ResMed AirSense 10 "Smart Start" Doesn't Shut Down

General Discussion on any topic relating to CPAP and/or Sleep Apnea.
User avatar
palerider
Posts: 32300
Joined: Wed Dec 16, 2009 5:43 pm
Location: Dallas(ish).

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by palerider » Mon May 06, 2019 1:17 pm

jschumann wrote:
Mon May 06, 2019 2:42 am
I have Aircurve 10 Vauto

Will try leaksense setting

John
Please update your equipment profile.

You won't find a "leaksense" setting, you will find a "leak alarm".

_________________
Mask: Bleep DreamPort CPAP Mask Solution
Additional Comments: S9 VPAP Auto
Get OSCAR

Accounts to put on the foe list: dataq1, clownbell, gearchange, lynninnj, mper!?, DreamDiver, Geer1, almostadoctor, sleepgeek, ajack, stom, mogy, D.H., They often post misleading, timewasting stuff.

User avatar
jschumann
Posts: 15
Joined: Thu Oct 15, 2009 6:09 pm

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by jschumann » Tue May 07, 2019 2:36 am

Thanks Pugsy, that did the trick!

_________________
MachineMask

PaulKTF
Posts: 190
Joined: Fri Jun 09, 2017 8:57 pm
Location: Wisconsin

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by PaulKTF » Sun Mar 29, 2020 10:47 am

I was having the same issue as the OP and what I did after reading this thread and getting the idea from one of palerider's posst in this topic was I was replaced the air filter (which I think was in a bad position which was apparently causing the issue) with a new one which solved the issue for me. :)

So if you're reading this and having the same issue as the OP, try simply removing and re-inserting your air filter or (if it's dirty) just replacing it with a new air filter and see if that solves the issue. It fixed the problem for me and saved me the hassle of getting my machine inspected and getting a "loaner" machine from my DME supplier.

_________________
Machine: ResMed AirSense™ 10 AutoSet™ CPAP Machine with HumidAir™ Heated Humidifier
Mask: Simplus Full Face CPAP Mask with Headgear
Additional Comments: Pressure is set to 10 cmH2O.
-Paul.

User avatar
chunkyfrog
Posts: 34397
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by chunkyfrog » Sun Mar 29, 2020 7:38 pm

That must have been one grody filter!
In the future, please replace a little more often.

_________________
Mask: AirFit™ P10 For Her Nasal Pillow CPAP Mask with Headgear
Additional Comments: Airsense 10 Autoset for Her

walshdr1
Posts: 1
Joined: Tue May 24, 2022 4:19 pm

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by walshdr1 » Tue May 24, 2022 4:24 pm

I have same cpap. Set 5-20 by doctor. My cpap was (for last 2 years) auto shutting off within seconds of taking off my mask. Few days ago I noticed it was not shutting off at all.

Reading some of this forum someone mentioned swapping out hose. Just did that and problem is fixed. Thanks for who ever gave that recommendation. I hadn’t swapped the hose w my replacement parts.

Out!

User avatar
zonker
Posts: 11048
Joined: Fri Jun 19, 2015 4:36 pm

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by zonker » Tue May 24, 2022 5:01 pm

walshdr1 wrote:
Tue May 24, 2022 4:24 pm
I have same cpap. Set 5-20 by doctor. My cpap was (for last 2 years) auto shutting off within seconds of taking off my mask. Few days ago I noticed it was not shutting off at all.

Reading some of this forum someone mentioned swapping out hose. Just did that and problem is fixed. Thanks for who ever gave that recommendation. I hadn’t swapped the hose w my replacement parts.

Out!
good job!

it's nice to know that some folk actually come here to read the forum first, before asking the question.

and thanks for letting us know it worked.
people say i'm self absorbed.
but that's enough about them.
Oscar-Win
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1-Win64.exe
Oscar-Mac
https://www.apneaboard.com/OSCAR/OSCAR-1.5.1.dmg

User avatar
chunkyfrog
Posts: 34397
Joined: Mon Jul 12, 2010 5:10 pm
Location: Nebraska--I am sworn to keep the secret of this paradise.

Re: ResMed AirSense 10 "Smart Start" Doesn't Shut Down

Post by chunkyfrog » Tue May 24, 2022 5:11 pm

Thank you for the heads-up.
My reserve hose has been sitting in its sealed packaging for at least a year.
Now I know when I probably should change it.
:lol: 8)