Re: Leak Data

Posted: Mon May 11, 2015 11:11 pm
by Pugsy
Imgur.com is less annoying than Photobucket should you ever want to post more images.

Re: Leak Data

Posted: Mon May 11, 2015 11:29 pm
by AfibApnea
A humidifier did not come with the BiPAP unit. (Normally, if one gets a BiPAP unit from Apria is the humidifier included?)

I assumed that the humidifier that came with the CPAP unit would fit the BiPAP since the plug in port (a six prong thingy below the air input) looked like the one in the BiPAP unit. What it says under the unit is: "System One Htd Humid, DOM, REF DS6HFLG, and IP22"

Fortunately, I don't normally use the humidifier. I find it makes the air seem "heavy".

Re: Leak Data

Posted: Tue May 12, 2015 6:32 am
by robysue
AfibApnea wrote:Wow, what a reply!
Now you can see why I wanted you to post the screen shots. By the way, you did an excellent job of marking up the screen shots so that I could tell exactly what you needed help with. Thanks for doing that.
And I was asking this question because I wanted to know whether to report to my RT that--according to the data--my mask was leaking. I was wondering if, given months of use, the rubber on the nasal mask changes (becomes less supple?) thus causing the annoying leaks. On the other hand, it could be that I'm getting older.
If this report is typical, then Large Leaks are not a problem. But as masks age, the cushions can loose their ability to seal. And sometimes the first sign of that are small, annoying leaks that are not large enough to get flagged as an Official Large Leak, but they need to be addressed anyway because small annoying leaks can interfere with our ability to sleep well because they're annoying. (A small leak that's not annoying enough to disrupt your sleep can be ignored. There's no reason to chase leaks that are small enough to NOT affect therapy AND also are NOT annoying enough to be disruptive of your sleep.)
How did you know I was using a PR System One REMstar Bipap Auto SV Advanced device?
The combination of the pressure curve and the leak curves. The combination is a dead giveaway.

Re: Leak Data

Posted: Tue May 12, 2015 6:33 am
by robysue
Pugsy wrote:Imgur.com is less annoying than Photobucket should you ever want to post more images.
Someday, but not until after my classes are over, I need to start using Imgur.com. But that will have to wait for now.

Re: Leak Data

Posted: Tue May 12, 2015 9:17 am
by palerider
AfibApnea wrote:Does the "P" indicate Pro (which appears in some of the other items listed).
the only part that matters is the number, the letters that come after indicate things like whether it has a heated hose, or humidifier, etc...

the 'pro' is printed on the face of the constant pressure machines that aren't bricks, whereas the bricks say 'plus' (even though it isn't).

again, the only bit that matters is the number.

Code: Select all

220;Philips Respironics;;PR System One REMStar 60 Series SE CPAP Machine
221;Philips Respironics;;PR System One REMStar 60 Series SE CPAP Machine
260;Philips Respironics;;PR System One REMstar 60 Series Plus CPAP Machine with C-Flex
261;Philips Respironics;;PR System One REMstar 60 Series Plus CPAP Machine with C-Flex
460;Philips Respironics;;PR System One REMStar 60 Series Pro CPAP Machine
461;Philips Respironics;;PR System One REMStar 60 Series Pro CPAP Machine
560;Philips Respironics;;PR System One REMStar 60 Series Auto CPAP Machine
561;Philips Respironics;;PR System One REMStar 60 Series Auto CPAP Machine
660;Philips Respironics;;PR System One REMStar 60 Series BiPAP Pro with Bi-Flex
661;Philips Respironics;;PR System One REMStar 60 Series BiPAP Pro with Bi-Flex
760;Philips Respironics;;PR System One REMStar 60 Series BiPAP Auto with Bi-Flex
761;Philips Respironics;;PR System One REMStar 60 Series BiPAP Auto with Bi-Flex
960;Philips Respironics;;PR System One REMStar 60 Series BiPAP autoSV Advanced
961;Philips Respironics;;PR System One REMStar 60 Series BiPAP autoSV Advanced
150;Philips Respironics;;PR System One REMstar DS150 CPAP Machine
151;Philips Respironics;;PR System One REMstar DS150 CPAP Machine
250;Philips Respironics;;PR System One REMstar Plus CPAP Machine with C-Flex
251;Philips Respironics;;PR System One REMstar Plus CPAP Machine with C-Flex
450;Philips Respironics;;PR System One REMstar Pro CPAP Machine with C-Flex Plus
451;Philips Respironics;;PR System One REMstar Pro CPAP Machine with C-Flex Plus
450AutoIQ;Philips Respironics;;PR System One REMstar Pro CPAP Machine with AutoIQ
451AutoIQ;Philips Respironics;;PR System One REMstar Pro CPAP Machine with AutoIQ
550;Philips Respironics;;PR System One REMstar Auto CPAP Machine with A-Flex
551;Philips Respironics;;PR System One REMstar Auto CPAP Machine with A-Flex
650;Philips Respironics;;PR System One REMstar BiPAP Pro with Bi-Flex
651;Philips Respironics;;PR System One REMstar BiPAP Pro with Bi-Flex
750;Philips Respironics;;PR System One REMstar BiPAP Auto with Bi-Flex
751;Philips Respironics;;PR System One REMstar BiPAP Auto with Bi-Flex
950;Philips Respironics;;PR System One REMstar BiPAP Auto SV Advanced
951;Philips Respironics;;PR System One REMstar BiPAP Auto SV Advanced
1150;Philips Respironics;;PR System One REMstar BiPAP AVAPS Machine
1151;Philips Respironics;;PR System One REMstar BiPAP AVAPS Machine
1060;Philips Respironics;System One 60 Series BiPAP S/T;
1061;Philips Respironics;System One 60 Series BiPAP S/T;
I think that list is correct

Re: Leak Data

Posted: Tue May 12, 2015 9:20 am
by palerider
Pugsy wrote:Imgur.com is less annoying than Photobucket should you ever want to post more images.
I was coming back to post this, I already gave him the link to krelvin's pages on arranging sleepyhead, taking a screenshot and posting it to imgur, but I'm guessing he didn't follow them through.

again for posterity:

https://sleep.tnet.com/resources/sleepyhead/shorganize
https://sleep.tnet.com/resources/sleepy ... screenshot
https://sleep.tnet.com/reference/tips/imgur

imgur is also much less annoying to LOOK at the pics, less clicks to get to a full size image. (ONE if it's posted right)

Re: Leak Data

Posted: Tue May 12, 2015 9:22 am
by palerider
AfibApnea wrote:A humidifier did not come with the BiPAP unit. (Normally, if one gets a BiPAP unit from Apria is the humidifier included?)

I assumed that the humidifier that came with the CPAP unit would fit the BiPAP since the plug in port (a six prong thingy below the air input) looked like the one in the BiPAP unit. What it says under the unit is: "System One Htd Humid, DOM, REF DS6HFLG, and IP22"

Fortunately, I don't normally use the humidifier. I find it makes the air seem "heavy".
that's probably what the p meant, 'no humidifier.

pugsy is right about what you should put on your equipment listing, the 50 series humidifier that you have listed will not fit the 60 series machine you have.

listing a 50 series humidifier and a 60 series machine leads to confusion because it makes people say "well, which do you have? a 50 or a 60 series'?

Re: Leak Data

Posted: Tue May 12, 2015 9:23 am
by palerider
robysue wrote:
Pugsy wrote:Imgur.com is less annoying than Photobucket should you ever want to post more images.
Someday, but not until after my classes are over, I need to start using Imgur.com. But that will have to wait for now.
it's so much easier, you'll save enough time to make up for any initial learning curve

Re: Leak Data

Posted: Tue May 12, 2015 9:31 am
by Pugsy
If you aren't using a humidifier then just leave that menu selection blank so nothing shows up.
People might wonder a bit as to why you don't (your personal preference is a good enough reason) but it will be less confusing if at least remove the 50 series humidifier choice.
You can always add a comment like "no humidifier needed at this time" in the comments section.
But did you know that with the 60 series heated hose you can actually have moisture without much added warmth to the air temp? So if there comes a time where you feel the added moisture would help your nose...there are ways to get it without having to breathe in sauna like air temps.

I don't care for very warm air at all in the warmer months but I need the warmer air in the winter months because my bedroom is always cold....but my nose desperately needs the added moisture and a lot of it.
So with the heated hose I can use warmer air in the winter and much cooler air in the summer and not have to sacrifice the moisture that my nose needs.
With the PR S1 60 series heated hose you can set it so it adds as little as 4 degrees to the room air temp and still get moisture choices separately.

Re: Leak Data

Posted: Wed May 13, 2015 6:24 pm
by AfibApnea
Now I'm real curious about my humidifier unit. (I had thought of leaving it in my equipment because I might use it when I use my CPAP machine.) Now I'm wondering if I just specified the wrong unit.

Here's its history.
- When I was sent my CPAP machine this humidifier came with it.
- When I was sent my BiPAP ASV machine a humidifier did not come with it.
I remember bringing this up with a RT who said to check the connection (the thingy below the air intake. The air intake on the humidifier is that part that accommodates the air outtake of the respirator unit.) If the humidifier (male) matched the respirator (female) then I was set to go.

It did. Therefore I expected that I lucked out and in fact have a respirator unit that works with the BiPAP machine.

I just tried hooking the two units together. They fit perfectly! Therefore I assume I do have a humidifier that works both with my CPAP and BiPAP machine. (I might even try it out this evening! First I'll have to reposition the respirator so that the humidifier doesn't cantilever over the edge of my bedside shelves. And I have to locate my distilled water...)

So
So, I guess I can assume it's a series 60. The question is: Which piece of data printed on the tags underneath the humidifier indicates (or does not indicate) the series number?
"System One Htd Humid, DOM
Ref: DS6HFLF
SN: H09450787
REV 00 Conforms to UL Std 60601-1
Certified to CAN/CSA Std. C22.2

IP22"

Re: Leak Data

Posted: Wed May 13, 2015 6:35 pm
by palerider
AfibApnea wrote:The question is: Which piece of data printed on the tags underneath the humidifier indicates (or does not indicate) the series number?

Ref: DS6HFLF
that one. just like the REF number on the blower unit.

but it's even easier. the 50 series humidifier has a 4 pin connector, the 60 series humidifer has a 6 pin connetor.

look at the sides of the machines, and the sides of the humidifiers, the pinouts should match.

Re: Leak Data

Posted: Wed May 13, 2015 7:48 pm
by AfibApnea
Wow, I guess I have one.

What's the "heated tube" thing all about? How is it different from PR humidifiers wo the heated tube?

Re: Leak Data

Posted: Wed May 13, 2015 7:53 pm
by Pugsy
The heated hose offers more customization of both the humidity that is delivered and separate customization of air temperature.
Originally thought mainly to prevent condensation in the mask or hose but there are other benefits as well.
Like in the winter I prefer a cold bedroom but I don't like a cold nose and with the heated hose I don't have ice cube nose which is real uncomfortable and what I dealt with without the heated hose.