setting in your machine correct?


Den

.
(5) REMstar Autos w/C-Flex & (6) REMstar Pro 2 CPAPs w/C-Flex - Pressure Setting = 14 cm.
"Passover" Humidification - ResMed Ultra Mirage FF - Encore Pro w/Card Reader & MyEncore software - Chiroflow pillow
User since 05/14/05

User avatar
Krelvin
Posts: 1977
Joined: Tue Jun 06, 2006 5:23 pm
Location: Metro Phx Area - Dry Heat!

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by Krelvin » Tue Aug 18, 2015 3:05 pm

Mine shuts off in about 5 seconds, min is never less than 10.6. Doesn't matter what type of mask, Resmed or F&P.
Current Settings PS 4.0 over 10.6-18.0 (cmH2O) - Resmed S9 VPAP Auto w/h5i Humidifier - Quattro Air FFM
TNET Sleep Resource Pages - CPAP Machine Database
Put your equip in your Signature - SleepyHead v1.0.0-beta-1
Kevin... alias Krelvin